air seems hot

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
stockman

air seems hot

Post by stockman » Tue Jun 14, 2016 11:47 am

Hello I am new to the world of CPAP . I have been on the machine for 7 weeks. I have a Resmed 10. My setting is 3 for the humidity and 74 temp for the hose. Full face mask as I breath many times through my mouth. The problem I have is it seems too hot when I turn on the machine for about 5 minutes. I did shut off the ramp off. Those 5 to about 8 minutes kill me and make me sweat until the air pressure goes up. The pressures starts at 4 then goes to 9 to 12 on an average. Any idea would help . Thanks to all. :

User avatar
Pugsy
Posts: 65323
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: air seems hot

Post by Pugsy » Tue Jun 14, 2016 11:55 am

What exactly are your pressure settings and exactly which model machine are you using (there are 4 different AirSense 10 models)?
If you are using the AutoSet machine and the minimum pressure is 4 starting point...ramp wasn't doing anything anyway because it can't go lower than 4 anyway...shutting off the ramp in that case didn't really change anything in terms of air movement.

If you don't know how to find out exactly what your various settings are this manual will explain how if it an AirSense machine.

https://sleep.tnet.com/home/files/resme ... -guide.pdf

You can set the hose air temperature lower than 74 degrees but if the problem is air movement itself that may not offer much relief.
Usually that problem is resolved only by increasing the starting pressure.
Ambient bedroom air temp may also be a factor....what temperature is your bedroom?

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

stockman

Re: air seems hot

Post by stockman » Tue Jun 14, 2016 12:22 pm

I have the air sense auto set resmed 10

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: air seems hot

Post by palerider » Tue Jun 14, 2016 12:28 pm

stockman wrote:I have the air sense auto set resmed 10
and, what's it set to?

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

stockman

Re: air seems hot

Post by stockman » Tue Jun 14, 2016 12:45 pm

hose set is 74 temp pressure 4 to 15 . I keep the room temp at 69 temp

User avatar
Pugsy
Posts: 65323
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: air seems hot

Post by Pugsy » Tue Jun 14, 2016 12:52 pm

stockman wrote:pressure 4 to 15
I assume you mean the minimum is 4 and the maximum is 15.
Ramp is useless with the minimum of 4.
You might want to consider trying something like a minimum of 6 to get a little more air moving so it doesn't seem so stifling. You can also reduce the hose air temp ...I think it will go down to 64 or 66 though it can't work like air conditioning since it also depends on room air temp but it will be cooler than 74 for sure.

So you may need to do one or both...to relieve your unwanted results.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

stockman

Re: air seems hot

Post by stockman » Tue Jun 14, 2016 1:20 pm

I will try it tonight . A big Thank You

User avatar
Goofproof
Posts: 16082
Joined: Mon Dec 05, 2005 3:16 pm
Location: Central Indiana, USA

Re: air seems hot

Post by Goofproof » Tue Jun 14, 2016 1:45 pm

Air is hot! Room temp plus machine temp plus compressing a gas (air) makes it warmer, add in the humidity from your body and the Machines humidifier, you have the feeling of warmer.

That's why I run a room a/c in the bedroom year around, no HH at all, keeping room at 62 or colder year around. Jim

Bedroom door closed furnace duct closed in the winter....
Use data to optimize your xPAP treatment!

"The art of medicine consists in amusing the patient while nature cures the disease." Voltaire

stockman

Re: air seems hot

Post by stockman » Wed Jun 15, 2016 8:10 am

Thank You I brought up the pressure to 6 and there was a big difference. I slept well last night problem may be solved