Philips Respironics System One Dial Doesn't Work
-
slf
Philips Respironics System One Dial Doesn't Work
My husband needs to adjust the pressure on his Philips Respironics System One REMstar Plus C-Flex. I've seen the instructions for how to do this but run into a problem immediately--when I turn the knob nothing happens. It should scroll through the menu but it just stays on "Therapy." I tried holding down any combination of buttons while attaching the power to see if I could unlock it, but nothing happens. I'm not sure if it is locked or if something is broken. Any advice?
A little more background...we are living overseas and will return to the US in about 3 months. He tried a sleep specialist here but the doctor seemed more motivated to part us with our money than to actually do something helpful. We'll probably buy a new machine when we return home but in the meantime we're stuck and my husband is snoring again and waking up tired. We'd like to try bumping up the pressure just a little to see if that helps.
A little more background...we are living overseas and will return to the US in about 3 months. He tried a sleep specialist here but the doctor seemed more motivated to part us with our money than to actually do something helpful. We'll probably buy a new machine when we return home but in the meantime we're stuck and my husband is snoring again and waking up tired. We'd like to try bumping up the pressure just a little to see if that helps.
Re: Philips Respironics System One Dial Doesn't Work
Scroll over to set up, then hold both buttons down for several seconds. That should get you to the clinical menu.
_________________
| Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier |
| Mask: DreamWear Nasal CPAP Mask with Headgear |
What you need to know before you meet your DME http://tinyurl.com/2arffqx
Taming the Mirage Quattro http://tinyurl.com/2ft3lh8
Swift FX Fitting Guide http://tinyurl.com/22ur9ts
Don't Pay that Upcharge! http://tinyurl.com/2ck48rm
Taming the Mirage Quattro http://tinyurl.com/2ft3lh8
Swift FX Fitting Guide http://tinyurl.com/22ur9ts
Don't Pay that Upcharge! http://tinyurl.com/2ck48rm
-
slf
Re: Philips Respironics System One Dial Doesn't Work
That's the problem--I can't scroll. The dial/knob does nothing when I turn it.
- Sir NoddinOff
- Posts: 4189
- Joined: Mon May 14, 2012 5:30 pm
- Location: California
Re: Philips Respironics System One Dial Doesn't Work
BUMP
I'm pretty sure somebody last year or the year before had that same problem with their Respironic's machine and a CPAPtalk member gave them advice on how to fix or unstick the dial. I looked high and low and couldn't find that post. PM palerider and ask him (you'll have to register first)... he's good on that kind of stuff. If you hit a brick wall there's always
http://www.acbio.com
I'm pretty sure somebody last year or the year before had that same problem with their Respironic's machine and a CPAPtalk member gave them advice on how to fix or unstick the dial. I looked high and low and couldn't find that post. PM palerider and ask him (you'll have to register first)... he's good on that kind of stuff. If you hit a brick wall there's always
http://www.acbio.com
_________________
| Mask: AirFit™ F10 Full Face Mask with Headgear |
| Additional Comments: Sleepyhead software v.0.9.8.1 Open GL and Encore Pro v2.2. |
I like my ResMed AirFit F10 FFM - reasonably low leaks for my ASV therapy. I'm currently using a PR S1 AutoSV 960P Advanced. I also keep a ResMed S9 Adapt as backup. I use a heated Hibernite hose. Still rockin' with Win 7 by using GWX to stop Win 10.
Re: Philips Respironics System One Dial Doesn't Work
I don't remember that one. the dial is attached to a rotary encoder, an electromechanical device that translates rotary motion into electrical pulses saying which direction the knob is being turned. unless, somehow, the knob has stripped out on the shaft of the encoder and so is turning without turning the shaft of the encoder, there's not much I can think of to do, unless one is good with a soldering iron and sourcing parts.Sir NoddinOff wrote:BUMP
I'm pretty sure somebody last year or the year before had that same problem with their Respironic's machine and a CPAPtalk member gave them advice on how to fix or unstick the dial. I looked high and low and couldn't find that post. PM palerider and ask him (you'll have to register first)... he's good on that kind of stuff. If you hit a brick wall there's always
http://www.acbio.com
does the knob sort of 'click' when turning, ie, you can feel notches as you turn it?
Get OSCAR
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
- Sir NoddinOff
- Posts: 4189
- Joined: Mon May 14, 2012 5:30 pm
- Location: California
Re: Philips Respironics System One Dial Doesn't Work
Yes, Respironic's knobs/dials have notches in their sweep (ie. detents).palerider wrote: does the knob sort of 'click' when turning, ie, you can feel notches as you turn it?
