?difference between Resmed S9 Autoset/S9 Escape Auto

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
User avatar
NormagiJoy
Posts: 41
Joined: Tue Jun 21, 2016 6:16 pm

?difference between Resmed S9 Autoset/S9 Escape Auto

Post by NormagiJoy » Tue Sep 27, 2016 9:35 am

I have found a *gently used* (low hours) Resmed S9. That's all it says, Resmed S9.

looked online and there are lots of different S9 machines!!!

I have yet to get the model number, waiting for an email response.

User avatar
Pugsy
Posts: 65399
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: ?difference between Resmed S9 Autoset/S9 Escape Auto

Post by Pugsy » Tue Sep 27, 2016 9:43 am

S9 models in the cpap/apap line

S9 Escape.....avoid ...no useful data at all
S9 Escape Auto...avoid...while it does offer a generic AHI (no breakdown into each category) that is all it offers and the working algorithm is the "old" algorithm

Full data S9 cpap/apap
S9 Elite....fixed cpap mode only
S9 AutoSet (and the pink AutoSet for Her)...has both cpap mode and apap mode.

For what you are wanting to do the AutoSet would be your best choice but it could be done with the Elite (take a little more time and work).
ResMed machines have the model name clearly stuck on the face of the machine so you are less likely to have to go the model number route if you have a nice clear picture of the top of the machine.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
NormagiJoy
Posts: 41
Joined: Tue Jun 21, 2016 6:16 pm

Re: ?difference between Resmed S9 Autoset/S9 Escape Auto

Post by NormagiJoy » Tue Sep 27, 2016 10:34 am

I knew I should have come here first.

Spent hours looking online trying to make comparisons of the different models,
now I have a headache!!!

if it's not an autoset, I'm not getting it.

THANK YOU!!!

User avatar
knothead
Posts: 538
Joined: Fri Jan 24, 2014 8:58 am
Location: Corpus Christi, Tx

Re: ?difference between Resmed S9 Autoset/S9 Escape Auto

Post by knothead » Tue Sep 27, 2016 10:55 am

Call these people, they will help you with a AutoSet...

http://oklahomacity.craigslist.org/hab/5778077367.html

_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier
Mask: Mirage Quattro™ Full Face CPAP Mask with Headgear
Additional Comments: BP down & in control, Resmed S9 Autoset for backup & travel.

User avatar
palerider
Posts: 32308
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: ?difference between Resmed S9 Autoset/S9 Escape Auto

Post by palerider » Tue Sep 27, 2016 12:44 pm

PM Grayghost4, see what he might be able to do.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
NormagiJoy
Posts: 41
Joined: Tue Jun 21, 2016 6:16 pm

Re: ?difference between Resmed S9 Autoset/S9 Escape Auto

Post by NormagiJoy » Tue Sep 27, 2016 12:57 pm

thanks for the heads up on craigslist and Grayghost. If this is not an autoset, I will check them out!!!

User avatar
JimW159
Posts: 804
Joined: Fri Jul 08, 2016 10:26 am
Location: Tarpon Springs, FL

Re: ?difference between Resmed S9 Autoset/S9 Escape Auto

Post by JimW159 » Tue Sep 27, 2016 3:05 pm

If the seller will provide clear pictures of the top of the blower half of the S9, you will see a word printed above the knob on the upper left. If is says "escape," so should you - very minimal data to assist you. If it says "Elite," it is a fully data-capable single pressure CPAP and very well respected; if "AutoSet," it is a fully data-capable APAP and by all accounts, an excellent machine with many comfort and performance features. I have one that I no longer use but has served me well - I replaced it with an AirSense 10 Auto; in terms of ease of use and comfort I find them both very similar. The AS10 has a significantly smaller footprint, but the S9 is perhaps a bit more flexible in being able to shed the humidifier if you wish.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Mask: Brevida™ Nasal Pillow CPAP Mask with Headgear
Additional Comments: BU Mach AirSense 10 AutoSet - Mask = F&P Brevida BU ResMed P10 - Pressure = 10-17 On CPAP since 12/05/2008 Prior ID on CPAPTalk.com = JimW203

User avatar
LSAT
Posts: 13412
Joined: Sun Nov 16, 2008 10:11 am
Location: SE Wisconsin

Re: ?difference between Resmed S9 Autoset/S9 Escape Auto

Post by LSAT » Tue Sep 27, 2016 4:00 pm

NormagiJoy wrote:I have found a *gently used* (low hours) Resmed S9. That's all it says, Resmed S9.

looked online and there are lots of different S9 machines!!!

