Safe to add extra air filter? Filter mod question.
-
- Posts: 62
- Joined: Tue Jan 03, 2017 3:41 pm
Safe to add extra air filter? Filter mod question.
I've got the dreamstation auto cpap and my disposable filter needs replacing every week or so or I'll smell dirt coming through my hose and disrupts my sleep. I don't understand why it gets dirty so fast. I keep my room clean and this machine has two filters and this disposable one is second in line. It looks like a hepa filter and the first one is foam.
Any suggestions or hacks I can try? Such as add another piece of foam before it or try washing the hepa disposable filter?
I know modifying anything may change my pressure but am hoping since it's an auto cpap that it will still give me proper levels without damaging my machine either.
Any suggestions or hacks I can try? Such as add another piece of foam before it or try washing the hepa disposable filter?
I know modifying anything may change my pressure but am hoping since it's an auto cpap that it will still give me proper levels without damaging my machine either.
_________________
Machine: DreamStation Auto CPAP Machine |
Mask: AirFit™ N20 Nasal CPAP Mask with Headgear |
Re: Safe to add extra air filter? Filter mod question.
Your time would be better spent cleaning your house, that's the real problem, and you are breathing inside your house all the time you are in it. Jimaxxell6307 wrote:I've got the dreamstation auto cpap and my disposable filter needs replacing every week or so or I'll smell dirt coming through my hose and disrupts my sleep. I don't understand why it gets dirty so fast. I keep my room clean and this machine has two filters and this disposable one is second in line. It looks like a hepa filter and the first one is foam.
Any suggestions or hacks I can try? Such as add another piece of foam before it or try washing the hepa disposable filter?
I know modifying anything may change my pressure but am hoping since it's an auto cpap that it will still give me proper levels without damaging my machine either.
Use data to optimize your xPAP treatment!
"The art of medicine consists in amusing the patient while nature cures the disease." Voltaire
"The art of medicine consists in amusing the patient while nature cures the disease." Voltaire
Re: Safe to add extra air filter? Filter mod question.
the pressure sensor is after the filter, in other words, minor changes there make no difference.axxell6307 wrote: I know modifying anything may change my pressure but am hoping since it's an auto cpap that it will still give me proper levels without damaging my machine either.
Get OSCAR
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
-
- Posts: 62
- Joined: Tue Jan 03, 2017 3:41 pm
Re: Safe to add extra air filter? Filter mod question.
What do you mean???Goofproof wrote:...and you are breathing inside your house all the time you are in it. Jimaxxell6307 wrote:I've got the dreamstation auto cpap and my disposable filter needs replacing every week or so or I'll smell dirt coming through my hose and disrupts my sleep. I don't understand why it gets dirty so fast. I keep my room clean and this machine has two filters and this disposable one is second in line. It looks like a hepa filter and the first one is foam.
Any suggestions or hacks I can try? Such as add another piece of foam before it or try washing the hepa disposable filter?
I know modifying anything may change my pressure but am hoping since it's an auto cpap that it will still give me proper levels without damaging my machine either.
I am disabled and more time cleaning is not going to happen. It's very clean as it is. It's done weekly with touch ups in between. Short of creating a clean room area, there will always be dust and dirt. Thus, the interest in filtration changes...
_________________
Machine: DreamStation Auto CPAP Machine |
Mask: AirFit™ N20 Nasal CPAP Mask with Headgear |
-
- Posts: 62
- Joined: Tue Jan 03, 2017 3:41 pm
Re: Safe to add extra air filter? Filter mod question.
Okay. Thanks.palerider wrote:the pressure sensor is after the filter, in other words, minor changes there make no difference.axxell6307 wrote: I know modifying anything may change my pressure but am hoping since it's an auto cpap that it will still give me proper levels without damaging my machine either.
_________________
Machine: DreamStation Auto CPAP Machine |
Mask: AirFit™ N20 Nasal CPAP Mask with Headgear |
Re: Safe to add extra air filter? Filter mod question.
Do you rinse your foam filter every wk? They're meant to be rinsed out, not just tossed... mine last for months,
Re: Safe to add extra air filter? Filter mod question.
the behavior of one brand of cpap does not necessarily translate accurately to another brand.Julie wrote:Do you rinse your foam filter every wk? They're meant to be rinsed out, not just tossed... mine last for months,
Get OSCAR
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
Re: Safe to add extra air filter? Filter mod question.
The filter will not be Dirty enough to change that often unless the air is dirty. If it's as bad as you say it's not healthy. Three or four weeks is normal time to change the filters.
Use a good sweeper with a hepa filter, Room Fans with a Good HEPA Filter. Have furnace checked for heat exchanger damage or vent piping holes or plugging up. A CO and New Smoke detector might be good.
Odd's are you are obsessing, but if not these are good ideas! The heat exchanger on my 10 year old gas failed I only found out when it's computer kept shutting the unit down.
OH! The good days, burning Coal for heat, clean not a option we lived with it. Having to clean the wallpaper, good old days. Jim
Use a good sweeper with a hepa filter, Room Fans with a Good HEPA Filter. Have furnace checked for heat exchanger damage or vent piping holes or plugging up. A CO and New Smoke detector might be good.
