AHI measured by CPAP devices

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
seekingwellness2016
Posts: 18
Joined: Sun Sep 04, 2016 3:49 pm

AHI measured by CPAP devices

Post by seekingwellness2016 » Tue Sep 13, 2016 7:21 am

How does a device like the ResMed AirSense 10 actually detect apneas/hypopneas, in order to define the AHI? Don't these by definition require measurement of oxygen desaturations and/or arousals? How can the estimated AHI possibly be accurate without measuring these parameters?

User avatar
LSAT
Posts: 13412
Joined: Sun Nov 16, 2008 10:11 am
Location: SE Wisconsin

Re: AHI measured by CPAP devices

Post by LSAT » Tue Sep 13, 2016 8:28 am

seekingwellness2016 wrote:How does a device like the ResMed AirSense 10 actually detect apneas/hypopneas, in order to define the AHI? Don't these by definition require measurement of oxygen desaturations and/or arousals? How can the estimated AHI possibly be accurate without measuring these parameters?
AHI is a measurement of apneas and hypopneas. An apnea is a full blockage of air for 10 seconds or longer...a hypopnea is a partial blockage of air for 10 seconds or longer. The CPAP device detects these blockages. That is what is measured. AHI does not measure O2. O2 along with AHI is used to determine Sleep Apnea.

User avatar
Pugsy
Posts: 65399
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: AHI measured by CPAP devices

Post by Pugsy » Tue Sep 13, 2016 8:38 am

These machines measure air flow or the reduction or lack of air flow.
While oxygen levels might be involved in the diagnostic sleep study to meet diagnosis criteria they aren't needed by the machine to measure air flow or to record the lack of or reduction in air flow using the criteria predetermined by the machine's algorithm for flagging apnea events.
The criteria is a certain % of reduction in air flow that lasts at least 10 seconds. Oxygen levels aren't needed for this flagging.
You don't need oxygen levels to get an AHI result.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

seekingwellness2016
Posts: 18
Joined: Sun Sep 04, 2016 3:49 pm

Re: AHI measured by CPAP devices

Post by seekingwellness2016 » Tue Sep 13, 2016 8:52 am

I guess the bottom line is that there are many different definitions of AHI, because there are different definitions of what constitutes an apnea or hypopnea.

User avatar
BlackSpinner
Posts: 9742
Joined: Sat Apr 25, 2009 5:44 pm
Location: Edmonton Alberta
Contact:

Re: AHI measured by CPAP devices

Post by BlackSpinner » Tue Sep 13, 2016 9:04 am

seekingwellness2016 wrote:I guess the bottom line is that there are many different definitions of AHI, because there are different definitions of what constitutes an apnea or hypopnea.
No there is only one. 10 secs or more of obstruction of air flow. It is very specific.
Drops in oxygen are not required.

_________________
Machine: PR System One REMStar 60 Series Auto CPAP Machine
Additional Comments: Quatro mask for colds & flus S8 elite for back up
71. The lame can ride on horseback, the one-handed drive cattle. The deaf, fight and be useful. To be blind is better than to be burnt on the pyre. No one gets good from a corpse. The Havamal

User avatar
chunkyfrog
Posts: 34544
Joined: Mon Jul 12, 2010 5:10 pm
Location: Nowhere special--this year in particular.

Re: AHI measured by CPAP devices

Post by chunkyfrog » Tue Sep 13, 2016 9:06 am

Google "apnea-hypopnea index".
Wikipedia's explanation is quite clear and fairly simple.
As you can plainly see, oxygen saturation is ANOTHER criteria to describe severity of apnea.
Both are important, but they are not the same thing.
Even without desaturation, there are negative effects on the body.

_________________
Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear
Additional Comments: Airsense 10 Autoset for Her

User avatar
LSAT
Posts: 13412
Joined: Sun Nov 16, 2008 10:11 am
Location: SE Wisconsin

Re: AHI measured by CPAP devices

Post by LSAT » Tue Sep 13, 2016 9:54 am

seekingwellness2016 wrote:I guess the bottom line is that there are many different definitions of AHI, because there are different definitions of what constitutes an apnea or hypopnea.
WRONG! There is only 1. Where are you finding YOUR information?

seekingwellness2016
Posts: 18
Joined: Sun Sep 04, 2016 3:49 pm

Re: AHI measured by CPAP devices

Post by seekingwellness2016 » Tue Sep 13, 2016 11:27 am

I'm done some of my own reading now on this topic, and am satisfied with the conclusions I've been able to draw myself.

