"Gurgling Sound" driving wife crazy

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
#1pipefitter

"Gurgling Sound" driving wife crazy

Post by #1pipefitter » Fri Feb 17, 2017 5:43 pm

I have a Dreamstation Auto Bipap with a P10 nasal pillows mask. My wife tells me that while asleep there are gurgling noises coming from me ( or the bipap) She says it sounds like someone blowing on a straw immersed in water, making bubbles. It wakes her up. I have no idea what is causing it. I don't have any water in my (heated) hose and none comes through to my mask.
Anyone have an idea what is causing this noise?

User avatar
Julie
Posts: 20056
Joined: Tue Feb 28, 2006 12:58 pm

Re: "Gurgling Sound" driving wife crazy

Post by Julie » Fri Feb 17, 2017 5:49 pm

Water in the hose - it's best to locate the machine at or below his head so water will run back to the machine and not to you.

JDS74
Posts: 3397
Joined: Sun Jan 23, 2011 2:57 pm
Location: South Carolina

Re: "Gurgling Sound" driving wife crazy

Post by JDS74 » Fri Feb 17, 2017 6:17 pm

As Julie suggests, the sound is from condensation in the hose at a low spot. Look closely at your hose at the lowest spot and see if there is some water condensed there.
If you lower the cpap machine so that the hose is completely up hill from the machine to you or install a hose hanger that has the same effect, the gurgling sound will go away.

_________________
Mask: Oracle HC452 Oral CPAP Mask
Humidifier: DreamStation Heated Humidifier
Additional Comments: EverFlo Q 3.0 Liters O2 PR DSX900 ASV
Oracle 452 Lessons Learned Updated
DSX900 AutoSV with HC150 extra humidifier and Hibernite heated hose
Settings: EPAP Min-10.0, EPAP Max-17, PS Min-3, PS Max-10, Max Pressure-20, Rate-Auto, Biflex-1.
Sleepyhead and Encore Pro 2.21.

#1pipefitter

Re: "Gurgling Sound" driving wife crazy

Post by #1pipefitter » Fri Feb 17, 2017 6:22 pm

I've checked my hose. There isn't any water in it. None.
The hose is heated as well.

User avatar
LSAT
Posts: 13346
Joined: Sun Nov 16, 2008 10:11 am
Location: SE Wisconsin

Re: "Gurgling Sound" driving wife crazy

Post by LSAT » Fri Feb 17, 2017 6:25 pm

#1pipefitter wrote:I've checked my hose. There isn't any water in it. None.
The hose is heated as well.
All it takes is a very small amount of water to "gurgle".....Follow the suggestions. Buy or make your own hose hanger. Any moisture will run down hill back to the humidifier. https://www.cpap.com/productpage/arden- ... ystem.html

D.H.
Posts: 3532
Joined: Sun Sep 04, 2016 7:07 pm

Re: "Gurgling Sound" driving wife crazy

Post by D.H. » Fri Feb 17, 2017 6:40 pm

It could be that you're overfilling the water tank. Make sure not to go over the line. Also, make sure that the tank is sitting correctly in its place and that the lid is properly closed.

_________________
MachineMask
Additional Comments: Auto PAP; 13.5 cmH2O min - 20 cmH2O max

Guest

Re: "Gurgling Sound" driving wife crazy

Post by Guest » Fri Feb 17, 2017 6:49 pm

#1pipefitter wrote:I've checked my hose. There isn't any water in it. None.
The hose is heated as well.
Disconnect the hose from the cpap then hang it so the entire hose hangs straight down. If no water comes out you better get in to see the doc.

User avatar
Okie bipap
Posts: 3567
Joined: Thu Oct 15, 2015 4:14 pm
Location: Central Oklahoma

Re: "Gurgling Sound" driving wife crazy

Post by Okie bipap » Fri Feb 17, 2017 7:37 pm

If there is any sort of a dip or low spot in the hose, moisture will collect there. I run my hose up over the headboard of my bed in order to keep the entire hose higher than my machine.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Mask: Evora Full Face Mask - Fitpack
Additional Comments: IPAP 20-25, ps 4, OSCAR software
Growing old is mandatory, but growing up is optional.

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: "Gurgling Sound" driving wife crazy

Post by palerider » Fri Feb 17, 2017 8:50 pm

JDS74 wrote:As Julie suggests, the sound is from condensation in the hose at a low spot. Look closely at your hose at the lowest spot and see if there is some water condensed there.
If you lower the cpap machine so that the hose is completely up hill from the machine to you or install a hose hanger that has the same effect, the gurgling sound will go away.
usually water in the hose makes a popping sound, not a gurgling sound. maybe an overfull humidifier?

