PR System One - disable/turn off bluetooth possible?

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
User avatar
GoodNightRest
Posts: 84
Joined: Thu Feb 12, 2015 12:36 pm

PR System One - disable/turn off bluetooth possible?

Post by GoodNightRest » Sat Aug 20, 2016 1:11 pm

Hi,

I am researching for a spare CPAP (still waiting for my prescription), and I am interested to get the same one that I have now (System One 560P - no Bluetooth).
Unfortunately they don't carry it anymore. The new version has built-in Bluetooth (seems to be true for all 60 series line).
I checked the manual online, searched the forum, I didn't see a way to disable/turn off the built-in Bluetooth feature.

I understand Bluetooth is just a small-power transmitter, but I just don't like it to be on near me the whole night.
(I always set my cellphone to airplane mode when I'm sleeping).

Did anybody find a way to turn off/disable the Bluetooth in these 60 series machine? On setup (clinician) page perhaps?

If it's not possible, any other brand or model recommendation around the same price range with Auto Flex that doesn't have Bluetooth, or Bluetooth can be disabled?
(If I couldn't find anything maybe I'll get the new 560P and use either aluminum foil or Faraday cage bag to cover it).

Thanks!

Guest

Re: PR System One - disable/turn off bluetooth possible?

Post by Guest » Sat Aug 20, 2016 1:42 pm

Check Craigslist or this forum for the model you want.

User avatar
GoodNightRest
Posts: 84
Joined: Thu Feb 12, 2015 12:36 pm

Re: PR System One - disable/turn off bluetooth possible?

Post by GoodNightRest » Sat Aug 20, 2016 2:14 pm

Guest wrote:Check Craigslist or this forum for the model you want.
I did, as I mentioned in my original post:
I checked the manual online, searched the forum, I didn't see a way to disable/turn off the built-in Bluetooth feature.

User avatar
chunkyfrog
Posts: 34545
Joined: Mon Jul 12, 2010 5:10 pm
Location: Nowhere special--this year in particular.

Re: PR System One - disable/turn off bluetooth possible?

Post by chunkyfrog » Sat Aug 20, 2016 2:26 pm

Wireless sleep trackers use BT. Not cpap machines. (Unless you use a bluetooth sd card)
They use a cellular modem like LTE. Still no big deal, unless you are paranoid about somebody "hacking" your machine.
The world is full of things that warrant more dread. Take cancer, for example.

_________________
Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear
Additional Comments: Airsense 10 Autoset for Her

Guest

Re: PR System One - disable/turn off bluetooth possible?

Post by Guest » Sat Aug 20, 2016 3:56 pm

GoodNightRest wrote:
Guest wrote:Check Craigslist or this forum for the model you want.
I did, as I mentioned in my original post:
I checked the manual online, searched the forum, I didn't see a way to disable/turn off the built-in Bluetooth feature.
i think we are talking about too two 2 different things

to buy. check craigslist to buy the model cpap you want.

User avatar
chunkyfrog
Posts: 34545
Joined: Mon Jul 12, 2010 5:10 pm
Location: Nowhere special--this year in particular.

Re: PR System One - disable/turn off bluetooth possible?

Post by chunkyfrog » Sat Aug 20, 2016 3:58 pm

There is no bluetooth in a cpap!

_________________
Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear
Additional Comments: Airsense 10 Autoset for Her

Guest

Re: PR System One - disable/turn off bluetooth possible?

Post by Guest » Sat Aug 20, 2016 4:09 pm

chunkyfrog wrote:There is no bluetooth in a cpap!
GoodNightRest wrote:I am interested to get the same one that I have now (System One 560P
Guest wrote:Check Craigslist or this forum for the model you want.
GoodNightRest wrote:I did, as I mentioned in my original post:

I checked the manual online, searched the forum, I didn't see a way to disable/turn off the built-in Bluetooth feature.
Guest wrote:i think we are talking about too two 2 different things

to buy. check craigslist to buy the model cpap you want.
Maybe three tree 3 different things

but hey have fun

amenite
Posts: 507
Joined: Sat Jan 31, 2015 7:02 pm

Re: PR System One - disable/turn off bluetooth possible?

