FF masks without support on the forehead! A real mystery !!!
- Oltremare
- Posts: 187
- Joined: Tue Dec 29, 2015 7:01 pm
- Location: On the other side of the sea - Italy
FF masks without support on the forehead! A real mystery !!!
My favorite mask is Resmed Mirage Quattro FF.
I bought a Airfit F10 and a Mirage Quattro FX FF.
Both masks without support forehead. I wear them faithfully respecting the instructions but it's not good! During sleep, both slide towards masks on until the base of the nose, and reach to touch my eyes with much hassle.
I doon't understand why this happens. My nose is not too big. My skin is not too oily.
I spent a lot of money for much hope of comfort but it is not working !!!
Does anyone know why this happens?
Thanks so much.
I bought a Airfit F10 and a Mirage Quattro FX FF.
Both masks without support forehead. I wear them faithfully respecting the instructions but it's not good! During sleep, both slide towards masks on until the base of the nose, and reach to touch my eyes with much hassle.
I doon't understand why this happens. My nose is not too big. My skin is not too oily.
I spent a lot of money for much hope of comfort but it is not working !!!
Does anyone know why this happens?
Thanks so much.
_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier |
Mask: ResMed AirFit™ F20 Full Face CPAP Mask with Headgear |
Additional Comments: Quattro FX FF mask > CMS50D+ pulse oximeter > Climate line > Settings: auto-CPAP pressure Min 11.00 Max 18.0 EPR 3 |
I hope you will forgive my language errors. To write in English I use a translator online. Thank you
Re: FF masks without support on the forehead! A real mystery !!!
Not sure why it happens, but since it does, you might want to look at other styles (on Cpap.com where there's a large variety).
Last edited by Julie on Wed Feb 17, 2016 4:34 pm, edited 1 time in total.
- Oltremare
- Posts: 187
- Joined: Tue Dec 29, 2015 7:01 pm
- Location: On the other side of the sea - Italy
Re: FF masks without support on the forehead! A real mystery !!!
Yeah, maybe I should change.
I think of the hybrids because only nasal masks are not good for me.
For now, however, I must look for a solution for the two new masks I've just bought.
I can't send them in the back.
Not even I can throw them.
Thank you
I think of the hybrids because only nasal masks are not good for me.
For now, however, I must look for a solution for the two new masks I've just bought.
I can't send them in the back.
Not even I can throw them.
Thank you
_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier |
Mask: ResMed AirFit™ F20 Full Face CPAP Mask with Headgear |
Additional Comments: Quattro FX FF mask > CMS50D+ pulse oximeter > Climate line > Settings: auto-CPAP pressure Min 11.00 Max 18.0 EPR 3 |
I hope you will forgive my language errors. To write in English I use a translator online. Thank you
- ChicagoGranny
- Posts: 15350
- Joined: Sun Jan 29, 2012 1:43 pm
- Location: USA
Re: FF masks without support on the forehead! A real mystery !!!
Hang on to them. You never know what changes in the future, and you might develop a liking to themOltremare wrote:For now, however, I must look for a solution for the two new masks I've just bought.
I am hesitant to recommend the mask I use because I am personally biased. But, this mask fits firm on my face and has worked perfectly for me - https://www.cpap.com/productpage/RespCa ... -Mask.html
Re: FF masks without support on the forehead! A real mystery !!!
If the Quattro is your favorite...WHY CHANGE ?Oltremare wrote:My favorite mask is Resmed Mirage Quattro FF.
I bought a Airfit F10 and a Mirage Quattro FX FF.
Both masks without support forehead. I wear them faithfully respecting the instructions but it's not good! During sleep, both slide towards masks on until the base of the nose, and reach to touch my eyes with much hassle.
I doon't understand why this happens. My nose is not too big. My skin is not too oily.
I spent a lot of money for much hope of comfort but it is not working !!!
Does anyone know why this happens?
Thanks so much.
_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier |
Mask: Fisher & Paykel Vitera Full Face Mask with Headgear (S, M, or L Cushion) |
Additional Comments: Back up is a new AS10. |
- Oltremare
- Posts: 187
- Joined: Tue Dec 29, 2015 7:01 pm
- Location: On the other side of the sea - Italy
Re: FF masks without support on the forehead! A real mystery !!!
Quattro is my favorite but it's not perfect. I have some problems, marks on my nose and is too bulky! I imagine summer, how hot will feel with Four?LSAT wrote:If the Quattro is your favorite...WHY CHANGE ?
All seek to improve our therapy.
_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier |
Mask: ResMed AirFit™ F20 Full Face CPAP Mask with Headgear |
Additional Comments: Quattro FX FF mask > CMS50D+ pulse oximeter > Climate line > Settings: auto-CPAP pressure Min 11.00 Max 18.0 EPR 3 |
I hope you will forgive my language errors. To write in English I use a translator online. Thank you
Re: FF masks without support on the forehead! A real mystery !!!
A couple ideas to look at.
F & P 431 has a section that rests under the chin. It can't slide up. I have used this one and at the time it was the only FF I could tolerate because the bottom of all the others would wind up in my mouth, like a bit in a horses mouth.
Amara View rests against the nostrils. It can't slide up.
