Still extremely fatigued on ASV, help?

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
Vennechris
Posts: 4
Joined: Thu Feb 04, 2016 3:07 pm

Still extremely fatigued on ASV, help?

Post by Vennechris » Thu Feb 04, 2016 4:09 pm

After two sleep studies showing AHI of 34 CA and 1 OA, I was prescribed the Aircurve 10 ASV, EPAP: 4, Min PS: 6, Max PS: 12.
After 8 months of consistent ASV usage, 8+ hours per night sleep with machine, with consistent AHI 0.0 - 0.4. I am still extremely fatigued, even more so than before treatment. Leak levels are very low. Flow limitation spikes occur throughout the night, but Sleep doc says they are meaningless, sleep doc is out of ideas and has no further help to offer. Consistently waking up exhausted, in a fog, with red glassy eyes, worsening memory problems, headaches. Adding Nuvigil makes slogging through the day bearable, but this is taking a toll. Any help would be greatly appreciated.

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Still extremely fatigued on ASV, help?

Post by palerider » Thu Feb 04, 2016 4:15 pm

Vennechris wrote:After two sleep studies showing AHI of 34 CA and 1 OA, I was prescribed the Aircurve 10 ASV, EPAP: 4, Min PS: 6, Max PS: 12.
After 8 months of consistent ASV usage, 8+ hours per night sleep with machine, with consistent AHI 0.0 - 0.4. I am still extremely fatigued, even more so than before treatment. Leak levels are very low. Flow limitation spikes occur throughout the night, but Sleep doc says they are meaningless, sleep doc is out of ideas and has no further help to offer. Consistently waking up exhausted, in a fog, with red glassy eyes, worsening memory problems, headaches. Adding Nuvigil makes slogging through the day bearable, but this is taking a toll. Any help would be greatly appreciated.
download sleepyhead, post some charts.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

Vennechris
Posts: 4
Joined: Thu Feb 04, 2016 3:07 pm

Re: Still extremely fatigued on ASV, help?

Post by Vennechris » Thu Feb 04, 2016 5:29 pm

Image

Image

Image

Image

JDS74
Posts: 3397
Joined: Sun Jan 23, 2011 2:57 pm
Location: South Carolina

Re: Still extremely fatigued on ASV, help?

Post by JDS74 » Thu Feb 04, 2016 5:53 pm

Check your PM.

_________________
Mask: Oracle HC452 Oral CPAP Mask
Humidifier: DreamStation Heated Humidifier
Additional Comments: EverFlo Q 3.0 Liters O2 PR DSX900 ASV
Oracle 452 Lessons Learned Updated
DSX900 AutoSV with HC150 extra humidifier and Hibernite heated hose
Settings: EPAP Min-10.0, EPAP Max-17, PS Min-3, PS Max-10, Max Pressure-20, Rate-Auto, Biflex-1.
Sleepyhead and Encore Pro 2.21.

Sleeprider
Posts: 1562
Joined: Tue May 06, 2008 5:57 pm
Location: Murrysville, PA

Re: Still extremely fatigued on ASV, help?

Post by Sleeprider » Thu Feb 04, 2016 6:00 pm

Vennechris, obviously there is a disconnect between AHI results and how you feel. see sI ome noticeable flow limitation, and might suggest an increase in EPAP min. Outside of the apnea treatment, have you had workups for testosterone and other possible sources? Are there any other health complications that might contribute to how you feel?

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Additional Comments: Sleepyhead software. Just changed from PRS1 BiPAP Auto DS760TS

Vennechris
Posts: 4
Joined: Thu Feb 04, 2016 3:07 pm

Re: Still extremely fatigued on ASV, help?

Post by Vennechris » Thu Feb 04, 2016 6:29 pm

No other health issues outside of the centrals. Have had nearly a full blood workup, (no vitamin tests), thyroid check, heart echo, etc. Going to see neurologist next week to exclude that. Running out of options, was hoping you folks might see something I don't. Will try raising EPAP to knock out flow limitations.

