Can't even take a nap
-
Thefrogshateme
- Posts: 16
- Joined: Wed Jun 08, 2016 4:52 pm
Can't even take a nap
I have a question about my sleep and wonder if anyone could help me. I got a machine in January.
Sleepmapper REMSTAR Auto A flex. Set at 17.
The reason I got a sleep study is because I kept busting out of my sleep the instant I go to sleep gasping for air. It just keeps on and on and on the moment I enter sleep. It's doing it now that I have a cpap machine also. I am drained and off from a hard days work just trying to take a nap with my wife. Trying to sleep 18 minutes with a ahi of 18. What could the problem be?
I have a SD card and put it in the computer but I don't know how to read it. I called Classic Sleep care LLC 8887072454 and they were no help at all. Just told me all they can read sent from the machine is the compliance and monthly ahi.
Sleepmapper REMSTAR Auto A flex. Set at 17.
The reason I got a sleep study is because I kept busting out of my sleep the instant I go to sleep gasping for air. It just keeps on and on and on the moment I enter sleep. It's doing it now that I have a cpap machine also. I am drained and off from a hard days work just trying to take a nap with my wife. Trying to sleep 18 minutes with a ahi of 18. What could the problem be?
I have a SD card and put it in the computer but I don't know how to read it. I called Classic Sleep care LLC 8887072454 and they were no help at all. Just told me all they can read sent from the machine is the compliance and monthly ahi.
Sleepmapper REMSTAR Auto A flex. Set at 17.
P1302925705CC
P1302925705CC
- ChicagoGranny
- Posts: 15419
- Joined: Sun Jan 29, 2012 1:43 pm
- Location: USA
Re: Can't even take a nap
Download and use Sleepyhead software - https://sleep.tnet.com/resources/sleepyheadThefrogshateme wrote:I have a SD card and put it in the computer but I don't know how to read it.
You can get plenty of help in this thread with using the software.
Look on the bottom of the machine and tell us the model number. They may have given you a discontinued machine. (Before you turn the machine over, make sure the humidifier tank is empty.)Thefrogshateme wrote: I got a machine in January.
Sleepmapper REMSTAR Auto A flex.
-
Thefrogshateme
- Posts: 16
- Joined: Wed Jun 08, 2016 4:52 pm
Re: Can't even take a nap
Okay I updated my model number in my signature. I have sleepyhead on the computer in the living room. Which area would I screen shot and post?
Sleepmapper REMSTAR Auto A flex. Set at 17.
P1302925705CC
P1302925705CC
- ChicagoGranny
- Posts: 15419
- Joined: Sun Jan 29, 2012 1:43 pm
- Location: USA
Re: Can't even take a nap
There are some examples on this page - viewtopic.php?f=1&t=111910&view=unread#unreadThefrogshateme wrote:Which area would I screen shot and post?
I could be wrong, but that looks more like a serial number. Check for the model number.P1302925705CC
Re: Can't even take a nap
That is a serial number...model numbers are usually 3 digit numbers with DS in front or behind the 3 digits
except for the new DreamStation machines which I think have 4 numbers but if it says Remstar on it then it isn't the new DreamStaton. Most likely either a 50 or 60 series PR System One Auto CPAP.
Remstar with AFlex and a SD card...it's going to be either 550/551 or 560/561.
except for the new DreamStation machines which I think have 4 numbers but if it says Remstar on it then it isn't the new DreamStaton. Most likely either a 50 or 60 series PR System One Auto CPAP.
Remstar with AFlex and a SD card...it's going to be either 550/551 or 560/561.
_________________
| Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
| Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/ |
I may have to RISE but I refuse to SHINE.
-
Thefrogshateme
- Posts: 16
- Joined: Wed Jun 08, 2016 4:52 pm
Re: Can't even take a nap
It just said this as I logged in now I'm scared I'm gonna die "The last time you used your Philips Respironics RemStar Auto with A-Flex...
was 0 days ago (on Wednesday, June 08, 2016)
You had an AHI of 17.18, which is considered horrible, please consult your doctor
You only had the mask on for 0 hours, 17 minutes and 28 seconds.
You had no major mask leaks (maximum was 6.00 L/min).
Your CPAP machine blasted you with a constant 15cmH2O of air"
was 0 days ago (on Wednesday, June 08, 2016)
You had an AHI of 17.18, which is considered horrible, please consult your doctor
You only had the mask on for 0 hours, 17 minutes and 28 seconds.
You had no major mask leaks (maximum was 6.00 L/min).
Your CPAP machine blasted you with a constant 15cmH2O of air"
Sleepmapper REMSTAR Auto A flex. Set at 17.
P1302925705CC
P1302925705CC
-
Thefrogshateme
- Posts: 16
- Joined: Wed Jun 08, 2016 4:52 pm
- ChicagoGranny
- Posts: 15419
- Joined: Sun Jan 29, 2012 1:43 pm
- Location: USA
Re: Can't even take a nap
Post the previous night's chart.Thefrogshateme wrote:You only had the mask on for 0 hours, 17 minutes and 28 seconds.
Re: Can't even take a nap
put the REF number in your signature, i'll have something like 560 in there.Thefrogshateme wrote:Okay I updated my model number in my signature. I have sleepyhead on the computer in the living room. Which area would I screen shot and post?
Get OSCAR
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
- zoocrewphoto
- Posts: 3732
- Joined: Mon Apr 30, 2012 10:34 pm
- Location: Seatac, WA
Re: Can't even take a nap
Welcome to the forum. You have come to a good place for help.
