Schumacher SJ1 with AirSense 10 Autoset

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
EagleJim471
Posts: 4
Joined: Thu Mar 31, 2016 9:20 am

Schumacher SJ1 with AirSense 10 Autoset

Post by EagleJim471 » Thu Mar 31, 2016 9:52 am

I do quite a bit of camping with the Boy Scouts and am new to my CPAP (as in I have only been on it for 2 nights so far). I have a ResMed AirSense 10 Autoset w Humidifier. I am not using any of the heated tubing, just the tubing that came standard with the machine. I do not want to pay $500 or more for the ResMed battery. I have been looking at a Schumacher SJ1 upon the recommendation of a friend with an older CPAP. However, I have never been a good decipherer of electrical speak, and the specs on the AirSense don't easily translate to the specs on the SJ1 for me. I am hoping someone can tell me if this would be an ample portable power source for my CPAP. All I need is a day or two of charge. When I camp longer times, it is almost always near an electrical source.

Specs from ResMed:
90W power supply unit
AC input range:100 – 240V, 50 – 60Hz 1.0 –1.5 A, Class II
115V, 400Hz 1.5A, Class II (nominal for aircraft use)
DC output: 24V 3.75A
Typical power consumption:
53W (57VA)
Peak power consumption:
104W (108VA

Specs for Shumacher SJ1
Peak Amps: 1200A
DC Power Outlet: 12V | 15A
AC Power Outlet: 120V | 1.7A
UBS Ports: 2A PER SIDE | 4A TOTAL
Converter: 120V, 60Hz; 200W continuous/400W peak
Coverter AC waveform: Modified sine wave

So, can I use this device safely for 1 or 2 nights? Should I go ahead and purchase the DC/DC converter from ResMed, and use the DC outlet, or am I just as well off using the standard 90W power supply with the AC power outlet? I would appreciate any and all advice. Thanks in advance.

_________________
MachineMask

HoseCrusher
Posts: 2744
Joined: Tue Oct 12, 2010 6:42 pm

Re: Schumacher SJ1 with AirSense 10 Autoset

Post by HoseCrusher » Thu Mar 31, 2016 10:01 am

IF Shumacher lives up to its specifications, the match looks reasonable.

A question for ResMed would be if the unit will operate properly using a Modified Sine Wave...

If not and IF the specifications are "real" the DC side should work.

_________________
Mask: Brevida™ Nasal Pillow CPAP Mask with Headgear
Additional Comments: Machine is an AirSense 10 AutoSet For Her with Heated Humidifier.
SpO2 96+% and holding...

User avatar
ChicagoGranny
Posts: 15449
Joined: Sun Jan 29, 2012 1:43 pm
Location: USA

Re: Schumacher SJ1 with AirSense 10 Autoset

Post by ChicagoGranny » Thu Mar 31, 2016 11:38 am

EagleJim471 wrote:EagleJim471
Have a look at this thread - viewtopic.php?t=102775

CapnLoki seems to be the resident expert on batteries. If you have questions and post in that thread, he will probably respond.

EagleJim471
Posts: 4
Joined: Thu Mar 31, 2016 9:20 am

Re: Schumacher SJ1 with AirSense 10 Autoset

Post by EagleJim471 » Thu Mar 31, 2016 1:54 pm

ChicagoGranny wrote:
EagleJim471 wrote:EagleJim471
Have a look at this thread - viewtopic.php?t=102775

CapnLoki seems to be the resident expert on batteries. If you have questions and post in that thread, he will probably respond.
Thanks. I'll give him a try.

_________________
MachineMask

EagleJim471
Posts: 4
Joined: Thu Mar 31, 2016 9:20 am

Re: Schumacher SJ1 with AirSense 10 Autoset

Post by EagleJim471 » Thu Mar 31, 2016 2:01 pm

HoseCrusher wrote:IF Shumacher lives up to its specifications, the match looks reasonable.

A question for ResMed would be if the unit will operate properly using a Modified Sine Wave...

If not and IF the specifications are "real" the DC side should work.
Thanks for the reply. I know the SJ1 is a lot heavier than the more expensive batteries made for CPAPs, but if I can get the same performance for a fraction of the price, I am willing to make the compromise.

_________________
MachineMask

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Schumacher SJ1 with AirSense 10 Autoset

Post by palerider » Thu Mar 31, 2016 2:56 pm

HoseCrusher wrote:IF Shumacher lives up to its specifications, the match looks reasonable.

A question for ResMed would be if the unit will operate properly using a Modified Sine Wave...

If not and IF the specifications are "real" the DC side should work.
there's nothing I see in those specs that says what the capacity of the unit is. it might work great, for 30 minutes.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

HoseCrusher
Posts: 2744
Joined: Tue Oct 12, 2010 6:42 pm

Re: Schumacher SJ1 with AirSense 10 Autoset

Post by HoseCrusher » Thu Mar 31, 2016 4:19 pm

Good catch.

It looks like the battery capacity is 18 amp hours. Back of the napkin calculation indicates that it would power that APAP unit for about 4 hours optimistically. Pessimistically about half that time. Realistically somewhere in the middle of those two.

_________________
Mask: Brevida™ Nasal Pillow CPAP Mask with Headgear
Additional Comments: Machine is an AirSense 10 AutoSet For Her with Heated Humidifier.
SpO2 96+% and holding...

User avatar
chunkyfrog
Posts: 34544
Joined: Mon Jul 12, 2010 5:10 pm
Location: Nowhere special--this year in particular.

