Do I need a converter for CPAP in Dubai & India?

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
mylescool
Posts: 6
Joined: Thu Oct 20, 2016 1:47 pm

Do I need a converter for CPAP in Dubai & India?

Post by mylescool » Thu Oct 20, 2016 6:16 pm

Travelling to Dubai & India. Don't intend to use CPAP on plane so don't need doctor's note, etc.

But I'm not sure a CPAP 24V made in USA will work in Dubai and India. I have plug adapters but do I need a voltage converter in UAE & India?

User avatar
avi123
Posts: 4509
Joined: Tue Dec 21, 2010 5:39 pm
Location: NC

Re: Do I need a converter for CPAP in Dubai & India?

Post by avi123 » Thu Oct 20, 2016 6:29 pm

Re: Electricity voltage in Dubai
Feb 24, 2009, 5:10 AM
Same as the UK, 3 pin and 220 - 240 volts 50hz

check your AC Adapter, my Resmed 90w AC Adapter says:

AC INPUT 100-140V~50-60Hz. So it fits, I only need the 3 prongs plug adapter.

_________________
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments:  S9 Autoset machine; Ruby chinstrap under the mask straps; ResScan 5.6
Last edited by avi123 on Thu Oct 20, 2016 6:43 pm, edited 1 time in total.
see my recent set-up and Statistics:
http://i.imgur.com/TewT8G9.png
see my recent ResScan treatment results:
http://i.imgur.com/3oia0EY.png
http://i.imgur.com/QEjvlVY.png

mylescool
Posts: 6
Joined: Thu Oct 20, 2016 1:47 pm

Re: Do I need a converter for CPAP in Dubai & India?

Post by mylescool » Thu Oct 20, 2016 6:41 pm

What does "check your power supply (brick)" mean?

User avatar
avi123
Posts: 4509
Joined: Tue Dec 21, 2010 5:39 pm
Location: NC

Re: Do I need a converter for CPAP in Dubai & India?

Post by avi123 » Thu Oct 20, 2016 6:45 pm

mylescool wrote:What does "check your power supply (brick)" mean?
See my new post.

All those countries that were under the British empire use 220 volts, 50 Hertz. If you get a shock there then it will be more than twice as painful as a shock in the USA.

_________________
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments:  S9 Autoset machine; Ruby chinstrap under the mask straps; ResScan 5.6
Last edited by avi123 on Thu Oct 20, 2016 7:14 pm, edited 1 time in total.
see my recent set-up and Statistics:
http://i.imgur.com/TewT8G9.png
see my recent ResScan treatment results:
http://i.imgur.com/3oia0EY.png
http://i.imgur.com/QEjvlVY.png

User avatar
Pugsy
Posts: 65109
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Do I need a converter for CPAP in Dubai & India?

Post by Pugsy » Thu Oct 20, 2016 6:48 pm

What brand machine are you using?
On the bottom or side or the power brick it will tell you what electrical supply it is compatible with.
Almost all machines today that I know of (especially those supplied in the US and the name brand machines) will make the conversion automatically for the difference in electricity and all you need is the adapter for the plugging into the wall socket.

If we knew exactly what brand you are using we could point to the actual specifications for that machine.
You mention 24 V which makes me think ResMed...the older ResMed's didn't have a power brick but the newer ones do but all ResMed's will do the conversion automatically within whatever power supply it uses.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
chunkyfrog
Posts: 34545
Joined: Mon Jul 12, 2010 5:10 pm
Location: Nowhere special--this year in particular.

Re: Do I need a converter for CPAP in Dubai & India?

Post by chunkyfrog » Thu Oct 20, 2016 7:00 pm

He also said made in USA.
All the 24v ones are from Oz, AFAIK, or someplace near there.
All the USA ones I am sure about are 12v.
So, what brand is this unusual machine?

_________________
Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear
Additional Comments: Airsense 10 Autoset for Her

D.H.
Posts: 3532
Joined: Sun Sep 04, 2016 7:07 pm

Re: Do I need a converter for CPAP in Dubai & India?

Post by D.H. » Thu Oct 20, 2016 7:57 pm

This seems to be a duplicate thread. Here's what I posted on the other thread.

Most (all?) CPAP machines are dual-voltage. However, you need to see what it says on you machine to be certain. If your machine has a "brick" (a big block in the middle of the power-cord that looks like a brick), it will be on that. If not, it will be on the bottom or side of your machine.