I have yet to get the model number, waiting for an email response.
Have you asked what "gently used (low hours) means? Ask what the run hours shown on the screen shows.

User avatar
NormagiJoy
Posts: 41
Joined: Tue Jun 21, 2016 6:16 pm

Re: ?difference between Resmed S9 Autoset/S9 Escape Auto

Post by NormagiJoy » Tue Sep 27, 2016 4:23 pm

I was looking at the cpap assistance program offered by the American Sleep Apnea Association, they confirmed it is a S9 autoset, so for $100 *donation* I got it, humidifer, filters, memory card, tubing and Quattro FX mask.

Hope I did ok???

I definitely did not want the S8 (with no humidifier).

User avatar
palerider
Posts: 32308
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: ?difference between Resmed S9 Autoset/S9 Escape Auto

Post by palerider » Tue Sep 27, 2016 4:36 pm

NormagiJoy wrote:I was looking at the cpap assistance program offered by the American Sleep Apnea Association, they confirmed it is a S9 autoset, so for $100 *donation* I got it, humidifer, filters, memory card, tubing and Quattro FX mask.

Hope I did ok???
the autoset is fine, as long as it says "autoset" not "escape auto"

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
knothead
Posts: 538
Joined: Fri Jan 24, 2014 8:58 am
Location: Corpus Christi, Tx

Re: ?difference between Resmed S9 Autoset/S9 Escape Auto

Post by knothead » Wed Sep 28, 2016 2:51 pm

S9 Autoset great machine. You can even spruce it up with Skinit stickers. Mine is covered with Texas A&M to remind me everyday where my money went...LOL

_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier
Mask: Mirage Quattro™ Full Face CPAP Mask with Headgear
Additional Comments: BP down & in control, Resmed S9 Autoset for backup & travel.

User avatar
NormagiJoy
Posts: 41
Joined: Tue Jun 21, 2016 6:16 pm

Re: ?difference between Resmed S9 Autoset/S9 Escape Auto

Post by NormagiJoy » Wed Sep 28, 2016 3:49 pm

" spruce it up with Skinit stickers"

I forgot about that!!! Do they still make them?

the only one I could find was on ebay....

User avatar
Pugsy
Posts: 65399
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: ?difference between Resmed S9 Autoset/S9 Escape Auto

Post by Pugsy » Wed Sep 28, 2016 3:56 pm

NormagiJoy wrote:" spruce it up with Skinit stickers"

I forgot about that!!! Do they still make them
Not long ago I was looking at the skin it website and all I could find was the ability to customize a skin for the S9.
They used to have a large inventory to choose from but all I could find recently was the "make your own" which was/is a little more costly than the regular ones.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
palerider
Posts: 32308
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: ?difference between Resmed S9 Autoset/S9 Escape Auto

Post by palerider » Wed Sep 28, 2016 4:36 pm

Pugsy wrote:
NormagiJoy wrote:" spruce it up with Skinit stickers"

I forgot about that!!! Do they still make them
Not long ago I was looking at the skin it website and all I could find was the ability to customize a skin for the S9.
They used to have a large inventory to choose from but all I could find recently was the "make your own" which was/is a little more costly than the regular ones.
but, the cool thing is that you can put *anything* on there

I did one for a friend that loves bunnies, so he's got bunnies on his, and did another for another friend that has tigers.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
NormagiJoy
Posts: 41
Joined: Tue Jun 21, 2016 6:16 pm

Re: ?difference between Resmed S9 Autoset/S9 Escape Auto

Post by NormagiJoy » Wed Sep 28, 2016 4:41 pm

if I decide to decorate mine, it will probably be stickers from the dollar store