Re: Schumacher SJ1 with AirSense 10 Autoset

Post by chunkyfrog » Thu Mar 31, 2016 4:40 pm

Considering the cost of proprietary power supplies for my Resmed machine,
I went for the Intellipap Autoadjust, since it will run cheaply off any deep cycle 12 volt DC battery.
Buying into our hosts' generous bundling, I added a few items to my order, including a battery pack.
I spent more than the Resmed battery would have been, but I have a complete backup/travel setup.
Even more good news: my Airsense 10 machine has a completely different power receptacle from my S9.
Yeah, I'd bet a magic crystal there is no (reasonably priced) adaptor.

_________________
Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear
Additional Comments: Airsense 10 Autoset for Her

User avatar
archangle
Posts: 9293
Joined: Sun Mar 27, 2011 11:55 am

Re: Schumacher SJ1 with AirSense 10 Autoset

Post by archangle » Thu Mar 31, 2016 5:18 pm

You will probably get more hours of usage with the ResMed 12V adapter than with the 120V output of the jumper battery. Unfortunately, I don't have enough information about the efficiency of the 120V inverter in the jumper battery. Gut feel is the penalty isn't too high. Maybe 20-30%, but that's just a gut feel.

Gut feel is that it will be borderline, but might run 2 nights. I have a similar Sears Diehard jumper system that ran my S9 for a short night with no problems. (I didn't run it all the way down.)

I think ResMed doesn't specifically say that A10 machines will run off of a modified sine wave source. They don't say it won't, but they don't say it will, the last time I checked. It's PROBABLY OK.

The machine will run longer if you don't turn on the humidifier. You can use it turned off with water in it and you'll get some humidity, or just leave the tank dry. There's an optional sideplate you can get to let you leave the water tank at home.

BTW, my S9 draws an average of less than 1 watt overnight even with humidifier.

Whatever you do, try it at home first.

_________________
Mask: Swift™ FX Nasal Pillow CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: Also SleepyHead, PRS1 Auto, Respironics Auto M series, Legacy Auto, and Legacy Plus
Please enter your equipment in your profile so we can help you.
Click here for information on the most common alternative to CPAP.
If it's midnight and a DME tells you it's dark outside, go and check for yourself.

Useful Links.

User avatar
CapnLoki
Posts: 1162
Joined: Tue Jun 25, 2013 6:26 pm
Location: North East

Re: Schumacher SJ1 with AirSense 10 Autoset

Post by CapnLoki » Thu Mar 31, 2016 6:44 pm

EagleJim471 wrote:I do quite a bit of camping with the Boy Scouts and am new to my CPAP (as in I have only been on it for 2 nights so far). I have a ResMed AirSense 10 Autoset w Humidifier. I am not using any of the heated tubing, just the tubing that came standard with the machine. I do not want to pay $500 or more for the ResMed battery. I have been looking at a Schumacher SJ1 upon the recommendation of a friend with an older CPAP. However, I have never been a good decipherer of electrical speak, and the specs on the AirSense don't easily translate to the specs on the SJ1 for me. I am hoping someone can tell me if this would be an ample portable power source for my CPAP. All I need is a day or two of charge. When I camp longer times, it is almost always near an electrical source.

Specs from ResMed:
90W power supply unit
AC input range:100 – 240V, 50 – 60Hz 1.0 –1.5 A, Class II
115V, 400Hz 1.5A, Class II (nominal for aircraft use)
DC output: 24V 3.75A
Typical power consumption:
53W (57VA)
Peak power consumption:
104W (108VA

Specs for Shumacher SJ1
Peak Amps: 1200A
DC Power Outlet: 12V | 15A
AC Power Outlet: 120V | 1.7A
UBS Ports: 2A PER SIDE | 4A TOTAL
Converter: 120V, 60Hz; 200W continuous/400W peak
Coverter AC waveform: Modified sine wave

So, can I use this device safely for 1 or 2 nights? Should I go ahead and purchase the DC/DC converter from ResMed, and use the DC outlet, or am I just as well off using the standard 90W power supply with the AC power outlet? I would appreciate any and all advice. Thanks in advance.
There are several relevant factors:

First, the battery is about 18 Amp-Hours. If you want it to last, you shouldn't go much below 50% (20% is the "hard limit" but at that point the lifetime is compromised) so figure about 10 Amp-Hours as the "useful" charge. Next, the inverter efficiency is is about 60% so that brings you down to about 6 Amp-hours. (If you get the 12-24 volt converter you can save most of this hit.)

So how much do you use? I haven't measured a ResMed but the comparable Respironics is about 4-6 Amp-hours a night without humidity. Even minimal humidity can easily double or triple this, so unless you're camping in the desert just forget about humidity. Bottom line, this system is good for about one night. With the converter you might stretch it to two nights.

As an alternative, I would suggest the system in the last post of this thread:
viewtopic.php?t=102775
The battery is same type as the in the Shumacher, but twice as large.The charger is high quality. You should get the 12-24 converter so you'll have plenty of power, enough for several nights, or enough to serve as power failure backup at home. The weight of the battery is about the same as the Shumacher.

_________________
Machine: DreamStation Auto CPAP Machine
Mask: Quattro™ Air Full Face Mask with Headgear
Humidifier: DreamStation Heated Humidifier
Additional Comments: Pressure 9-20, average ~9.5; often use battery power while off-grid
Hark, how hard he fetches breath . . .  Act II, Scene IV, King Henry IV Part I, William Shakespeare
Choosing a Battery thread: http://www.cpaptalk.com/viewtopic/t1140 ... ttery.html