What does mask setting actually change (on resmed machines)

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
Wonderbrah
Posts: 30
Joined: Wed Jun 22, 2016 4:41 am

What does mask setting actually change (on resmed machines)

Post by Wonderbrah » Sun Aug 14, 2016 2:58 am

In my Airsense 10 settings page, it asks what type of mask you're using: pillows, nasal, full face mask. I'm wondering what if anything this setting changes. Is it just for data purposes so the type of mask is listed for clinician's awareness maybe, or does it actually change something related to pressure output?

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: What does mask setting actually change (on resmed machines)

Post by palerider » Sun Aug 14, 2016 3:31 am

Wonderbrah wrote:In my Airsense 10 settings page, it asks what type of mask you're using: pillows, nasal, full face mask. I'm wondering what if anything this setting changes. Is it just for data purposes so the type of mask is listed for clinician's awareness maybe, or does it actually change something related to pressure output?
mask setting is not recorded in the data.

as to what it does, that depends on who you ask. the two main theories are
"it has something to do with leak calculations" (which to me, doesn't hold water, due to the variances of intentional leak among masks of the same type)
or,
it helps to compensate for differences in air flow resistance (which increases from ffm, to nasal to pillow masks) to allow the machine to more accurately deliver the intended pressure.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
Jay Aitchsee
Posts: 2936
Joined: Sun May 22, 2011 12:47 pm
Location: Southwest Florida

Re: What does mask setting actually change (on resmed machines)

Post by Jay Aitchsee » Sun Aug 14, 2016 5:03 am

Here it comes. You knew it would
viewtopic.php?f=1&t=105637#p1000814
Jay Aitchsee wrote:
palerider wrote:I've never seen that from resmed, (I'm sure there's a lot of stuff I haven't seen, I'd be interested in seeing any references.
Here's a couple:
ResScan Report Interpretation Guide, page 32 wrote: DEFINITIONS
Leak: This refers to unintentional leak,
which is the leak value after deduction
of the intentional mask leak
.emphasis added
ResScan Report Interpretation Guide, page 56 wrote: ResMed devices calculate unintentional leak:
Unintentional leak = Total leak – Intentional leak from mask CO2 washout
You have to set the mask type. emphasis added
http://www.resmed.eu/us/assets/html/niv/leak-management.html?nc=dealers wrote: Interface selection
ResMed offers an exceptional selection of vented masks designed to ensure adequate washout of carbon dioxide while minimizing leak and maintaining patient comfort. All of these benefits are extremely important for patients utilizing NIV therapy.

Many ResMed devices offer an interface selection menu. Selecting the suitable interface enables the device to factor in the intentional leak emitting from the mask vent, which allows for more accurate unintentional leak reporting and alarm accuracy
. emphasis added
Additionally, here's a paper you may find interesting:

http://scholar.google.com/scholar_url?u ... i=scholarr

Monitoring of Noninvasive Ventilation by
Built-in Software of Home Bilevel Ventilators
A Bench Study
In this study,we first realized that data provided by ventilator software were not measured in the same way in all devices; this is not always explicitly stated by the manufacturers. One device estimated leaks only under expiratory positive airway pressure (device E) (these results are thus misleading and cannot be compared with those of other devices); furthermore, some devices subtract from their estimation of leaks the intentional leaks expected for a given type of mask at a given pressure setting (devices F and G), while others report the sum of intentional and unintentional leaks. Once this was taken into account, the reliability of leak assessment was shown to be highly variable from one device to another. emphasis added
Devices F & G are ResMed which were found to be among the most accurate in leak reporting.

_________________
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Additional Comments: S9 Auto, P10 mask, P=7.0, EPR3, ResScan 5.3, SleepyHead V1.B2, Windows 10, ZEO, CMS50F, Infrared Video
Last edited by Jay Aitchsee on Sun Aug 14, 2016 8:07 am, edited 1 time in total.

User avatar
Pugsy
Posts: 65136
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: What does mask setting actually change (on resmed machines)

Post by Pugsy » Sun Aug 14, 2016 7:55 am

Jay Aitchsee wrote:Here it comes. You knew it would
Glad you dug it up and saved me the time...and I knew you would. Thank you for saving me some work.

I think that a generic intentional value per each setting is used as a starting point at whatever pressure and excess leak is calculated from that starting point. It's going to be close but it's never going to be 100% accurate because of the small variations in vent rates among the masks in the same class.
It's going to be close enough though to get well within the ball park for determining if excess leak is bad enough to cause a problem.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: What does mask setting actually change (on resmed machines)

Post by palerider » Sun Aug 14, 2016 5:56 pm

Jay Aitchsee wrote:Here it comes. You knew it would
well, actually no, I'd forgotten

thanks!

I admit that I still have problems with the concept, because of the published vent rates of the various masks, it just doesn't make any sense. :scratching head: add in the fact that the manuals all state "As a result of manufacturing variations, the vent flow rate may vary." which really makes it a "why bother" situation if it's just for making a good guess at excess leak rate.

I don't know, what I'd like to find is a patent that discusses it, something *technical*, not just an aside in a manual.

