Need info for setup RS series M machine

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
User avatar
LAGrandpa
Posts: 66
Joined: Mon Feb 08, 2016 9:48 pm
Location: Louisiana

Need info for setup RS series M machine

Post by LAGrandpa » Mon Aug 15, 2016 4:57 pm

I have gotten displaced from my lifesaving CPAP machine because of the flooding here in south Louisiana and a good friend has loaned me a Respironics RemstarPro M series machine to get me by. Model number 1022334. Can someone please tell me how to get into the clinical menu.

_________________
Mask: ComfortGel Blue Nasal CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: SleepyHead Software Buckweat pillow also have Intellipap auto machine

User avatar
Pugsy
Posts: 65418
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Need info for setup RS series M machine

Post by Pugsy » Mon Aug 15, 2016 5:23 pm

Short version.
Respironics REMstar M Series (CPAP or BiPAP): Unplug the power cord from the back of the CPAP machine, then hold down both the left and right arrow keys while simultaneously plugging in the power supply. Once the device beeps twice, release the buttons. After you hear the two beeps, you'll see the word setup in the lower right corner of the LCD screen. When you select setup you'll be in the clinician's mode where you can adjust every feature of the CPAP machine, including the pressure setting.
If you run into something you are unsure of in the menu
http://www.apneaboard.com/manuals51/REM ... 0Guide.pdf
this is the quick setup guide download..it's for the Auto but I think you can figure things out...it has pictures.

If you want the full provider manual you can request it via email here
http://www.apneaboard.com/adjust-cpap-p ... tup-manual

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
Pugsy
Posts: 65418
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Need info for setup RS series M machine

Post by Pugsy » Mon Aug 15, 2016 5:25 pm

Oh...my bad manners...forgive me please.

My thoughts and prayers are with you. I have been watching the news and it's been scary.
Is there anything else you might need that we could help with? Just ask...

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
Wulfman...
Posts: 6695
Joined: Mon Sep 01, 2014 6:41 pm
Location: Nearest fishing spot

Re: Need info for setup RS series M machine

Post by Wulfman... » Mon Aug 15, 2016 6:11 pm

LAGrandpa wrote:I have gotten displaced from my lifesaving CPAP machine because of the flooding here in south Louisiana and a good friend has loaned me a Respironics RemstarPro M series machine to get me by. Model number 1022334. Can someone please tell me how to get into the clinical menu.
Sorry to hear of your misfortune.
Here are the text instructions for the M Series Pro which I've been posting on the forum for years.

Den

.


Menu for M Series Pro

1. Hold down the <- -> buttons while plugging in the power on the back, wait for 2 beeps, release buttons.
2. Press the + key. <- -> buttons move to next field, -/+ keys decrement/increment
3. Check the following field(s):

-Therapy Mode = CPAP

-CPAP Pressure = NN (4 - 20 cm.)

-C-Flex Setting = (options are off, 1, 2 or 3)

-Ramp Time = (options are 05 to 45 min)

-Ramp Start Pressure = cm (Min. 4cm)

-Mask Alert Feature = On (On/Off)

-Auto Off Feature = Off (On/off)

-Show AHI/Leak Feature = On (On/Off)

Press On/Off button to exit.





==========Resetting LCD Display Data====================

To reset LCD Therapy data (does NOT erase any data from SmartCard):

1. Hold down the (<-) (->) buttons while plugging in the power on the back,
wait for 2 beeps or 5 seconds, release buttons.
2. Press the (->) key and navigate to Therapy screen.
3. Press and HOLD down the (-) minus key for 5 seconds and/or until therapy data is reset to Zero.
4. Press the (->) key to scroll back to main menu, press (+) key to enter Setup or
On/Off button to exit.


.
(5) REMstar Autos w/C-Flex & (6) REMstar Pro 2 CPAPs w/C-Flex - Pressure Setting = 14 cm.
"Passover" Humidification - ResMed Ultra Mirage FF - Encore Pro w/Card Reader & MyEncore software - Chiroflow pillow
User since 05/14/05

User avatar
Sir NoddinOff
Posts: 4189
Joined: Mon May 14, 2012 5:30 pm
Location: California

Re: Need info for setup RS series M machine

Post by Sir NoddinOff » Tue Aug 16, 2016 3:41 pm

Today I heard that after a week the flood waters are finally going down. I wonder why the flood levels kept increasing for so long. I saw some pictures over the week and all I could think of was 'mildew'... heat, residual moisture and still air, the perfect recipe. Whew, I hate mildew and don't think I could cope with that.

_________________
Mask: AirFit™ F10 Full Face Mask with Headgear
Additional Comments: Sleepyhead software v.0.9.8.1 Open GL and Encore Pro v2.2.
I like my ResMed AirFit F10 FFM - reasonably low leaks for my ASV therapy. I'm currently using a PR S1 AutoSV 960P Advanced. I also keep a ResMed S9 Adapt as backup. I use a heated Hibernite hose. Still rockin' with Win 7 by using GWX to stop Win 10.

User avatar
LAGrandpa
Posts: 66
Joined: Mon Feb 08, 2016 9:48 pm
Location: Louisiana

Re: Need info for setup RS series M machine

Post by LAGrandpa » Thu Aug 18, 2016 11:40 am

Sorry for the late response but I just got internet access back today. Thanks for the responses guys. I got the info I needed and all is well. Thanks again.