auto on/off doesn't work properly

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
prodoula

auto on/off doesn't work properly

Post by prodoula » Thu Feb 11, 2016 10:32 am

I have a Resmed AirSense 10 AutoSet for Her. I have the auto on/off turned on. The machine comes on automatically when I put the mask on, but doesn't turn off automatically. I've let it run for up to 10 minutes after taking the mask off and it still doesn't shut off. The auto off feature did work about 2 times when I first go the machine about 2 weeks ago, but after the first couple of times it stopped working. Any suggestions?

User avatar
Julie
Posts: 20051
Joined: Tue Feb 28, 2006 12:58 pm

Re: auto on/off doesn't work properly

Post by Julie » Thu Feb 11, 2016 10:48 am

Turn it off when you're not using it. It should still come on auto. when you want it to.

User avatar
kteague
Posts: 7784
Joined: Tue May 16, 2006 8:30 pm
Location: West and Midwest

Re: auto on/off doesn't work properly

Post by kteague » Thu Feb 11, 2016 5:28 pm

I personally don't use the auto on or auto off feature after a few times "back in the day" having them malfunction. I would do what Julie suggests and just not use the feature. However, if you have a warranteed machine and it is indeed malfunctioning, that is a consideration. Maybe some others who use that particular machine can say if theirs works properly or if this machine is prone to this issue.

_________________
Mask: TAP PAP Nasal Pillow CPAP Mask with Improved Stability Mouthpiece
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: Bleep/DreamPort for full nights, Tap Pap for shorter sessions

User avatar
K5MOW
Posts: 155
Joined: Mon Oct 04, 2010 9:50 am
Location: Friendswood TX
Contact:

Re: auto on/off doesn't work properly

Post by K5MOW » Thu Feb 11, 2016 5:32 pm

I love the auto on feature with any CPAP. But I have never used the auto off.
Moderate to Severe Sleep Apnea

Airsense 10 Autoset AirStart 10 CPAP

Mirage Quattro Full Face Mask

Roger

User avatar
chunkyfrog
Posts: 34544
Joined: Mon Jul 12, 2010 5:10 pm
Location: Nowhere special--this year in particular.

Re: auto on/off doesn't work properly

Post by chunkyfrog » Thu Feb 11, 2016 5:38 pm

I use the auto feature to turn my machine on.
I shut it off manually, because of the noise and irritating delay.
I also suspect this feature was designed primarily for cushion-type masks,
and may only work sporadically with nasal cushions or even low pressure.

_________________
Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear
Additional Comments: Airsense 10 Autoset for Her

User avatar
stienman
Posts: 89
Joined: Thu Feb 04, 2016 12:40 pm
Location: Ann Arbor, MI

Re: auto on/off doesn't work properly

Post by stienman » Fri Feb 12, 2016 8:23 am

What hose and mask are you using? Are you sure you have it detect both on and off, I believe you can set it to auto start without auto stopping, so maybe the auto stopping is turned off.

I have an S9 which should be similar in how it detects usage, and I keep the auto on/off turned on, and it works fine. I like this feature because I'm also using a separate pulse oximeter logger, and I can get it ready to record, slip the nasal pillows on and breath into it while pressing the record button on the oximeter. This allows sleepyhead to align the two separate recordings very closely.

The auto off function works well too - I slip it off, set it aside, and within a second or two it turns on the display and starts the cool down cycle.

Make sure you have the correct mask type selected in the machine (may need the clinical manual to change this).

Also if you have additional hoses, or a non thin hose (it's designed for the slimline or heated climate line hoses) then the algorithm may not detect a mask off event. Try it with the official hose, and only one connected if you have any extensions or unusual setups.

If you're using it with a mask/pillow/etc that isn't a Resmed brand, try each of the mask type settings just in case the outflow rate is different enough that it doesn't detect it.

Lastly, it should definitely detect a mask off event if you disconnect the hose from the mask hose. This probably isn't as convenient as pressing the button on the machine, though.