Oracle is mouth only and never touches the nose.
I don't know if these will work for you but they will each address the problem you have.
F & P 431 has a section that rests under the chin. It can't slide up. I have used this one and at the time it was the only FF I could tolerate because the bottom of all the others would wind up in my mouth, like a bit in a horses mouth.
Amara View rests against the nostrils. It can't slide up.
Oracle is mouth only and never touches the nose.
I don't know if these will work for you but they will each address the problem you have.
_________________
Mask: SleepWeaver 3D Soft Cloth Nasal CPAP Mask with Headgear |
Additional Comments: AurCurve 10 ASV Also using Sleaplyhead 1.1, ResScan 6 and CMS50i |
- Oltremare
- Posts: 187
- Joined: Tue Dec 29, 2015 7:01 pm
- Location: On the other side of the sea - Italy
Re: FF masks without support on the forehead! A real mystery !!!
Thanks wexler, I will study your ideas, they are very interesting.bwexler wrote:A couple ideas to look at.
F & P 431 has a section that rests under the chin. It can't slide up. I have used this one and at the time it was the only FF I could tolerate because the bottom of all the others would wind up in my mouth, like a bit in a horses mouth.
Amara View rests against the nostrils. It can't slide up.
Oracle is mouth only and never touches the nose.
I don't know if these will work for you but they will each address the problem you have.
Before doing this, I would understand if I can solve the problem with the 2 masks I've just bought.
Has anyone had the same problem of slipping?
_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier |
Mask: ResMed AirFit™ F20 Full Face CPAP Mask with Headgear |
Additional Comments: Quattro FX FF mask > CMS50D+ pulse oximeter > Climate line > Settings: auto-CPAP pressure Min 11.00 Max 18.0 EPR 3 |
I hope you will forgive my language errors. To write in English I use a translator online. Thank you
Re: FF masks without support on the forehead! A real mystery !!!
Several users have reported having their mask slide up during the night. They made a strap, using either parts of old head gear or velcro, that will attach to one lower mask strap and go under the chin to the opposite strap.Oltremare wrote:My favorite mask is Resmed Mirage Quattro FF.
I bought a Airfit F10 and a Mirage Quattro FX FF.
Both masks without support forehead. I wear them faithfully respecting the instructions but it's not good! During sleep, both slide towards masks on until the base of the nose, and reach to touch my eyes with much hassle.
I doon't understand why this happens. My nose is not too big. My skin is not too oily.
I spent a lot of money for much hope of comfort but it is not working !!!
Does anyone know why this happens?
Thanks so much.
I hope this may help. (and translate correctly)
_________________
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear |
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control |
Additional Comments: CMS-50F & 50IW - SleepyHead 1.0.0-Beta |
Re: FF masks without support on the forehead! A real mystery !!!
Have you tried a liner? They helped to stablize my husband's F10. PadaCheek is nice. RemZzz's gives free samples. PM Becky at RemZzz stated that on this page. viewtopic/p1057667/Especially-for-newly ... l#p1057667 and you can cut your own from old tee shirt material.
_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear |
Additional Comments: Sleepyhead Software |
ResMed Aircurve 10 VAUTO EPAP 11 IPAP 15 / P10 pillows mask / Sleepyhead Software / Back up & travel machine Respironics 760
- ChicagoGranny
- Posts: 15350
- Joined: Sun Jan 29, 2012 1:43 pm
- Location: USA
Re: FF masks without support on the forehead! A real mystery !!!
How is your hose management? There should be slack in the hose so that it does not pull on the mask.Oltremare wrote:Has anyone had the same problem of slipping?
- Okie bipap
- Posts: 3567
- Joined: Thu Oct 15, 2015 4:14 pm
- Location: Central Oklahoma
Re: FF masks without support on the forehead! A real mystery !!!
When I was using the F&P Simplus, I used to run two pieces of tape trombone mask to down under my chin. This kept the mask from riding up on my face.
_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
Mask: Evora Full Face Mask - Fitpack |
Additional Comments: IPAP 20-25, ps 4, OSCAR software |
Growing old is mandatory, but growing up is optional.
Re: FF masks without support on the forehead! A real mystery !!!
autoINcorrect?Okie bipap wrote:When I was using the F&P Simplus, I used to run two pieces of tape trombone mask to down under my chin. This kept the mask from riding up on my face.
Get OSCAR
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
- Okie bipap
- Posts: 3567
- Joined: Thu Oct 15, 2015 4:14 pm
- Location: Central Oklahoma
Re: FF masks without support on the forehead! A real mystery !!!
That's what I get for trying to use my tablet instead of my computer.
_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
Mask: Evora Full Face Mask - Fitpack |
Additional Comments: IPAP 20-25, ps 4, OSCAR software |
Growing old is mandatory, but growing up is optional.
Re: FF masks without support on the forehead! A real mystery !!!
It sounds like your head size or shape may be preventing the headgear from resting low enough for the straps to pull the masks to your face at the proper angle, or, the headgear's size is just too small for you. A chin strap that runs between the two lower straps could be a good solution, if you can tolerate it. Good luck.
_________________
Mask: ResMed AirFit™ F20 Full Face CPAP Mask with Headgear |
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control |