Odd's are you are obsessing, but if not these are good ideas! The heat exchanger on my 10 year old gas failed I only found out when it's computer kept shutting the unit down.
OH! The good days, burning Coal for heat, clean not a option we lived with it. Having to clean the wallpaper, good old days. Jim
Use data to optimize your xPAP treatment!
"The art of medicine consists in amusing the patient while nature cures the disease." Voltaire
"The art of medicine consists in amusing the patient while nature cures the disease." Voltaire
- chunkyfrog
- Posts: 34545
- Joined: Mon Jul 12, 2010 5:10 pm
- Location: Nowhere special--this year in particular.
Re: Safe to add extra air filter? Filter mod question.
Many people never consider a hard floor in the bedroom.
Carpeting is a major source of allergens.
If your bedroom is still carpeted, have it removed.
Any fibers in the bedroom need to be removable and launderable.
Some asthmatics go so far as using an air bed, like in hospitals.
Garment bags can be used to isolate out of season clothes in the closet.
Carpeting is a major source of allergens.
If your bedroom is still carpeted, have it removed.
Any fibers in the bedroom need to be removable and launderable.
Some asthmatics go so far as using an air bed, like in hospitals.
Garment bags can be used to isolate out of season clothes in the closet.
_________________
Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear |
Additional Comments: Airsense 10 Autoset for Her |
Re: Safe to add extra air filter? Filter mod question.
Odd's are you aren't smelling dirt, you are smelling something in the room (house) or the HH. The last time I smelled dirt, I was tilling a garden, and that smelled good! Even better was eating the tomatoes and corn. Jim
Do you have Smoking going on in your home, or have others smoking there. Do You have someone there cooking burning food and pans. (I do ) I don't have a good answer for that problem, I married it! What's so hard turning the fire on the stove down, water boils at 212F no matter how mush heat you add, after that food burns and pans melt.
Do you have Smoking going on in your home, or have others smoking there. Do You have someone there cooking burning food and pans. (I do ) I don't have a good answer for that problem, I married it! What's so hard turning the fire on the stove down, water boils at 212F no matter how mush heat you add, after that food burns and pans melt.
Use data to optimize your xPAP treatment!
"The art of medicine consists in amusing the patient while nature cures the disease." Voltaire
"The art of medicine consists in amusing the patient while nature cures the disease." Voltaire
- chunkyfrog
- Posts: 34545
- Joined: Mon Jul 12, 2010 5:10 pm
- Location: Nowhere special--this year in particular.
Re: Safe to add extra air filter? Filter mod question.
212F? Nothing browns at that temperature.
Can't even make toast.
Can't even make toast.
_________________
Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear |
Additional Comments: Airsense 10 Autoset for Her |
-
- Posts: 1795
- Joined: Mon Mar 27, 2017 5:45 pm
- Location: Central Virginia
Re: Safe to add extra air filter? Filter mod question.
Perhaps that is the case for your machine. Other are different.Goofproof wrote:The filter will not be Dirty enough to change that often unless the air is dirty. If it's as bad as you say it's not healthy. Three or four weeks is normal time to change the filters.
I change my Resmed filter once a month but they recommend twice a month.
_________________
Machine: AirSense 11 Autoset |
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear |
Additional Comments: ClimateLineAir Heated Tube & Sleepyhead software |
Please visit our sponsor, CPAP.com at https://www.cpap.com/ for all your CPAP needs.
- chunkyfrog
- Posts: 34545
- Joined: Mon Jul 12, 2010 5:10 pm
- Location: Nowhere special--this year in particular.
Re: Safe to add extra air filter? Filter mod question.
I change my hypoallergenic filters at 3 weeks, when they begin to look soiled.
Any longer, and I sneeze more than I like to.
One or two feels good, but five or more in a row in a nuisance.
Any longer, and I sneeze more than I like to.
One or two feels good, but five or more in a row in a nuisance.
_________________
Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear |
Additional Comments: Airsense 10 Autoset for Her |
Re: Safe to add extra air filter? Filter mod question.
Hard boiled eggs, asparagus, peas, tomatoes, don't need browning, neither do stainless steel cookware. . I made it as easy as I could, I bought Induction cookers, full temp controlled 100 F to 485 F, ten degree steps, Timers, all new SS cookware some non-stick Nice. But not Burn proof. I can't fault her, I just need to learn how to eat carbon, and we all need more medal in our diets. Jimchunkyfrog wrote:212F? Nothing browns at that temperature.
Can't even make toast.
Use data to optimize your xPAP treatment!
"The art of medicine consists in amusing the patient while nature cures the disease." Voltaire
"The art of medicine consists in amusing the patient while nature cures the disease." Voltaire
Re: Safe to add extra air filter? Filter mod question.
Candles, kerosene heaters, or wood stoves and the like can also emit many particles into the air that would need to be filtered.Goofproof wrote:Do you have Smoking going on in your home