Yes. The machine will ramp up from the beginning ramp pressure but it won't back down. Until you turn the machine OFF and back ON.Tformhals wrote:Pressure was set at 6-12 with an EPR of 3. I need to check out the auto ramp feature I was probably 5 minutes in and it had jumped to 12. So I turned the ramp time on to 15 mins and fell asleep. I didn't know it had an auto feature. Am I understanding correctly that the machine will ramp up but not back down?
The machine WILL back the pressure all the way down to your minimum pressure setting if the breathing is exceptionally stable and there no events, snoring or flow limitations happening.
Since your titrated pressure was 8cm and you were using 8cm before, I'd suggest thinking about using these settings:
AutoRamp ON
Starting Ramp Pressure 6cm
Min Pressure 7 cm
Max Pressure 10 cm
See if the machine wants to go higher than 10cm when you are asleep. You can always increase the max pressure later if you need to.
As for EPR: Set it where you need to set it in order to be comfortable. Yes, it may be that you'll need a bit more pressure if you use EPR = 2 or 3, but you might not. If you can't tell any difference in comfort when you are trying to get to sleep, then sure, turn it off. But if you are UNCOMFORTABLE with EPR off, then turn it back on to a level that maximizes your comfort.