For those who are interested, there are in fact several different definitions of AHI.

See, for example, the following abstract/article:

http://www.ncbi.nlm.nih.gov/pubmed/19238801

where AHI(recommended), AHI(alternative), and AHI(Chicago) are all described.

User avatar
Goofproof
Posts: 16082
Joined: Mon Dec 05, 2005 3:16 pm
Location: Central Indiana, USA

Re: AHI measured by CPAP devices

Post by Goofproof » Tue Sep 13, 2016 11:44 am

Congrats, you learned to read, now it's time for realizing everything you read isn't always true, As long as people get paid they are going to publish papers, it's how they stroke themselves and get paid. The government pays be the word, your tax dollars, it doesn't have to make sense or be true. Jim
Use data to optimize your xPAP treatment!

"The art of medicine consists in amusing the patient while nature cures the disease." Voltaire

User avatar
Julie
Posts: 20051
Joined: Tue Feb 28, 2006 12:58 pm

Re: AHI measured by CPAP devices

Post by Julie » Tue Sep 13, 2016 11:52 am

There are posters here who've been here for years, knowledgable and very experienced in dealing with OSA/Cpap, some could easily give much better courses on AHI etc. than many/most of the authors of old or just badly researched papers you've found to prove what's unproveable! You're just plain WRONG about AHI (the most basic and easy aspect of Cpap features) and you're doing yourself an injustice by wasting your time with that stuff trying to convince the people here who've worked with it all forever. Use your time more productively! Stop trying to justify a mistake!
Last edited by Julie on Tue Sep 13, 2016 12:02 pm, edited 1 time in total.

User avatar
BlackSpinner
Posts: 9742
Joined: Sat Apr 25, 2009 5:44 pm
Location: Edmonton Alberta
Contact:

Re: AHI measured by CPAP devices

Post by BlackSpinner » Tue Sep 13, 2016 12:02 pm

seekingwellness2016 wrote:I'm done some of my own reading now on this topic, and am satisfied with the conclusions I've been able to draw myself.

For those who are interested, there are in fact several different definitions of AHI.

See, for example, the following abstract/article:

http://www.ncbi.nlm.nih.gov/pubmed/19238801

where AHI(recommended), AHI(alternative), and AHI(Chicago) are all described.
Nobody these days cares about this anymore. All sleep studies use 10 seconds or more of obstruction only. All cpap machines use the same. Industry standards prevail. Suck it up.

_________________
Machine: PR System One REMStar 60 Series Auto CPAP Machine
Additional Comments: Quatro mask for colds & flus S8 elite for back up
71. The lame can ride on horseback, the one-handed drive cattle. The deaf, fight and be useful. To be blind is better than to be burnt on the pyre. No one gets good from a corpse. The Havamal

User avatar
palerider
Posts: 32308
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: AHI measured by CPAP devices

Post by palerider » Tue Sep 13, 2016 12:03 pm

seekingwellness2016 wrote:I'm done some of my own reading now on this topic, and am satisfied with the conclusions I've been able to draw myself.
well, now we've established one fact: seekingwellness2016 is the type that lines to come in, ask questions and then argue about the answers given. also, likes to dig up old stuff (2009 Feb) for arguing purposes. you and avi will get along great.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
LSAT
Posts: 13412
Joined: Sun Nov 16, 2008 10:11 am
Location: SE Wisconsin

Re: AHI measured by CPAP devices

Post by LSAT » Tue Sep 13, 2016 12:04 pm

seekingwellness2016 wrote:I'm done some of my own reading now on this topic, and am satisfied with the conclusions I've been able to draw myself.

For those who are interested, there are in fact several different definitions of AHI.

See, for example, the following abstract/article:

http://www.ncbi.nlm.nih.gov/pubmed/19238801

where AHI(recommended), AHI(alternative), and AHI(Chicago) are all described.
The study you are posting is from 2009...maybe you can find something more current. Research continues......
I don't know what you are trying to prove...that you don't have Sleep Apnea?

User avatar
palerider
Posts: 32308
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: AHI measured by CPAP devices

Post by palerider » Tue Sep 13, 2016 12:10 pm

LSAT wrote:
seekingwellness2016 wrote:I'm done some of my own reading now on this topic, and am satisfied with the conclusions I've been able to draw myself.
The study you are posting is from 2009...maybe you can find something more current. Research continues......
I don't know what you are trying to prove...that you don't have Sleep Apnea?
that he's *obviously* smarter than the rest of us?
cuz we dun't know nuthin 'bout ahi or how machines guesstimate it.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.