Jay Aitchsee wrote:In case someone else needs it:You can just copy and paste the line, no need to type:remstarcpap wrote:Hi all,
If you have a ResMed machine and you take out your SD card to read it using Sleepyhead or ResScan, especially on a Windows 8.1 machine, you can see the dreaded "Invalid SD card erase data? Yes No" message when you put the card back in the machine.
Here's how to fix it without erasing the card.
First of all, what is happening? Windows writes a Folder called System Volume Information onto the card when you put it into your PC. Resmed doesn't like this, and refuses to accept the card.
Take the card out of your Cpap machine, and plug it back into your PC. A Windows Explorer window will open up showing the files on the card. If you don't see the System Volume Information folder it's probably there, but you have your Windows Folder Options set to hide system files.
I think this procedure will work regardless. Hold down the shift key, and right click inside the Explorer window showing your SD card. Click on the option Open Command Window Here.
Now type this command into the black command window:
rmdir "System Volume Information" /s /q
Type it exactly as shown, with the quotes, and with a space before and after the /s and /q commands.
This removes the folder. Pull out the SD card, and try it in your CPAP, now it should work fine.
The next time you remove your SD card from your CPAP, flip the little locking tab on the side, which should in most cases prevent Windows from writing any files to the card.
rmdir "System Volume Information" /s /q
The other thing you can do is to put your SD card into another machine and format it. Then unplug the ResMed AirSense machine. Put the blank card into the cpap while it's powerd off. after the card is in the cpap, apply power to the cpap. Now start up the cpap, breathe through your mask for about 2 - mins, and push the power off button on the top of the machind. the data in the ResMed's memory will repopulate the SD card.
I've done this numerous times so I know it works,
Since then, I've acquired a Toshiba Air Flash W-03 SDcard and non longer remove the card... saves me a ton of time and not having to worry about mucking up the card in the CPAP
The only thing is if you let sleepy head point at your sd card, it puts your profile there, so you'll have to recreate the user for sleepyhead
Best of luck with however you do it.
 
                 
                         
                         
                         
                         
                         
                         
                         
			 
	