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Still extremely fatigued on ASV, help?

Post by palerider » Thu Feb 04, 2016 7:16 pm

I second the notion of raising the epap a bit, if it were me, I'd raise the maxps some, since you appear to be maxing out on that at times.

I'd like to see your minute vent trace.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Still extremely fatigued on ASV, help?

Post by palerider » Thu Feb 04, 2016 7:17 pm

Sleeprider wrote:Vennechris, obviously there is a disconnect between AHI results and how you feel. see sI ome noticeable flow limitation, and might suggest an increase in EPAP min. Outside of the apnea treatment, have you had workups for testosterone and other possible sources? Are there any other health complications that might contribute to how you feel?
crazy flow limitations traces seem to be a feature of the resmed ASV machines, I haven't ever seen an explanation for it.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

Vennechris
Posts: 4
Joined: Thu Feb 04, 2016 3:07 pm

Re: Still extremely fatigued on ASV, help?

Post by Vennechris » Thu Feb 04, 2016 7:50 pm

Image

Image

Here is the minute vent

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Still extremely fatigued on ASV, help?

Post by palerider » Thu Feb 04, 2016 8:03 pm

Vennechris wrote:Here is the minute vent
it looks ok, I was wanting to see if there were drops in it at any points.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

Sleeprider
Posts: 1562
Joined: Tue May 06, 2008 5:57 pm
Location: Murrysville, PA

Re: Still extremely fatigued on ASV, help?

Post by Sleeprider » Thu Feb 04, 2016 8:41 pm

I have to admit, I'm a bit flummoxed. Even your tidal volume looks great. Hope the increased EPAP helps a bit.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Additional Comments: Sleepyhead software. Just changed from PRS1 BiPAP Auto DS760TS

Guest

Re: Still extremely fatigued on ASV, help?

Post by Guest » Fri Feb 05, 2016 5:33 am

Do any ASV users out there feel that the pressure changes prevent them from reaching a deeper level of sleep?

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Still extremely fatigued on ASV, help?

Post by palerider » Fri Feb 05, 2016 2:05 pm

Guest wrote:Do any ASV users out there feel that the pressure changes prevent them from reaching a deeper level of sleep?
not as much as having a ton of centrals would.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
chrisj
Posts: 23
Joined: Mon Jan 18, 2021 12:06 pm

Re: Still extremely fatigued on ASV, help?

Post by chrisj » Thu Mar 18, 2021 6:01 am

Hi Vennechris,

Did you ever get to the bottom of this? Seems a little like my experience too.

Cheers

Vennechris wrote:
Thu Feb 04, 2016 4:09 pm
After two sleep studies showing AHI of 34 CA and 1 OA, I was prescribed the Aircurve 10 ASV, EPAP: 4, Min PS: 6, Max PS: 12.
After 8 months of consistent ASV usage, 8+ hours per night sleep with machine, with consistent AHI 0.0 - 0.4. I am still extremely fatigued, even more so than before treatment. Leak levels are very low. Flow limitation spikes occur throughout the night, but Sleep doc says they are meaningless, sleep doc is out of ideas and has no further help to offer. Consistently waking up exhausted, in a fog, with red glassy eyes, worsening memory problems, headaches. Adding Nuvigil makes slogging through the day bearable, but this is taking a toll. Any help would be greatly appreciated.

_________________
MaskHumidifier
Additional Comments: Complex Sleep Apnea. OSCAR v1.2.0 with Radical 7 pulse Co-oximeter. Non smoker/non drinker. Also using ResMed S9 Adapt SV

User avatar
Julie
Posts: 20056
Joined: Tue Feb 28, 2006 12:58 pm

Re: Still extremely fatigued on ASV, help?

Post by Julie » Thu Mar 18, 2021 7:56 am

They haven't been here since '16... I wouldn't hold my breath.