Newbie here with a ramp question

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
User avatar
Hopeful50
Posts: 177
Joined: Fri Feb 12, 2016 3:50 pm
Location: Cupertino, CA

Newbie here with a ramp question

Post by Hopeful50 » Sat Feb 13, 2016 9:32 am

Newly diagnosed, and just picked up my machine two days ago. Doc prescribed a pressure setting of 10 with 30-minute ramp. The therapist (she was awesome!) set the machine to start at 5, with the autoramp feature ON, so it wouldn't start to ramp up until I was asleep. First night was fine . . . it took me over 40 minutes to fall asleep, but the level stayed at 5 while I was awake. Last night (second night), within 5-7 minutes of startup, the machine would quickly ramp up to a pressure of 10 over a period of 2-3 minutes. Several times I started over, only to have the same thing happen again. Finally I changed the ramp feature from auto to 45 minutes and was able to fall asleep that way.

Any ideas what I've done to mess up the Auto Ramp feature?

_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: Pressure 13-20, EPR 3

User avatar
Julie
Posts: 20035
Joined: Tue Feb 28, 2006 12:58 pm

Re: Newbie here with a ramp question

Post by Julie » Sat Feb 13, 2016 9:42 am

Hi, your prescribed pressure is not an unusually high one by any means, and setting such a long time for it to be reached (using the ramp) does mean you're not fully treated for all that time. Most of us actually stop using the ramp altogether within a few days unless our scripted pressure is very high or we're particularly sensitive to any pressure. I would suggest that you possibly lower the time setting by a fair amount and try to relax, go to sleep without continuously resetting the ramp downward, essentially starting its climb over again. I bet you can get used to the air 'blast' quickly if you try and eventually lower and/or get off the ramp entirely in future. It may seem like a hurdle now, but I think you could overcome it soon if you allow it to happen, you won't be blasted out of bed and will be fully treated much sooner.
Last edited by Julie on Sat Feb 13, 2016 9:44 am, edited 1 time in total.

Cardsfan
Posts: 1509
Joined: Wed Apr 29, 2015 3:02 pm
Location: Close to St. Louis, MO

Re: Newbie here with a ramp question

Post by Cardsfan » Sat Feb 13, 2016 9:44 am

Do you have the manual that came with the machine? Or you can find it online.

When ramp is set to Auto, the device will detect when you have fallen asleep and the automatically rise to the prescribed treatment pressure. You can either change it to manual and pick a set time. Or, adjust up your starting pressure to 6 or 7 and your jump up to 10 won't seem so noticeable. Many of us find a setting of 5 uncomfortable, like we are starving for air.

_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier
Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear
Additional Comments:  CPAP 10 cmH20., User since 1/1/15.

User avatar
Hopeful50
Posts: 177
Joined: Fri Feb 12, 2016 3:50 pm
Location: Cupertino, CA

Re: Newbie here with a ramp question

Post by Hopeful50 » Sat Feb 13, 2016 10:12 am

Sorry, Cardsfan, I guess I didn't explain properly. The ramp was initially set to Auto, but behaved as is it was detecting that I was asleep when I really wasn't. (Still had the light on while reading.)

_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: Pressure 13-20, EPR 3

User avatar
K5MOW
Posts: 155
Joined: Mon Oct 04, 2010 9:50 am
Location: Friendswood TX
Contact:

Re: Newbie here with a ramp question

Post by K5MOW » Sat Feb 13, 2016 10:22 am

Cardsfan wrote:Do you have the manual that came with the machine? Or you can find it online.

When ramp is set to Auto, the device will detect when you have fallen asleep and the automatically rise to the prescribed treatment pressure. You can either change it to manual and pick a set time. Or, adjust up your starting pressure to 6 or 7 and your jump up to 10 won't seem so noticeable. Many of us find a setting of 5 uncomfortable, like we are starving for air.
I use the ramp maybe the first couple of days and now I can't stay on the ramp it feels like you are starving for air. I have been on CPAP now about six years.

Roger
Moderate to Severe Sleep Apnea

Airsense 10 Autoset AirStart 10 CPAP

Mirage Quattro Full Face Mask

Roger

Cardsfan
Posts: 1509
Joined: Wed Apr 29, 2015 3:02 pm
Location: Close to St. Louis, MO

Re: Newbie here with a ramp question

Post by Cardsfan » Sat Feb 13, 2016 10:32 am

The machine does not know if you are awake or asleep. It is responding to your breathing patterns. The sensation of a gust of air will go away after a few days. Your body adjusts to it.
Are you asking if something is wrong with the machine?

_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier
Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear
Additional Comments:  CPAP 10 cmH20., User since 1/1/15.