Post by amenite » Sat Aug 20, 2016 4:32 pm

chunkyfrog wrote:There is no bluetooth in a cpap!
I suppose you could call it an APAP, but our hosts describe it as a CPAP ... with bluetooth ...

https://www.cpap.com/productpage/pr-sys ... tooth.html

Product Features
The PR 60 Series Auto with Bluetooth is an innovative auto-adjusting CPAP machine that offers the same features as the popular non-bluetooth model but with the added benefit of viewing and tracking therapy data over a bluetooth connection.

User avatar
GoodNightRest
Posts: 84
Joined: Thu Feb 12, 2015 12:36 pm

Re: PR System One - disable/turn off bluetooth possible?

Post by GoodNightRest » Sat Aug 20, 2016 5:45 pm

amenite wrote:
chunkyfrog wrote:There is no bluetooth in a cpap!
I suppose you could call it an APAP, but our hosts describe it as a CPAP ... with bluetooth ...

https://www.cpap.com/productpage/pr-sys ... tooth.html

Product Features
The PR 60 Series Auto with Bluetooth is an innovative auto-adjusting CPAP machine that offers the same features as the popular non-bluetooth model but with the added benefit of viewing and tracking therapy data over a bluetooth connection.
Thanks amenite. Yes, that's the machine I am interested in and I plan to get it from our host once I got my prescription.

Additional note:
I am pretty sure a Bluetooth device is always transmitting even though it's not connected to other device (because we can discover that device - Bluetooth discovery).
With my computer, only if I disable the Bluetooth, then it can't be discovered, thus I believe stop transmitting.

I just don't like the idea of electronics device transmitting signal near my head for 8 hours when I am sleeping. As I mentioned in my original post, I even set my cellphone to airplane mode (cut off all transmission) when I sleep.

amenite
Posts: 507
Joined: Sat Jan 31, 2015 7:02 pm

Re: PR System One - disable/turn off bluetooth possible?

Post by amenite » Sat Aug 20, 2016 5:59 pm

I actually have one of these new in the box unopened from the blowout sale several weeks back. I was going to leave it that way but if you're in no hurry I'll pull out the manual tonight or tomorrow. You could always contact our host and ask them about it. I'm sure they have access to all the available info on it including the clinician and technical manuals.

User avatar
Goofproof
Posts: 16087
Joined: Mon Dec 05, 2005 3:16 pm
Location: Central Indiana, USA

Re: PR System One - disable/turn off bluetooth possible?

Post by Goofproof » Sat Aug 20, 2016 6:30 pm

As long as the machine is a few feet away, my foil hat will protect me. How do you think the T.V. and Radio get their signal? We live in a sea of electric signal waste. Unless you live in a Faraday cage, you will never be free from electric polution, that's why I got my foil hat.

P.S. Now they have a Tulsa Wireless car charger, you just park over it and it charges your car, no need to hook up the cable. Jim

If my neighbor gets one I will need a thicker hat.
Use data to optimize your xPAP treatment!

"The art of medicine consists in amusing the patient while nature cures the disease." Voltaire

User avatar
Snoozeless
Posts: 47
Joined: Wed Jan 20, 2016 4:19 pm
Location: Texas

Re: PR System One - disable/turn off bluetooth possible?

Post by Snoozeless » Sat Aug 20, 2016 9:56 pm

I have the same machine but I leave the Bluetooth on. Here's what you do:

Turn the knob to Setup. Push the knob down and press the ramp button. Hold until you hear two beeps. Then turn the knob to Bluetooth. It probably says "on." Turn the knob so that it says "off" and press the knob. That will save the setting.

