System One water usage/Dry mouth
-
larryb5061
- Posts: 2
- Joined: Tue Jan 26, 2016 10:27 am
System One water usage/Dry mouth
I can use my Respronics System One for 2 to 3 nights and the water level rarely changes. I use a dehumidifier in my house so I know its not a ambient moisture thing in the house. I have tried different settings on the cpap unit itself with no changes. I wake up some nights with a dry mouth. Any suggestions??
Re: System One water usage/Dry mouth
Please fill in your equipment to help us give you answers. wiki/index.php/Registering_Equipment_in_User_Profile Please remove the water and turn over your machine. on the bottom of the blower part is a REF# that identifies the machine. And this link might help you identify the machine. https://sleep.tnet.com/cmd
_________________
| Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
| Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear |
| Additional Comments: Sleepyhead Software |
ResMed Aircurve 10 VAUTO EPAP 11 IPAP 15 / P10 pillows mask / Sleepyhead Software / Back up & travel machine Respironics 760
-
Sleeprider
- Posts: 1562
- Joined: Tue May 06, 2008 5:57 pm
- Location: Murrysville, PA
Re: System One water usage/Dry mouth
Larry, turn off the System One humidification and use the manual settings. You will need to be in the clinical menu. http://www.apneaboard.com/dreamstation- ... structions
Once in setup, turn System One humidification off, then select 1-3 for humidity level. At setting 2, I will use the entire tank in 10-11 hours and need to refill daily. Get the heated hose if you don't have it, cause you're going to have rain-out if you make it this humid without heat or a cover.
Once in setup, turn System One humidification off, then select 1-3 for humidity level. At setting 2, I will use the entire tank in 10-11 hours and need to refill daily. Get the heated hose if you don't have it, cause you're going to have rain-out if you make it this humid without heat or a cover.
_________________
| Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
| Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear |
| Additional Comments: Sleepyhead software. Just changed from PRS1 BiPAP Auto DS760TS |
-
larryb5061
- Posts: 2
- Joined: Tue Jan 26, 2016 10:27 am
Re: System One water usage/Dry mouth
Thank you, however my machine does not look anything like that. I am at work at the time. will provide more information when I am home.Sleeprider wrote:Larry, turn off the System One humidification and use the manual settings. You will need to be in the clinical menu. http://www.apneaboard.com/dreamstation- ... structions
Once in setup, turn System One humidification off, then select 1-3 for humidity level. At setting 2, I will use the entire tank in 10-11 hours and need to refill daily. Get the heated hose if you don't have it, cause you're going to have rain-out if you make it this humid without heat or a cover.
-
Sleeprider
- Posts: 1562
- Joined: Tue May 06, 2008 5:57 pm
- Location: Murrysville, PA
Re: System One water usage/Dry mouth
My apologies...I posted the wrong link. Try this: http://www.apneaboard.com/pr-system-one ... structions
_________________
| Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
| Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear |
| Additional Comments: Sleepyhead software. Just changed from PRS1 BiPAP Auto DS760TS |
Re: System One water usage/Dry mouth
well, go back and fix it, to avoid confusing future generations.Sleeprider wrote:My apologies...I posted the wrong link. Try this: http://www.apneaboard.com/pr-system-one ... structions
Get OSCAR