doing fine until I awake with the machine roaring

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
bluetrout
Posts: 92
Joined: Mon Apr 11, 2016 10:10 pm

doing fine until I awake with the machine roaring

Post by bluetrout » Thu May 05, 2016 2:24 pm

Image

Everything goes great until I wake up, usually with the machine at a high pressure forcing a lot of air which leaks. Can't get back to sleep with that going on so I stop the machine and get to sleep without it for another few hours. I'd like to get the full benefit from the CPAP but not sure how.

User avatar
Pugsy
Posts: 65328
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: doing fine until I awake with the machine roaring

Post by Pugsy » Thu May 05, 2016 2:34 pm

Exactly what pressure is the machine blowing when it is "roaring".
Per your report the highest it ever went was 10.5.
If that is too much you can try lowering the maximum and keep a small range or go to cpap mode and maybe get by with something like 9 maybe....

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

bluetrout
Posts: 92
Joined: Mon Apr 11, 2016 10:10 pm

Re: doing fine until I awake with the machine roaring

Post by bluetrout » Thu May 05, 2016 3:01 pm

Hmmm. You're right Pugsy. I was looking at the AHI going up at about 1:45 however the pressure wasn't high. The leak peaked around then too. Wonder if that's what woke me up. The feeling was that the air was pouring into the mask and that it was leaking everywhere. IT seemed loud and so the "roaring".

User avatar
Pugsy
Posts: 65328
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: doing fine until I awake with the machine roaring

Post by Pugsy » Thu May 05, 2016 3:42 pm

Now the machine when it is leaking can sound like it is "roaring" and go silent when the seal is fixed and the pressure the machine is spitting out is really the same. It really doesn't go into gale force winds in an effort to reseal things. Instead it will actually drop the pressure first.

I am betting the leak woke you up and that's what you experienced and what seemed like huge winds which is really quite normal to perceive things that way.
That said.... some people do find that the changes in pressure when using apap mode are a disruptive factor and other can sleep through some pretty dramatic real changes. When I was using apap I was using 10 minimum and 20 max and I often saw it hit 18 cm...but it was never noticed while I was sleeping. It was only noted once I saw the report. Even now I often go above 20 cm for short periods of time and I never know it.
Others might find that a small 1 or 2 cm pressure change wakes them up.

The AHI did go up but in the face of the large leak occurring at around the same time as the increase in AHI .....I would be first inclined to think the big leak allowed the apnea events to sneak past the defenses. In the face of a big leak these machines often don't know that they need to increase the pressure because of the apneas because the sensors get compromised. They are doing good to flag the sensors...much less respond. And if you look at the pressure line during that large leak it didn't do much.

If this happens often then I would probably investigate what I could do to reduce the happenings but if it is a rare occurrence I don't know that I would do much. You are so new to therapy and there's just so much to try to adjust to and worrying overly much about such a short time in large leak can create more stress than is really warranted. I know it is annoying and it wakes you up and that of course is unwanted but we all have this happen from time to time...even me. It happened a lot more when I was new to therapy and more sensitive to the least little thing. I sleep through those short periods of large leak now. On the rare occasion it wakes me up now I just have a short brain wave that tells me to "fix the mask" and it takes me a millisecond and I roll over and go back to sleep. I will admit that it bothered my sleep a lot more when I was new to therapy.

I hesitate to suggest that you start doing extraordinary things to prevent this occurrence because I hate to pile stuff on a newbie that can cause stress unless I really think it is necessary.

Next time you put your mask on an turn on the machine and get the seal...take a brief second to purposely break the seal and note the loud wind noise...then put the mask back so it seals and note how quickly the wind dies down. It's not really "wind" and it really doesn't change but the seal quietens things. Perhaps that will reassure you that nothing is wrong and ease your mind and thus you can go quickly back to sleep.

One time I woke up and my nasal pillow was in my ear...yep...blowing those seemingly high winds right into my ear.
Now that woke me up. but I quickly realized what it was and just moved things back in place and things quieted down immediately.

