1st night sleepyhead results, pg2

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
zlman
Posts: 18
Joined: Tue May 24, 2016 7:27 pm

1st night sleepyhead results, pg2

Post by zlman » Tue May 24, 2016 7:41 pm

I'm buying this on fri, RESMED S9 Auto-Set respiratory system w/ heated humidity reservoir.

I'm fairly certain I have apnea, sleep is never restful. Wives have told me I stop breathing at night, etc.

I found the above unit with 230 hrs for $250. I need direction on initial set-up and suggestions for use. I've looked through the stickied items at the top of the page but there's so much info it seems impossible to get through it all, that and I'm not sure what I should be reading and whats too advanced for a beginner.

This will be a journey and I'll be back often with ?'s, for now please point me to links or pages that I should start with.

Thank you for any and all help.
Last edited by zlman on Sat May 28, 2016 8:57 am, edited 1 time in total.

User avatar
LSAT
Posts: 13412
Joined: Sun Nov 16, 2008 10:11 am
Location: SE Wisconsin

Re: Buying, set-up. No medical diagnosis. On my own.

Post by LSAT » Tue May 24, 2016 7:45 pm

zlman wrote:I'm buying this on fri, RESMED S9 Auto-Set respiratory system w/ heated humidity reservoir.

I'm fairly certain I have apnea, sleep is never restful. Wives have told me I stop breathing at night, etc.

I found the above unit with 230 hrs for $250. I need direction on initial set-up and suggestions for use. I've looked through the stickied items at the top of the page but there's so much info it seems impossible to get through it all, that and I'm not sure what I should be reading and whats too advanced for a beginner.

This will be a journey and I'll be back often with ?'s, for now please point me to links or pages that I should start with.

Thank you for any and all help.
It's going to be a shot in the dark...here's the Clinical Manual.....Start with the settings Min 6...Max 20
http://osahelp.com/clinician-manuals/S9 ... al-eng.pdf

zlman
Posts: 18
Joined: Tue May 24, 2016 7:27 pm

Re: Buying, set-up. No medical diagnosis. On my own.

Post by zlman » Tue May 24, 2016 7:50 pm

LSAT wrote:
zlman wrote:I'm buying this on fri, RESMED S9 Auto-Set respiratory system w/ heated humidity reservoir.

I'm fairly certain I have apnea, sleep is never restful. Wives have told me I stop breathing at night, etc.

I found the above unit with 230 hrs for $250. I need direction on initial set-up and suggestions for use. I've looked through the stickied items at the top of the page but there's so much info it seems impossible to get through it all, that and I'm not sure what I should be reading and whats too advanced for a beginner.

This will be a journey and I'll be back often with ?'s, for now please point me to links or pages that I should start with.

Thank you for any and all help.
It's going to be a shot in the dark...here's the Clinical Manual.....Start with the settings Min 6...Max 20
http://osahelp.com/clinician-manuals/S9 ... al-eng.pdf
Thats a start. Thanks!

User avatar
Julie
Posts: 20051
Joined: Tue Feb 28, 2006 12:58 pm

Re: Buying, set-up. No medical diagnosis. On my own.

Post by Julie » Tue May 24, 2016 8:30 pm

Go to the lightbulb top of the pg (Cpap Wiki) and start reading... and reading... and reading.

User avatar
palerider
Posts: 32308
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Buying, set-up. No medical diagnosis. On my own.

Post by palerider » Tue May 24, 2016 8:31 pm

zlman wrote:I'm buying this on fri, RESMED S9 Auto-Set respiratory system w/ heated humidity reservoir.

I'm fairly certain I have apnea, sleep is never restful. Wives have told me I stop breathing at night, etc.

I found the above unit with 230 hrs for $250. I need direction on initial set-up and suggestions for use. I've looked through the stickied items at the top of the page but there's so much info it seems impossible to get through it all, that and I'm not sure what I should be reading and whats too advanced for a beginner.

This will be a journey and I'll be back often with ?'s, for now please point me to links or pages that I should start with.

Thank you for any and all help.
you have downloaded sleepyhead, right? (nod your head and say "yes")

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

thecpapguy!
Posts: 75
Joined: Sun Dec 06, 2015 10:26 pm

Re: Buying, set-up. No medical diagnosis. On my own.

Post by thecpapguy! » Tue May 24, 2016 11:29 pm

I definitely would get very active, like you already are, in the forums and online research if you are going to initiate the therapy on your own choice. So that is an excellent first step! I listed out the steps I would take in finding some information, getting your right pressure settings, getting the right mask, and ultimately monitoring to make sure positive pressure therapy stays working.

1. You said you are getting an S9, make sure it is an "autoset". Auto CPAP will likely be better for you in this scenario. I think most users would agree.

2. Find your right settings to start. You can use "Google" or "Youtube" to find how to unlock the Resmed S9 clinical menu to adjust your pressures.

3. Starting at a pressure of of "6 cmH2O" to "20 cmH2O' could help for five to ten days.

4. DOWNLOAD SLEEPY HEAD OR THE MANUFACTURER DOWNLOADING SOFTWARE! Get this software so you can download, do a review, or even upload it to the forum and seek advice on changing your pressures moving forward. You may start at the pressure settings of 6-20, but it will likely be more comfortable, and beneficial, if you change these pressure to a more narrow setting after five to ten days. For example: after reviewing your download you may consistently maintain a pressure of "12 cmH2O" for the majority of the night. So change your pressures from 6-20 to 10-14.

5. Finding the right mask is always the trick when private paying. Evaluate your needs when it comes to a mask and do not assume that a nasal style or a full face style will exclusively work for you! You may feel that a full face mask is all that will work for you, but a nasal but may very tolerable. Although you are private paying for the machine, you may consider using a supplier for your mask fitting. It isn't that these providers can better serve you with a mask fitting technician (most DME providers are highly incompetent), but they may have more of a selection and exchanging options for you.