User avatar
OkyDoky
Posts: 2870
Joined: Mon Aug 25, 2014 5:18 pm

Re: Newbie here with a ramp question

Post by OkyDoky » Sat Feb 13, 2016 11:10 am

This is what auto ramp is supposed to do from your Clinical Manual and here is a link to the manual. https://sleep.tnet.com/home/files/resme ... -guide.pdf

In CPAP mode, the pressure increases from a low pressure (Start Pressure) to the prescribed
treatment pressure.
Ramp Time can be set to Off, 5 to 45 minutes or Auto. When Ramp Time is set to Auto, the device
will detect sleep onset and then gradually increase from the start pressure to the minimum
treatment pressure at a rate of 1 cm H2O per minute. However, if sleep onset is not detected, the
device will reach the target pressure within 30 minutes.
ResMed Aircurve 10 VAUTO EPAP 11 IPAP 15 / P10 pillows mask / Sleepyhead Software / Back up & travel machine Respironics 760

User avatar
Hopeful50
Posts: 177
Joined: Fri Feb 12, 2016 3:50 pm
Location: Cupertino, CA

Re: Newbie here with a ramp question

Post by Hopeful50 » Sat Feb 13, 2016 1:23 pm

Yes, Cardsfan and Okydoky, I'm trying to figure out why the machine isn't behaving like the book says it should. I got into bed with everything in place, the machine started up (with Ramp set to Auto) and I started reading a book. Within 5 minutes I could feel the pressure increasing, checked the display, and it was moving through 7.2, 7.3, 7.4 . . . and hit 10 within two minutes. It wasn't uncomfortable, it wasn't scary, it's just not what's supposed to happen with Ramp set to Auto. The only thing I can figure is that my breathing was so regular it mimicked sleep.

_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: Pressure 13-20, EPR 3

User avatar
OkyDoky
Posts: 2870
Joined: Mon Aug 25, 2014 5:18 pm

Re: Newbie here with a ramp question

Post by OkyDoky » Sat Feb 13, 2016 1:56 pm

Well it does state stable breathing indicates patient has fallen asleep. So maybe you fooled it.
Image
ResMed Aircurve 10 VAUTO EPAP 11 IPAP 15 / P10 pillows mask / Sleepyhead Software / Back up & travel machine Respironics 760

User avatar
Goofproof
Posts: 16087
Joined: Mon Dec 05, 2005 3:16 pm
Location: Central Indiana, USA

Re: Newbie here with a ramp question

Post by Goofproof » Sat Feb 13, 2016 2:07 pm

Hopeful50 wrote:Yes, Cardsfan and Okydoky, I'm trying to figure out why the machine isn't behaving like the book says it should. I got into bed with everything in place, the machine started up (with Ramp set to Auto) and I started reading a book. Within 5 minutes I could feel the pressure increasing, checked the display, and it was moving through 7.2, 7.3, 7.4 . . . and hit 10 within two minutes. It wasn't uncomfortable, it wasn't scary, it's just not what's supposed to happen with Ramp set to Auto. The only thing I can figure is that my breathing was so regular it mimicked sleep.
The whole idea is to get to sleep with your XPAP providing the needed treatment to hold your airway open as you sleep. Your watching and playing with your XPAP, instead of just using it, make the ramp counterproductive for treatment.

Many find the ramp uncessary, for them it just one more distraction. Jim
Use data to optimize your xPAP treatment!

"The art of medicine consists in amusing the patient while nature cures the disease." Voltaire

User avatar
Krelvin
Posts: 1977
Joined: Tue Jun 06, 2006 5:23 pm
Location: Metro Phx Area - Dry Heat!

Re: Newbie here with a ramp question

Post by Krelvin » Sat Feb 13, 2016 2:19 pm

Many just turn the ramp off and never use it.
Current Settings PS 4.0 over 10.6-18.0 (cmH2O) - Resmed S9 VPAP Auto w/h5i Humidifier - Quattro Air FFM
TNET Sleep Resource Pages - CPAP Machine Database
Put your equip in your Signature - SleepyHead v1.0.0-beta-1
Kevin... alias Krelvin

User avatar
Okie bipap
Posts: 3566
Joined: Thu Oct 15, 2015 4:14 pm
Location: Central Oklahoma

Re: Newbie here with a ramp question

Post by Okie bipap » Sat Feb 13, 2016 5:01 pm

When I first started treatment, I had my ramp set on 45 minutes, but I found I would lay awake until it came up to full pressure. After a few days, I decided to just turn the ramp off and haven't used it since.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Mask: Evora Full Face Mask - Fitpack
Additional Comments: IPAP 20-25, ps 4, OSCAR software
Growing old is mandatory, but growing up is optional.

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Newbie here with a ramp question

Post by palerider » Sat Feb 13, 2016 5:26 pm

Hopeful50 wrote:Any ideas what I've done to mess up the Auto Ramp feature?
autoramp works by looking at your breathing pattern and making a calculated guess as to whether you're awake or asleep.

you must have been breathing in such a manner that the machine thought you were asleep, at which point it starts increasing pressure by one cm per minute till it gets to your prescribed pressure.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Newbie here with a ramp question

Post by palerider » Sat Feb 13, 2016 5:27 pm

Hopeful50 wrote:It wasn't uncomfortable, it wasn't scary,
then ignore it and let it do its thing.