Finally...I never really panic much with odd, rare, random, off beat things showing up like the AHI and leak relationship. I don't worry until I start seeing a pattern and something happening often or nightly on a rather consistent basis. I don't make changes to anything base on a one night report like this. We never know if the ugliness was a fluke or not and chasing flukes means you end up chasing your tail and rarely works out productively.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
robysue
Posts: 7520
Joined: Sat Sep 18, 2010 2:30 pm
Location: Buffalo, NY
Contact:

Re: doing fine until I awake with the machine roaring

Post by robysue » Thu May 05, 2016 4:20 pm

bluetrout wrote: Everything goes great until I wake up, usually with the machine at a high pressure forcing a lot of air which leaks. Can't get back to sleep with that going on so I stop the machine and get to sleep without it for another few hours. I'd like to get the full benefit from the CPAP but not sure how.
As pugsy points out, the highest your pressure went on this night is 10.5, and that was only during the PR Search algorithm's test pressure increases.

There is evidence that you woke up around 0:20 and hit the ramp button. The pressure had been running at around 10cm for about 20 minutes at that time. There was a small, but long leak around that time: The estimated excess leak rate was between 10 and 15 L/min, which is small enough to ignore--except for the fact that you say leaks wake you up.

There's a lot of ugliness going on between 1:20 and 2:00. The excess leak rate spikes to 30 L/min around 1:50. The PR machines base official Large Leaks on the total leak rate, which peaks at 60 L/min, and that's flirting with Large Leak territory. But this leak lasts only 15-20 minutes, and the leak is only flirting with Large Leak territory for 5-10 minutes. Again, that's not enough to worry about---except you say that leaks wake you up. There are several events between 1:20 and 2:00 as well. The pressure is increased from 9cm to 10cm after the leak has spiked. The pressure increase seems to be in response to a small cluster of 2 or 3 Hs along with an OA scored right before the pressure increase. About 5 minutes after the pressure is increased to 10cm, your breathing settles down and the largish leak has disappeared.

It's possible that some of the ugliness between 1:20 and 2:00 is sleep-wake-junk that occurred when you were sort of awake, but still sort of asleep. You may have been barely awake enough to notice the leak and were fiddling to fix it. In any case, it looks like you settled back down into sleep by 2:00am.

You wake up and turn the machine OFF for good at around 3:15. At the time you turned the machine off the pressure was at 9cm and the excess leak rate was below 10 L/min. In other words, the data doesn't agree with your perceptions of what was going on when you woke up and turned the machine off.

So the question becomes: Why are you perceiving the pressure is "high" and that are a lot of leaks when neither is the case?

It could be that you are just "waking up"---possibly from a perfectly normal post-REM wake. But instead of noticing that everything is "normal", your semi-awake brain notices that there's this thing on your nose and that a lot of air is coming from the thing. You're probably noticing the airflow from the exhaust vents in the mask and thinking that you've sprung a serious leak. And you probably are then attributing the "leak" that isn't really a leak to the machine's increased pressure, even though the pressure hasn't increased. And then for some reason you decide that you can't fix the leak quickly enough to get back to sleep. By this time, you've taken the mask off and you go back to sleep.
bluetrout wrote:Hmmm. You're right Pugsy. I was looking at the AHI going up at about 1:45 however the pressure wasn't high. The leak peaked around then too. Wonder if that's what woke me up.
My guess is that the leak around 1:45 did wake you up. But you fixed the leak and the flow rate then returns to what looks like a normal sleep breathing pattern.
The feeling was that the air was pouring into the mask and that it was leaking everywhere. IT seemed loud and so the "roaring".
Was this right before you took the mask off?

If the exhaust vent is aimed in such a way that it hits the bedcovers and reflects back in your face, it can feel like there is air blowing around everywhere and it can be very noisy. If the hose is touching your pillow or your head, the conducted sound of both your breathing and the machine's fan can make it seem like everything is very loud and roaring.

Finally, I'll offer this: When you woke up around 0:20, your instincts were to press the ramp button and go back to sleep. That's not an unreasonable thing to do since it appears that your ramp only lasts for 10 minutes. Perhaps if you'd just pressed the ramp button to get the "excess pressure and air blowing everywhere" under control when you woke up and took the mask off, you would have quickly returned back to sleep.