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
zoocrewphoto
Posts: 3732
Joined: Mon Apr 30, 2012 10:34 pm
Location: Seatac, WA

Re: "Gurgling Sound" driving wife crazy

Post by zoocrewphoto » Sat Feb 18, 2017 1:16 am

I get gurgling in the hose if I stay at my sister's place where the house temp is set lower. Or if I share a hotel room with somebody who runs the air conditioner a lot. To prevent this, I raise the temp of my humidifier settings about 6 degrees. I usually have it set fairly low as I like cool air. That solves my problem completely.

_________________
Mask: Quattro™ FX Full Face CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: Resmed S9 autoset pressure range 11-17
Who would have thought it would be this challenging to sleep and breathe at the same time?

DonnieRB
Posts: 2
Joined: Fri Jan 04, 2019 4:12 am

Re: "Gurgling Sound" driving wife crazy

Post by DonnieRB » Fri Jan 04, 2019 4:38 am

I have been on cpap for 22 years. And for the last year. When i sleep. I have a horrible gurgling clicking noise coming from my throat ! NOT the machine! I have recorded it for my doctor. He never heard a noise like it before. I had a sleep study. They heard it. Turned up the cpap from 8 to 10. Said the noise stopped! NO it didn’t!
So i don’t use the cpap anymore because it keeps the wife awake or she sleeps in the spare room. This is unacceptable to me. My equipment is all new. I use a nose mask. Wondering if a full face mask would do it? Horrible idea! Imput please!

DonnieRB
Posts: 2
Joined: Fri Jan 04, 2019 4:12 am

Re: "Gurgling Sound" driving wife crazy

Post by DonnieRB » Fri Jan 04, 2019 4:41 am

I have an app called Sleep Talk. Sound attavated.

User avatar
remstarcpap
Posts: 318
Joined: Mon Feb 16, 2009 2:17 pm

Re: "Gurgling Sound" driving wife crazy

Post by remstarcpap » Mon Jan 07, 2019 9:52 pm

If no H2O in the hose, then it might be that you are mouth breathing/gurgling. I do that when I don't tape. The other place you can get gurgles is in the mask itself. My P10 gurgles if their is too much humidity. Maybe ask your wife to wake you up in the middle of the night when she hears it, so that you can reproduce it and figure out which of these options are causing the gurgle.

Good gurgle luck!

_________________
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: Tape on mouth

User avatar
chunkyfrog
Posts: 34545
Joined: Mon Jul 12, 2010 5:10 pm
Location: Nowhere special--this year in particular.

Re: "Gurgling Sound" driving wife crazy

Post by chunkyfrog » Mon Jan 07, 2019 10:57 pm

Have her record it on her phone.
God help you if it's a premature death rattle.
(Totally unlikely)

_________________
Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear
Additional Comments: Airsense 10 Autoset for Her

Patpruchnik
Posts: 2
Joined: Sun Dec 01, 2019 9:19 pm

Re: "Gurgling Sound" driving wife crazy

Post by Patpruchnik » Sun Dec 01, 2019 10:10 pm

I have had similar experience with a gurgling sound as I sleep. For years I was able to identify it only through my spouse since it occurs while I am asleep. However, In the recent past I have been able to hear it in the middle of the night while trying to get back to sleep. I believe the cause to be two fold:

1). POSTNASAL drip
2). BACKUP AIR from the CPAP which forces the POSTNASAL drip to reverberate in the back of the throat.

To mitigate the POSTNASAL drip I have increased my daily treatment to be (a) The use of a bed wedge, (b) the use of a saline sinus wash twice a day (I use NielMed), (c). Application of two drops of fluticasone (Flonase) spray in each nostrils at bed time.

I have a Resmed Airfit P10 CPAP.
In order to mitigate the back-flow of air, (a) my sleep doctor has changed a setting that automatically adjusts airflow only as needed with the idea that this might lessen the problem. (B) In addition, I contacted the manufacturer to activate the “pressure relief” option which is intended to make it easier when breathing out. This, in the hope of further reducing the problem.

I believe that these measures have substantially reduced the problem. I say this because the air leakage indicator on my CPAP has dropped into the normal range for the first time in years. I believe this leakage was due to the backed up air escaping through my lips as I slept. But I have to admit that I cannot be completely sure if the gurgling has subsided because my spouse has not yet returned to the bedroom!

Hope this helps.