Leak Rate Acceptable?

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
VolDog
Posts: 7
Joined: Sat May 28, 2016 10:07 am

Leak Rate Acceptable?

Post by VolDog » Sun Jun 05, 2016 6:16 am

Can someone please tell me if the following leak rate is acceptable:

AHIApnea Hypopnea Index 0.48
ResMed AirCurve 10
ASVModel 37043 - 22151825854
PAP Mode: ASVAuto
Min EPAP 6.0 Max IPAP 17.0 PS 0.0-5.0 (cmH2O)
Date Sleep Wake Hours
6/4/16 22:01:02 06:20:03 08:19:01
Large Leak 3.59%
Clear Airway 0.00
Unclassified Apnea 0.00
Obstructive 0.00
Hypopnea 0.48

Event Breakdown

Statistics
Channel Min Med 95% Max
EPAPExpiratory Pressure (cmH2O)
W-Avg: 6.27 4.08 6.22 6.84 7.66
IPAPInspiratory Pressure (cmH2O)
W-Avg: 8.04 4.08 7.66 11.28 12.62
Minute Vent.Amount of air displaced per minute (L/min)
W-Avg: 9.05 0.00 8.62 13.38 29.75
Resp. RateRate of breaths per minute (Breaths/min)
W-Avg: 16.08 0.00 15.80 20.00 29.80
Flow Limit.Graph showing severity of flow limitations (Severity (0-1))
W-Avg: 0.00 0.00 0.00 0.00 1.00
Leak RateRate of detected mask leakage (L/min)
W-Avg: 5.94 0.00 3.60 21.60 43.20
SnoreGraph displaying snore volume (?)
W-Avg: 0.00 0.00 0.00 0.00 0.14
Insp. TimeTime taken to breathe in (Seconds)
W-Avg: 1.68 0.22 1.56 2.46 5.06
Exp. TimeTime taken to breathe out (Seconds)
W-Avg: 2.09 0.32 2.08 2.68 5.92
Target Vent.Target Minute Ventilation? (L/min)
W-Avg: 8.16 4.25 8.25 10.12 11.88
Tidal VolumeAmount of air displaced per breath (ml)
W-Avg: 563.39 0.00 540.00 800.00 1840.00

Total time in apnea 00:00:40
Time over leak redline 3.594%
Machine Settings

ModeCPAP Mode ASVAuto
Min EPAPLower Expiratory Pressure 6.00 cmH2O
Max EPAPHigher Expiratory Pressure 12.00 cmH2O
Min IPAPLower Inspiratory Pressure 6.00 cmH2O
Max IPAPHigher Inspiratory Pressure 17.00 cmH2O
PS MinPressure Support Minimum 0.00 cmH2O
PS MaxPressure Support Maximum 5.00 cmH2O
Antibacterial FilterAntibacterial Filter No
Climate ControlClimate Control Manual
Humidifier StatusHumidifier Enabled Status On
Humidity LevelHumidity Level 3
MaskResMed Mask Setting Nasal
Patient AccessPatient Access Yes
RampRamp Enable Off
Smart StartMachine auto starts by breathing Off
TemperatureClimateLine Temperature 27 ºC
Temperature EnableClimateLine Temperature Enable 2
Session Information
CPAP Sessions
On Date Start End Duration
ResMed Session #1465092062
Click to disable this session. 6/4/16 22:01:02 06:20:03 08h 19m 01s

User avatar
Pugsy
Posts: 65257
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Leak Rate Acceptable?

Post by Pugsy » Sun Jun 05, 2016 6:33 am

VolDog wrote:Large Leak 3.59%
Means for about 3 1/2 % of the night you were in large leak territory....
Out of roughly 8 1/4 hours of sleep that amounts to approx 17 minutes.
Not the end of the world by any means and acceptable in my book unless the leaks wake me up and then that's a different thing.
Leak RateRate of detected mask leakage (L/min)
W-Avg: 5.94 0.00 3.60 21.60 43.20
The 21.20 L/min is a 95% leak number...all that means is that you were at OR BELOW that number of 95% of the night...5% you were above but the max you hit was 43.20. I disregard maximums because they can look really bad just from a quick reseating of the mask. It does NOT mean that you were at 21.20 L/min leak for 95% of the night...remember the "or below" part of the definition.

ResMed's line in the sand is 24 L/min and for 95% of the night you were below that number and probably a lot below it. 95% numbers are easily skewed with just a little higher numbers being added into the mix.
ResMed doesn't even give a Mr Frowny face for leak until you spend 30% or more of the night in large leak territory...which I don't quite agree with.

I would say that your leak numbers are well within acceptable limits providing those leaks didn't wake you up often during the night.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

VolDog
Posts: 7
Joined: Sat May 28, 2016 10:07 am

Re: Leak Rate Acceptable?

Post by VolDog » Sun Jun 05, 2016 6:41 am

Thanks for your response! It's very much appreciated.

I'm happy to be getting sleep without constantly waking up as before. I think I can thank my doctor for prescribing Prilosec as the Gerd apparently was waking me up every 1.5 hours prior to taking it.

Last question, I noticed that Sleepyhead said I went to bed at 10:01 when it was actually 11:01. How do I fix that?

User avatar
Pugsy
Posts: 65257
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Leak Rate Acceptable?

Post by Pugsy » Sun Jun 05, 2016 6:49 am

VolDog wrote:Last question, I noticed that Sleepyhead said I went to bed at 10:01 when it was actually 11:01. How do I fix that?
There is a clock on the machine that you can set...just don't try to set it until after 12 Noon on that clock...or within an hour of using it. I usually tell people to just set it right at bedtime.

For some reason using the DST box option in SH doesn't/hasn't worked for me.

The clinical manual should explain how to get to the clock.

I don't bother with changing as DST comes and goes. It's enough to know I got X amount of hours of sleep.
Now if I was trying to sync SH report with pulse ox report...then I would make the effort to keep times exact.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.