Power in both India and Dubai is 230 volts 50 hertz. U.S. power is 120 volts 60 hertz. That will most likely be a non-issue. What will be an issue is the shape of the electric sockets (also called "power points"). Please follow the link below to see what you'll need in any country. Also, I pack three normal size extension cords, rather than one long one; I find it more convenient. Note also that since there's a U.K. influence in both these places, what we call an extension cord might be known as an "extension lead." Beware of the "M" type socket as the "universal" adapters won't fit into it.

Here's the last caveat. Most power strips that are sold in the U.S. are also surge protectors and won't work on a 230 volt system.

Link ===> http://www.worldstandards.eu/electricit ... y-country/

_________________
MachineMask
Additional Comments: Auto PAP; 13.5 cmH2O min - 20 cmH2O max
Last edited by D.H. on Fri Oct 21, 2016 5:15 am, edited 1 time in total.

mylescool
Posts: 6
Joined: Thu Oct 20, 2016 1:47 pm

Re: Do I need a converter for CPAP in Dubai & India?

Post by mylescool » Thu Oct 20, 2016 8:07 pm

The machine is a ResMed. Brick says 90W.

User avatar
Pugsy
Posts: 65109
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Do I need a converter for CPAP in Dubai & India?

Post by Pugsy » Thu Oct 20, 2016 8:22 pm

Okay...it has a power brick and ResMed...the S9 and the AirSense use power bricks.

Does it look like this one?
https://www.cpap.com/productpage/resmed ... chine.html
If so here's the spec page
https://www.cpap.com/productpage/resmed ... #specs-tab
and this means you don't need any sort of converter...it can handle it on its own

Code: Select all

Unit Input Range 100-240V, 50-60Hz,
Or does it look like this one
https://www.cpap.com/productpage/resmed ... idair.html
and here's the spec page
https://www.cpap.com/productpage/resmed ... #specs-tab
AC Input Range - 90W Power Supply: 100 - 240V, 50 - 60Hz, 1.0 - 1.5A, Class II
So...all you need is an adapter for the plug to the outlet...if your machine is one of the models above.
No special converter is needed.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

D.H.
Posts: 3532
Joined: Sun Sep 04, 2016 7:07 pm

Re: Do I need a converter for CPAP in Dubai & India?

Post by D.H. » Fri Oct 21, 2016 5:19 am

mylescool wrote:The machine is a ResMed. Brick says 90W.
Turn over the brick, it's on the other side.

_________________
MachineMask
Additional Comments: Auto PAP; 13.5 cmH2O min - 20 cmH2O max

mylescool
Posts: 6
Joined: Thu Oct 20, 2016 1:47 pm

Re: Do I need a converter for CPAP in Dubai & India?

Post by mylescool » Fri Oct 21, 2016 2:22 pm

Machine doesn't look like any that Pugsy posted. This machine was made in Singapore, S9, H5i IPX1.

D.H., there's no more information on the other side of the brick.

User avatar
chunkyfrog
Posts: 34545
Joined: Mon Jul 12, 2010 5:10 pm
Location: Nowhere special--this year in particular.

Re: Do I need a converter for CPAP in Dubai & India?

Post by chunkyfrog » Fri Oct 21, 2016 2:33 pm

The machine that says H5i is the HUMIDIFIER.
The other box is the "BLOWER", or the actual cpap. The identification is on that part.
Right on top, by the ON button, it will likely say either Elite, Autoset, Escape, or Escape Auto.

_________________
Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear
Additional Comments: Airsense 10 Autoset for Her

mylescool
Posts: 6
Joined: Thu Oct 20, 2016 1:47 pm

Re: Do I need a converter for CPAP in Dubai & India?

Post by mylescool » Fri Oct 21, 2016 2:41 pm

It's Escape

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Do I need a converter for CPAP in Dubai & India?

Post by palerider » Fri Oct 21, 2016 3:12 pm

chunkyfrog wrote:The machine that says H5i is the HUMIDIFIER.
The other box is the "BLOWER", or the actual cpap. The identification is on that part.
Right on top, by the ON button, it will likely say either Elite, Autoset, Escape, or Escape Auto.
doesn't matter, they all use the same power brick... and if the OP were to read the fine print on the brick, all the questions would be answered.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
Pugsy
Posts: 65109
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Do I need a converter for CPAP in Dubai & India?

Post by Pugsy » Fri Oct 21, 2016 3:23 pm

mylescool wrote:This machine was made in Singapore, S9, H5i IPX1.
It's the S9 model. The first link I posted was to the S9 Elite...you have a S9 Escape...the power specs are the same no matter if it is an Elite or Escape or whatever S9 model.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.