For the record, the pressures I mentioned above are entirely hypothetical. They are not meant to be the exact pressure that you always use. They are simply examples. I hope this information helps! Some information on my blog is below. Maybe it will help clarify a little of the confusion for you and help with what mask interfaces are out there for you.

http://cpapguy.blogspot.com/2015/11/the ... -apap.html
http://cpapguy.blogspot.com/2016/01/ful ... tions.html
http://cpapguy.blogspot.com/2016/01/nas ... tions.html
http://cpapguy.blogspot.com/2016/02/nas ... tions.html

User avatar
sleepylynn
Posts: 70
Joined: Sat Oct 31, 2015 5:20 pm

Re: Buying, set-up. No medical diagnosis. On my own.

Post by sleepylynn » Wed May 25, 2016 1:15 pm

There's a ton of advice and instruction on this forum. I did everything on my own, too, and I would never have been able to figure out how to get the right machine settings without this place. The 'Pugsy's Pointers' thread is a fantastic resource. Best of luck!

_________________
Machine: PR System One REMStar 60 Series Auto CPAP Machine
Mask: DreamWear Nasal CPAP Mask with Headgear
Additional Comments: SleepyHead Software

zlman
Posts: 18
Joined: Tue May 24, 2016 7:27 pm

Re: Buying, set-up. No medical diagnosis. On my own.

Post by zlman » Wed May 25, 2016 6:50 pm

Very excited for fri to arrive, have most of the long weekend to set-up and experiment with the S9.

klv329
Posts: 500
Joined: Mon Aug 31, 2015 6:43 pm
Location: Bradenton, Florida

Re: Buying, set-up. No medical diagnosis. On my own.

Post by klv329 » Thu May 26, 2016 8:45 am

What type of mask are you going to start off with?

_________________
Machine: ResMed AirCurve 10 ASV Machine with Heated Humidifier
Additional Comments: Resmed Aircurve 10 ASVAUTO Min Epap 10.4, Max Epap 11.6, PS 1.6-12.0

User avatar
avi123
Posts: 4509
Joined: Tue Dec 21, 2010 5:39 pm
Location: NC

Re: Buying, set-up. No medical diagnosis. On my own.

Post by avi123 » Thu May 26, 2016 9:36 am

Zalman, I prefer to use ResScan software over Sleepyhead.

Image

Image

_________________
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments:  S9 Autoset machine; Ruby chinstrap under the mask straps; ResScan 5.6
see my recent set-up and Statistics:
http://i.imgur.com/TewT8G9.png
see my recent ResScan treatment results:
http://i.imgur.com/3oia0EY.png
http://i.imgur.com/QEjvlVY.png

User avatar
lordvader
Posts: 156
Joined: Sat Mar 20, 2010 10:37 am
Location: WA

Re: Buying, set-up. No medical diagnosis. On my own.

Post by lordvader » Thu May 26, 2016 10:09 am

avi123 wrote:Zalman, I prefer to use ResScan software over Sleepyhead.

Image

Image
Where can I find the ResScan software?
Machine: ResMed AirSense 11 Autoset
Mask: Fisher & Paykel Zest Q

User avatar
Pugsy
Posts: 65411
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Buying, set-up. No medical diagnosis. On my own.

Post by Pugsy » Thu May 26, 2016 12:16 pm

lordvader wrote:Where can I find the ResScan software?
You will need to go over to the apneaboard forum and register and request that a link be emailed to you.

http://www.apneaboard.com/adjust-cpap-p ... tup-manual

ResScan shows the same data that SleepyHead shows but just presents it visually in a different manner.
It's huge and it is often really slow.
Also...if you use a Mac it won't work at all so don't bother....ResScan is Windows only software and to use it on a Mac you have to put virtual Windows on the Mac....which can be done but involves some work.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
Pesser
Posts: 547
Joined: Sat Jan 11, 2014 4:41 pm
Location: Nova Scotia Canada

Re: Buying, set-up. No medical diagnosis. On my own.

Post by Pesser » Thu May 26, 2016 12:26 pm

palerider wrote:
zlman wrote:I'm buying this on fri, RESMED S9 Auto-Set respiratory system w/ heated humidity reservoir.

I'm fairly certain I have apnea, sleep is never restful. Wives have told me I stop breathing at night, etc.

I found the above unit with 230 hrs for $250. I need direction on initial set-up and suggestions for use. I've looked through the stickied items at the top of the page but there's so much info it seems impossible to get through it all, that and I'm not sure what I should be reading and whats too advanced for a beginner.

This will be a journey and I'll be back often with ?'s, for now please point me to links or pages that I should start with.

Thank you for any and all help.
you have downloaded sleepyhead, right? (nod your head and say "yes")
Nice to see you with your old avatar. Your eyes are closer together than before!

_________________
MachineMaskHumidifier
Additional Comments: Fear of devil with halo

User avatar
palerider
Posts: 32308
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Buying, set-up. No medical diagnosis. On my own.

Post by palerider » Thu May 26, 2016 7:20 pm

Pesser wrote:Nice to see you with your old avatar. Your eyes are closer together than before!
really? *grabs mirror, adjusts halo* I'm sure it's just a trick of the light

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
palerider
Posts: 32308
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Buying, set-up. No medical diagnosis. On my own.

Post by palerider » Thu May 26, 2016 7:20 pm

lordvader wrote:
avi123 wrote:Zalman, I prefer to use ResScan software over Sleepyhead.

Image

Image
Where can I find the ResScan software?
resscan is much more difficult to use, and provides less information, but, to each his own.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.