It will help if we can see the flow limitation and snore graphs. My first impression is that you appear to be using ramp, and starting from 4. As your machine ramps up, it does not respond to flow limitations, so it can only prevent an event if the pressure happens to be high enough. And it isn't. During your 17 minutes, the ramp appears to make it to 11.
I would start by having the ramp start higher, and if you are comfortable with it, shorten it. You might want to experiment with it while awake and see what pressures you are comfortable with. Some people struggle with 7 or 8 while some of us start at 10 or even 18 without the ramp. I start at 11 without ramp, and that prevents most of my events, so I get treatment right away. I would not be comfortable starting at 15 even though my pressure spikes to 17 sometimes without waking me.
If you have trouble with start pressures more than 7 or, start with what you can, and over time, raise the minimum. The earlier your machine can get to a useful pressure, the earlier you can get treatment and not be jerked out of sleep. You may need to practice during the day while reading or watching tv.
It will help if we can see the flow limitation and snore graphs. My first impression is that you appear to be using ramp, and starting from 4. As your machine ramps up, it does not respond to flow limitations, so it can only prevent an event if the pressure happens to be high enough. And it isn't. During your 17 minutes, the ramp appears to make it to 11.
I would start by having the ramp start higher, and if you are comfortable with it, shorten it. You might want to experiment with it while awake and see what pressures you are comfortable with. Some people struggle with 7 or 8 while some of us start at 10 or even 18 without the ramp. I start at 11 without ramp, and that prevents most of my events, so I get treatment right away. I would not be comfortable starting at 15 even though my pressure spikes to 17 sometimes without waking me.
If you have trouble with start pressures more than 7 or, start with what you can, and over time, raise the minimum. The earlier your machine can get to a useful pressure, the earlier you can get treatment and not be jerked out of sleep. You may need to practice during the day while reading or watching tv.
_________________
| Mask: Quattro™ FX Full Face CPAP Mask with Headgear |
| Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control |
| Additional Comments: Resmed S9 autoset pressure range 11-17 |
Who would have thought it would be this challenging to sleep and breathe at the same time?
Re: Can't even take a nap
Short naps don't count.
Pick a night with several hours of real sleep and share the detailed report for that night.
Pick a night with several hours of real sleep and share the detailed report for that night.
_________________
| Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
| Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/ |
I may have to RISE but I refuse to SHINE.
- zoocrewphoto
- Posts: 3732
- Joined: Mon Apr 30, 2012 10:34 pm
- Location: Seatac, WA
Re: Can't even take a nap
I got the impression that he can't sleep with the machine as he keeps waking up gasping. I don't think he has any full nights of data. He's probably not making it past the ramp time.
_________________
| Mask: Quattro™ FX Full Face CPAP Mask with Headgear |
| Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control |
| Additional Comments: Resmed S9 autoset pressure range 11-17 |
Who would have thought it would be this challenging to sleep and breathe at the same time?
Re: Can't even take a nap
It would still be better to see a more prolonged attempt at sleep. Impossible to tell anything with 17 minutes and nap at that which is notorious for flagging all sorts of crap that isn't real.zoocrewphoto wrote:I got the impression that he can't sleep with the machine as he keeps waking up gasping. I don't think he has any full nights of data. He's probably not making it past the ramp time.
So to OP....do you have anything where you actually may have went to sleep?
_________________
| Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
| Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/ |
I may have to RISE but I refuse to SHINE.
Re: Can't even take a nap
I agree, the first thing that I'd do is to either get rid of the ramp, or move up the starting ramp pressure. In my opinion (and experience) you're starting your ramp at too low of a pressure. When I started, I first moved my ramp up to start at 6cm, 4cm left me feeling like I wasn't getting enough air, then turned it off entirely. I had my range set to 8-10cm at that point, so turning off the ramp wasn't really difficult *smile*. In fact, it was nice to get that extra air, after being starved for it at 4cm.
So that's something to try, move the ramp starting pressure higher. And maybe shorten the ramp time as well after that, if you're comfortable with doing that. That moves your pressure higher faster. But you don't have to go any faster than you're comfortable with. My main impression is that you're starting out with a lot less air than you need, and your cpap is just the thing to give you more air to get past that initial suffocation feeling that wakes you right up *smile*.
Once you get past the initial "lack of air" problem, and you can keep your mask on for a while, you can look at your data and see what's up over a long enough period of sleeping to figure out what comes next *smile*.
So that's something to try, move the ramp starting pressure higher. And maybe shorten the ramp time as well after that, if you're comfortable with doing that. That moves your pressure higher faster. But you don't have to go any faster than you're comfortable with. My main impression is that you're starting out with a lot less air than you need, and your cpap is just the thing to give you more air to get past that initial suffocation feeling that wakes you right up *smile*.
Once you get past the initial "lack of air" problem, and you can keep your mask on for a while, you can look at your data and see what's up over a long enough period of sleeping to figure out what comes next *smile*.
_________________
| Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier |
| Mask: ResMed AirFit N30 Nasal CPAP Mask with Headgear |
- ChicagoGranny
- Posts: 15419
- Joined: Sun Jan 29, 2012 1:43 pm
- Location: USA
Re: Can't even take a nap
According to your chart, it is set at 15. And yes, the others are correct about the ramp being set waaay too low. It's no wonder:Thefrogshateme wrote:REMSTAR Auto A flex. Set at 17.
You just aren't getting enough.Thefrogshateme wrote:gasping for air
Since your prescription is 15, I would set the ramp beginning pressure to 10. Let's get you some air!