I note that the paper refers to the VPAP III ST and VPAP IV ST-A machines, which are both older tech, and, as a result, dumber tech. (earlier ASV machines had to have a sensing tube to measure mask pressure, because either the processor wasn't fast enough, or the software smart enough, to model the mask pressure out of the flow and pressure data in the unit itself... obviously this has been overcome with more advanced processing power in the S9 and later machines). I can see that with older, dumber machines, having to have these things set would make more sense in calculating leak, especially with fewer masks available, and probably different leak profiles among them.

I've seen two mask vent rate profile charts, from resmed, that were published at different times, and they show *different* leak amounts... implying that either one or both of the charts are wrong, or resmed has changed the mask vent rates..... well, of course I can't find one of them now, but I did find another one, and put them here..

some things of interest, looking at Resmed_pressure_vs_flow_rates.pdf, there's three different flow rates for nasal masks, and two for pillows. compared to 1016852_pressure-vs-flow_amer_eng.pdf however, the rates for the swift LT are different, but still nasal.

in mask-comp-list_amer_eng.pdf, it says that airfit N10 masks should use the pillow setting. interestingly enough, if you look at the *resistance* values in the manual, the n10 looks more like a pillow mask, than a nasal mask...
the following are some pillow masks (and the n10) the numbers are pressure drop at 50 and then 100lpm pressure

p10 .4 - 1.4
n10 .4 - 1.2
swiftfx .4 - 1.4

and some other nasal masks, pressure drop at 50 and 100lpm.
mirage activa .1 - .7
mirage fx : .2 - 1.0
mirage softgel .3 - .9

you see that generally, the pressure drop for a nasal mask (except the n10) is much less than the drop for a pillow mask.

anyway.. *shrug* I dunno.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
Jay Aitchsee
Posts: 2936
Joined: Sun May 22, 2011 12:47 pm
Location: Southwest Florida

Re: What does mask setting actually change (on resmed machines)

Post by Jay Aitchsee » Sun Aug 14, 2016 7:43 pm

palerider wrote:I admit that I still have problems with the concept, because of the published vent rates of the various masks, it just doesn't make any sense. :scratching head: add in the fact that the manuals all state "As a result of manufacturing variations, the vent flow rate may vary." which really makes it a "why bother" situation if it's just for making a good guess at excess leak rate.
I'm not saying you're wrong. I've just posted the only things I have found published on it. I suppose it's possible the setting could do both, used for establishing unintentional leak and for setting resistance values. And the function could have changed over the years as more masks were developed. I'm pretty sure the settings definition themselves have changed. The patent relied on for the current machines probably goes back to the S6 or the S7, or maybe even earlier. Without trying to dig out the info, I think the last major change, was the with the S9 and the FOT used to detect CA, I know the S8 had an interface selection setting, which is the vintage of most of the quotes above re: unintentional leak, and I "think" the S7 did too, but I don't want to look it up tonight. There are probably some old threads about it from circa 2006- 2005 that should make some interesting reading - if there was nothing better to do - probably a link to some patent material, too.

I think the VPAP series references some "advanced leak detection" or something like that, again without looking it up, so maybe that series does use a little different methodology.

I think it reasonable to say something like, "The ResMed interface setting is used by the machine to compensate for the differences in flow resistance of the various masks and in the calculation of unintentional leak". Someone will find this quote with Google, repost it and it will become so. And then we can just use it to answer the next time the question comes up. Whaddya think?

_________________
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Additional Comments: S9 Auto, P10 mask, P=7.0, EPR3, ResScan 5.3, SleepyHead V1.B2, Windows 10, ZEO, CMS50F, Infrared Video

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: What does mask setting actually change (on resmed machines)

Post by palerider » Sun Aug 14, 2016 8:47 pm

Jay Aitchsee wrote:I'm not saying you're wrong. I've just posted the only things I have found published on it. I suppose it's possible the setting could do both, used for establishing unintentional leak and for setting resistance values. And the function could have changed over the years as more masks were developed. I'm pretty sure the settings definition themselves have changed. The patent relied on for the current machines probably goes back to the S6 or the S7, or maybe even earlier. Without trying to dig out the info, I think the last major change, was the with the S9 and the FOT used to detect CA, I know the S8 had an interface selection setting, which is the vintage of most of the quotes above re: unintentional leak, and I "think" the S7 did too, but I don't want to look it up tonight. There are probably some old threads about it from circa 2006- 2005 that should make some interesting reading - if there was nothing better to do - probably a link to some patent material, too.

I think the VPAP series references some "advanced leak detection" or something like that, again without looking it up, so maybe that series does use a little different methodology.

I think it reasonable to say something like, "The ResMed interface setting is used by the machine to compensate for the differences in flow resistance of the various masks and in the calculation of unintentional leak". Someone will find this quote with Google, repost it and it will become so. And then we can just use it to answer the next time the question comes up. Whaddya think?
sounds like a plan.

also, note, I'm not saying that the 'used in leak calculations' is wrong, because I have no proof, but it doesn't make sense to me... at least, not with modern equipment, with older stuff, I could see it being needed for that.

speaking of the vpaps, on the treament screen, on the LCD, there's a 'leak' number... and you can see the leak amount, in real time. when you first start the machine, it registers each breath as a leak, until it calculates what the baseline should be, then it settles down to showing zero leak.
https://www.youtube.com/watch?v=1eeDtO-k47c
breathing, then after it settled down to zero, pulled one of the o2 port caps to create a leak... then plugged up up... back to zero leak, then removed the mask.

maybe the vpaps have to dynamically calculate baseline leak, because of the varying pressure?