_________________
| Mask: AirFit™ F10 Full Face Mask with Headgear |
| Additional Comments: Sleepyhead software v.0.9.8.1 Open GL and Encore Pro v2.2. |
Last edited by Sir NoddinOff on Sat Jul 16, 2016 3:55 pm, edited 1 time in total.
I like my ResMed AirFit F10 FFM - reasonably low leaks for my ASV therapy. I'm currently using a PR S1 AutoSV 960P Advanced. I also keep a ResMed S9 Adapt as backup. I use a heated Hibernite hose. Still rockin' with Win 7 by using GWX to stop Win 10.
Re: Philips Respironics System One Dial Doesn't Work
I meant, the broken oneSir NoddinOff wrote:Yes, Respironic's knobs/dials have stops in their sweep (ie. detents).palerider wrote: does the knob sort of 'click' when turning, ie, you can feel notches as you turn it?
Get OSCAR
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
- Sir NoddinOff
- Posts: 4189
- Joined: Mon May 14, 2012 5:30 pm
- Location: California
Re: Philips Respironics System One Dial Doesn't Work
Okay, I see what you meant, that's another way of asking if the knob is stripped . As I recall last time I popped a PR case, the knob itself had a 'D' shape aperture including a similar shaped metal spring that applied tension between the knob and the shaft. Those are hard to strip even if you try. Anyway, the OP disappeared, therefore they probably sent it back to the DME, screwed up pressing it down and HOLDING, or simply decided to stop rotating their knob from now on. Well, that sounded fairly masturbatory. Odd where these threads go sometimes.palerider wrote:I meant, the broken oneSir NoddinOff wrote:Yes, Respironic's knobs/dials have stops in their sweep (ie. detents).palerider wrote: does the knob sort of 'click' when turning, ie, you can feel notches as you turn it?
_________________
| Mask: AirFit™ F10 Full Face Mask with Headgear |
| Additional Comments: Sleepyhead software v.0.9.8.1 Open GL and Encore Pro v2.2. |
I like my ResMed AirFit F10 FFM - reasonably low leaks for my ASV therapy. I'm currently using a PR S1 AutoSV 960P Advanced. I also keep a ResMed S9 Adapt as backup. I use a heated Hibernite hose. Still rockin' with Win 7 by using GWX to stop Win 10.
-
slf
Re: Philips Respironics System One Dial Doesn't Work
Thanks everyone! Yes, the knob does sort of click when I turn it. I pulled it off to have a look and it does fit snugly on the "D" post so it sounds like I'm out of luck. No soldering iron or spare parts or any clue what to do with them if I did. I guess we'll have to wait until we're back in the USA.
- Sir NoddinOff
- Posts: 4189
- Joined: Mon May 14, 2012 5:30 pm
- Location: California
Re: Philips Respironics System One Dial Doesn't Work
Too bad it happened when you were out of the country. A sleep doctor won't be able to help you much in fixing a machine... try to find a equipment supplier or DME in whatever country you're in . How you would do that I can't say but maybe start a new post and ask for help finding a DME in whatever country your are in; in other words maybe somebody in that country can point you in the right direction (put that country right in the subject line to focus attention). I wouldn't stop using the machine just because your husband can't change the pressure - a slightly off pressure is better than nothing IMO. Again, here's the website for Accurate BioMed which is in Florida.slf wrote:Thanks everyone! Yes, the knob does sort of click when I turn it. I pulled it off to have a look and it does fit snugly on the "D" post so it sounds like I'm out of luck. No soldering iron or spare parts or any clue what to do with them if I did. I guess we'll have to wait until we're back in the USA.
http://www.acbio.com
I just had another thought since you are thinking of getting another machine when you return. Register and send a PM to Grayghost4... he frequently sends his machines overseas and has contacts in many places, plus he knows most of the regulations.
memberlist.php?mode=viewprofile&u=72737
_________________
| Mask: AirFit™ F10 Full Face Mask with Headgear |
| Additional Comments: Sleepyhead software v.0.9.8.1 Open GL and Encore Pro v2.2. |
I like my ResMed AirFit F10 FFM - reasonably low leaks for my ASV therapy. I'm currently using a PR S1 AutoSV 960P Advanced. I also keep a ResMed S9 Adapt as backup. I use a heated Hibernite hose. Still rockin' with Win 7 by using GWX to stop Win 10.