This menu is where you can change all your settings such as pressure, change it from CPAP to APAP, etc.

egzbuen
Posts: 188
Joined: Thu May 05, 2016 10:06 am

Re: PR System One - disable/turn off bluetooth possible?

Post by egzbuen » Sat Aug 20, 2016 10:50 pm

GoodNightRest wrote:Hi,

I am researching for a spare CPAP (still waiting for my prescription), and I am interested to get the same one that I have now (System One 560P - no Bluetooth).
Unfortunately they don't carry it anymore. The new version has built-in Bluetooth (seems to be true for all 60 series line).
I checked the manual online, searched the forum, I didn't see a way to disable/turn off the built-in Bluetooth feature.

I understand Bluetooth is just a small-power transmitter, but I just don't like it to be on near me the whole night.
(I always set my cellphone to airplane mode when I'm sleeping).

Did anybody find a way to turn off/disable the Bluetooth in these 60 series machine? On setup (clinician) page perhaps?

If it's not possible, any other brand or model recommendation around the same price range with Auto Flex that doesn't have Bluetooth, or Bluetooth can be disabled?
(If I couldn't find anything maybe I'll get the new 560P and use either aluminum foil or Faraday cage bag to cover it).

Thanks!
I have the same model. You can turn off BT in the clinician menu. Haven't checked if it's in the regular menu though.

Press the knob and the ramp button simultaneously and wait for the beep.

Good luck.

Sent from heaven

User avatar
GoodNightRest
Posts: 84
Joined: Thu Feb 12, 2015 12:36 pm

Re: PR System One - disable/turn off bluetooth possible?

Post by GoodNightRest » Sun Aug 21, 2016 8:49 pm

amenite wrote:I actually have one of these new in the box unopened from the blowout sale several weeks back. I was going to leave it that way but if you're in no hurry I'll pull out the manual tonight or tomorrow. You could always contact our host and ask them about it. I'm sure they have access to all the available info on it including the clinician and technical manuals.
Hi amenite, thank you so much for the offer to open it up.
Thank you, I already got the answer from Snoozeless and egzbuen

User avatar
GoodNightRest
Posts: 84
Joined: Thu Feb 12, 2015 12:36 pm

Re: PR System One - disable/turn off bluetooth possible?

Post by GoodNightRest » Sun Aug 21, 2016 8:50 pm

Snoozeless wrote:I have the same machine but I leave the Bluetooth on. Here's what you do:

Turn the knob to Setup. Push the knob down and press the ramp button. Hold until you hear two beeps. Then turn the knob to Bluetooth. It probably says "on." Turn the knob so that it says "off" and press the knob. That will save the setting.

This menu is where you can change all your settings such as pressure, change it from CPAP to APAP, etc.
egzbuen wrote:
GoodNightRest wrote:Hi,

I am researching for a spare CPAP (still waiting for my prescription), and I am interested to get the same one that I have now (System One 560P - no Bluetooth).
Unfortunately they don't carry it anymore. The new version has built-in Bluetooth (seems to be true for all 60 series line).
I checked the manual online, searched the forum, I didn't see a way to disable/turn off the built-in Bluetooth feature.

I understand Bluetooth is just a small-power transmitter, but I just don't like it to be on near me the whole night.
(I always set my cellphone to airplane mode when I'm sleeping).

Did anybody find a way to turn off/disable the Bluetooth in these 60 series machine? On setup (clinician) page perhaps?

If it's not possible, any other brand or model recommendation around the same price range with Auto Flex that doesn't have Bluetooth, or Bluetooth can be disabled?
(If I couldn't find anything maybe I'll get the new 560P and use either aluminum foil or Faraday cage bag to cover it).

Thanks!
I have the same model. You can turn off BT in the clinician menu. Haven't checked if it's in the regular menu though.

Press the knob and the ramp button simultaneously and wait for the beep.

Good luck.

Sent from heaven
Thank you so much Snoozeless and egzbuen!