_________________
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: CMS50F PulseOx, Sleepyhead

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: auto on/off doesn't work properly

Post by palerider » Fri Feb 12, 2016 8:35 am

stienman wrote:What hose and mask are you using? Are you sure you have it detect both on and off, I believe you can set it to auto start without auto stopping, so maybe the auto stopping is turned off.
only respironics have separate on/off settings.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

Thatgirl
Posts: 385
Joined: Wed Sep 17, 2014 2:13 pm

Re: auto on/off doesn't work properly

Post by Thatgirl » Fri Feb 12, 2016 9:11 am

Yeah, same issue here. Auto on works fine. It doesn't go off, though. I just turn it off. I'm sure it's because of mask I'm using.

_________________
MachineMask
Additional Comments: Vauto EPAP min 5, IPAP max 25, PS 6
Aircurve 10 Vauto, Dreamwear mask, Vauto mode: EPAP min 5, IPAP max 20, PS 6
Usually ~EPAP 6, IPAP 12
Loving Sleepyhead!

User avatar
Okie bipap
Posts: 3567
Joined: Thu Oct 15, 2015 4:14 pm
Location: Central Oklahoma

Re: auto on/off doesn't work properly

Post by Okie bipap » Fri Feb 12, 2016 10:00 am

I tried using the auto start option far about a month, but I ended up turning it off. If I put my mask on and wanted to use the mask fit function, the therapy would start before I could start the mask fit function. Some night it takes a couple tries to get my mask to seal properly.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Mask: Evora Full Face Mask - Fitpack
Additional Comments: IPAP 20-25, ps 4, OSCAR software
Growing old is mandatory, but growing up is optional.

prodoula

Re: auto on/off doesn't work properly

Post by prodoula » Fri Feb 12, 2016 10:16 am

To answer some questions, I use the climate line resmed hose, and the resmed p10 nasal pillows. I have my machine mask setting on pillows. My machine only has auto on/off, not separate auto on and auto off.

I toyed with the machine quite a bit last night and found that if EPR is turned off, the auto off works great. If EPR is on and set to 3 it doesn't work. 2 and 1 work intermittently, with 1 working better than 2.

So I think to answer my own question, the auto off doesn't work (or works well) based on the EPR setting. I prefer an EPR of 3, so I guess I'll just continue to turn the machine off manually when I take the mask off.

But I do think ResMed should have written in the clinician's manual that auto off only consistently works with EPR turn off.

Anyway, I hope this helps someone else who is stumped because the advertised "auto off" feature doesn't work on their machine.

SewTired
Posts: 1736
Joined: Thu Apr 16, 2015 8:33 am
Location: Minneapolis area

Re: auto on/off doesn't work properly

Post by SewTired » Fri Feb 12, 2016 12:38 pm

prodoula wrote:To answer some questions, I use the climate line resmed hose, and the resmed p10 nasal pillows. I have my machine mask setting on pillows. My machine only has auto on/off, not separate auto on and auto off.

I toyed with the machine quite a bit last night and found that if EPR is turned off, the auto off works great. If EPR is on and set to 3 it doesn't work. 2 and 1 work intermittently, with 1 working better than 2.

So I think to answer my own question, the auto off doesn't work (or works well) based on the EPR setting. I prefer an EPR of 3, so I guess I'll just continue to turn the machine off manually when I take the mask off.

But I do think ResMed should have written in the clinician's manual that auto off only consistently works with EPR turn off.

Anyway, I hope this helps someone else who is stumped because the advertised "auto off" feature doesn't work on their machine.
You posted before I had time too! YES, EPR affects auto off! I thought it was just because I had changed to a different mask, but then discovered that the auto-off stopped working when I put on the EPR.

_________________
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Additional Comments: Sleepyhead
Diabetes 2, RLS & bradycardia
Airsense For Her; Settings: range 8-12, Airfit P10 (M)

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: auto on/off doesn't work properly

Post by palerider » Fri Feb 12, 2016 8:41 pm

EPR lowers your effective pressure.

if the pressure is too low, the machine can have trouble detecting the mask off situation.

raise your min pressure a couple, and even with EPR, it should still detect mask off and shut down.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.