_________________
Machine: DreamStation BiPAP® Auto Machine
Mask: Swift™ FX Nasal Pillow CPAP Mask with Headgear
Additional Comments: PR System DreamStation and Humidifier. Max IPAP = 9, Min EPAP=4, Rise time setting = 3, minPS = 3, maxPS=5

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: doing fine until I awake with the machine roaring

Post by palerider » Thu May 05, 2016 5:13 pm

bluetrout wrote:Everything goes great until I wake up, usually with the machine at a high pressure forcing a lot of air which leaks.
your pressures aren't high.

work on your mask leaks, look into mask liners.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: doing fine until I awake with the machine roaring

Post by palerider » Thu May 05, 2016 5:14 pm

bluetrout wrote:The feeling was that the air was pouring into the mask and that it was leaking everywhere. IT seemed loud and so the "roaring".
the air is 'pouring' into the mask to try and compensate for the pressure lost due to the leak. fix that, and it all goes back to normal.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

bluetrout
Posts: 92
Joined: Mon Apr 11, 2016 10:10 pm

Re: doing fine until I awake with the machine roaring

Post by bluetrout » Thu May 05, 2016 7:52 pm

A wonderful amount of good information from the posters. Thank you. What you said makes a great deal of sense....and is reassuring. Everything you all noticed rings true. I am sort of amazed by your observations. I'm absorbing the advice and will put it into practice and will report back. Will work on 1. correcting mask leaks 2. awareness of the noise when the mask is away from my face as in a leak situation 3. relaxing when awakened by whatever is going on and returning to sleep.
This forum is impressive. Really impressive.

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: doing fine until I awake with the machine roaring

Post by palerider » Thu May 05, 2016 8:47 pm

I wake up several times at night, (light sleeper) I turn over, poke the mask if needed and try to get back to the dream I was having...

obsessing about waking up just makes it worse.

it's *normal* to almost wake up several times during he night, in between sleep cycles... one doesn't always remember it.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

bluetrout
Posts: 92
Joined: Mon Apr 11, 2016 10:10 pm

Re: doing fine until I awake with the machine roaring

Post by bluetrout » Sun May 08, 2016 4:22 pm

OK it's been a few nights now so I've got some additional data. First I thought I'd respond to the previous posts that offered so much advice. Pugsy, yes I think the "roaring" was the machine leaking and responding to the leak. I've gotten used to that as you suggested I would. Robysue, I think you were correct in suggesting I was just waking up and then overreacting to the mask and the air and the whole weird situation when half awake....and then solving the perceived problem by removing the mask and returning to sleep.
Now to bring things up to date: I am continuing to have difficulty with both FFMs rubbing on the bridge of my nose to the point of causing a sore that doesn't heal. Not a big, worrisome sore but a sore non the less. I have a fabric cover for the Simplus mask and I also use paper tape on my nose for both of them. Still the sore is there. The Simplus is more comfortable but the Comfort Gel Blue FFM seems to give better AHI scores. My preferred mask for comfort is the F&P Eson nasal mask but I wake up when wearing that when I sort of "burp" air. Once that "burping" starts it continues and I cannot get back to sleep. When I "burp" it is not, I think, from mouth breathing but rather air being forced past my closed teeth and lips. The nasal mask does not bother the bridge of my nose.
I'm attaching the graph from Thursday night when I was wearing the blue gel FFM.
Image

bluetrout
Posts: 92
Joined: Mon Apr 11, 2016 10:10 pm

Re: doing fine until I awake with the machine roaring

Post by bluetrout » Sun May 08, 2016 4:30 pm

now Friday night wearing the Simples with cover:

Image
Last edited by bluetrout on Sun May 08, 2016 5:04 pm, edited 1 time in total.

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: doing fine until I awake with the machine roaring

Post by palerider » Sun May 08, 2016 4:34 pm

broken images.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

bluetrout
Posts: 92
Joined: Mon Apr 11, 2016 10:10 pm

Re: doing fine until I awake with the machine roaring

Post by bluetrout » Sun May 08, 2016 4:37 pm

And finally Saturday night wearing the blue gel FFM. This one is a mystery to me. I have no recollection of that long time from midnight to 1:15 or so. I must have turned the machine off and then on but I have no memory of that. I don't know what I am asking for here other than my concern is that I am not getting a good night's sleep and the AHI numbers are not staying down and the masks are a problem with comfort. Overall, I'm pretty tired and foggy and most likely not thinking very clearly.

Image
Last edited by bluetrout on Sun May 08, 2016 5:00 pm, edited 1 time in total.

bluetrout
Posts: 92
Joined: Mon Apr 11, 2016 10:10 pm

Re: doing fine until I awake with the machine roaring

Post by bluetrout » Sun May 08, 2016 4:38 pm

palerider wrote:broken images.
how so? They appear ok to me....

User avatar
Pugsy
Posts: 65328
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: doing fine until I awake with the machine roaring

Post by Pugsy » Sun May 08, 2016 4:52 pm

This is what I get for the Thursday and Friday night images...so Palerider isn't the only one having broken links.
If you can see them and we can't....I don't know what happened.

Image

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.