Broken Heater Hose

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
User avatar
Leland
Posts: 25
Joined: Tue Jan 09, 2007 10:50 pm
Location: A bit east of San Francisco

Broken Heater Hose

Post by Leland » Sun Jan 10, 2016 11:58 am

I have a Respironics System One with a heated hose, but the heater in the hose no longer functions. I tried a new hose but the problem remains. Do any of the readers here know how to repair this unit?

User avatar
Pugsy
Posts: 65261
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Broken Heater Hose

Post by Pugsy » Sun Jan 10, 2016 12:32 pm

Are you saying that your machine will no longer recognize the heated hose when it is plugged in or it recognizes the heated hose and gives you the appropriate menu choices but those choices don't seem to be working?

There was a known problem with the 50 series machines and humidifiers not working because of a loose connection in the blower unit but I wasn't aware of that problem showing up in the 60 series machines.

What is the manufacture date of your machine? It's on a little sticker on the bottom of the blower unit.
Same sticker where the model number is...it's an 8 digit number in year/month/day format.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
Leland
Posts: 25
Joined: Tue Jan 09, 2007 10:50 pm
Location: A bit east of San Francisco

Re: Broken Heater Hose

Post by Leland » Fri Jan 29, 2016 12:41 pm

I took my unit into Apria yesterday. The problem was in the three electrical contacts at the outlet of the humidifier. The technician said that I could probably have repaired it myself by somehow cleaning the corrosion off the contacts. Instead, he replaced the entire humidifier, heated hose and main (external) powers supply. I was surprised that it was still covered by warranty since I got the machine a bit over three years ago. It works perfectly now. Kudos to Apria as I was met at the door by the technician before I even sat down. Getting the appointment only took four days. The entire repair/replacement took less than ten minutes.

User avatar
OkyDoky
Posts: 2869
Joined: Mon Aug 25, 2014 5:18 pm

Re: Broken Heater Hose

Post by OkyDoky » Fri Jan 29, 2016 1:22 pm

Leland wrote:I took my unit into Apria yesterday. The problem was in the three electrical contacts at the outlet of the humidifier. The technician said that I could probably have repaired it myself by somehow cleaning the corrosion off the contacts. Instead, he replaced the entire humidifier, heated hose and main (external) powers supply. I was surprised that it was still covered by warranty since I got the machine a bit over three years ago. It works perfectly now. Kudos to Apria as I was met at the door by the technician before I even sat down. Getting the appointment only took four days. The entire repair/replacement took less than ten minutes.
Do they charge you a continuing maintenance fee twice a year? Some of their contracts do. Even so it paid off this time.
ResMed Aircurve 10 VAUTO EPAP 11 IPAP 15 / P10 pillows mask / Sleepyhead Software / Back up & travel machine Respironics 760

JDS74
Posts: 3397
Joined: Sun Jan 23, 2011 2:57 pm
Location: South Carolina

Re: Broken Heater Hose

Post by JDS74 » Fri Jan 29, 2016 5:34 pm

I had one failure on a 50 series machine and PR replaced the blower and one failure on my 60 series machine and PR replaced that blower as well. I don't think it is very common but it does happen.

_________________
Mask: Oracle HC452 Oral CPAP Mask
Humidifier: DreamStation Heated Humidifier
Additional Comments: EverFlo Q 3.0 Liters O2 PR DSX900 ASV
Oracle 452 Lessons Learned Updated
DSX900 AutoSV with HC150 extra humidifier and Hibernite heated hose
Settings: EPAP Min-10.0, EPAP Max-17, PS Min-3, PS Max-10, Max Pressure-20, Rate-Auto, Biflex-1.
Sleepyhead and Encore Pro 2.21.

Mopheus
Posts: 75
Joined: Wed Dec 09, 2015 5:16 pm

Re: Broken Heater Hose

Post by Mopheus » Sat Jan 30, 2016 10:15 am

My hose stopped heating a while ago, I thought it must be the hose, perhaps its the unit. Is there any way to tell, other than getting another hose, or bringing it into the DME?

_________________
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Additional Comments: PR sytem one Aflex. 560 p

JDS74
Posts: 3397
Joined: Sun Jan 23, 2011 2:57 pm
Location: South Carolina

Re: Broken Heater Hose

Post by JDS74 » Sat Jan 30, 2016 1:56 pm

Change the humidifier setting to a high value. Put some water in the humidifier chamber and turn pre-heat on (Press the alarm silence button AND the control knob at the same time and hold for about 8 seconds - it will beep and the alarm silence button will light up blue). Then wait for about 20 minutes or so. Remove the humidifier tank and check the temperature of the heater plate on the bottom of the water chamber. If it is cold or only slightly warm, it is likely that the unit heater is not getting enough power from the blower unit and that will have to be repaired. If you unit is less than 2 years old or your DME has you on some sort of maintenance payment contract, take it back to your DME for repairs. If older than 2 years and no maintenance agreement, then contact AccBio for repairs. If you want to fix it yourself,
search for "Fix: Respironics System One (50 series) Humidifier Problems" for a detailed description of this problem on a Series 50 machine. I believe the Series 60 machines have a similar connection issue and may be able to be repaired in the same fashion. If you try this on your own, you may end up with a dead machine so be careful.

If it is hot or at least very warm, then the remaining check is to see if you are getting rainout during the night when the humidifier level is set to a high value. Rainout is a symptom that the hose is not heating properly even though the humidifier is heating. At this point, swapping out the heated hose is the best way to make sure.

_________________
Mask: Oracle HC452 Oral CPAP Mask
Humidifier: DreamStation Heated Humidifier
Additional Comments: EverFlo Q 3.0 Liters O2 PR DSX900 ASV
Oracle 452 Lessons Learned Updated
DSX900 AutoSV with HC150 extra humidifier and Hibernite heated hose
Settings: EPAP Min-10.0, EPAP Max-17, PS Min-3, PS Max-10, Max Pressure-20, Rate-Auto, Biflex-1.
Sleepyhead and Encore Pro 2.21.

Mopheus
Posts: 75
Joined: Wed Dec 09, 2015 5:16 pm

Re: Broken Heater Hose

Post by Mopheus » Sat Jan 30, 2016 2:34 pm

My water gets warm. I had rain out, so my wife made me hose cozy. It is most likely the hose, but It could also be the contacts on the machine like the OP problem. I should test the contacts with a multi meter, there should be something there to read?

_________________
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Additional Comments: PR sytem one Aflex. 560 p

Mopheus
Posts: 75
Joined: Wed Dec 09, 2015 5:16 pm

Re: Broken Heater Hose

Post by Mopheus » Sun Jan 31, 2016 8:05 pm

OK the hose is good, it is not getting juice from the unit. I have '50 unit. I will try for a replacement.

_________________
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Additional Comments: PR sytem one Aflex. 560 p

Sleeprider
Posts: 1562
Joined: Tue May 06, 2008 5:57 pm
Location: Murrysville, PA

Re: Broken Heater Hose

Post by Sleeprider » Sun Jan 31, 2016 8:47 pm

Mopheus wrote:OK the hose is good, it is not getting juice from the unit. I have '50 unit. I will try for a replacement.
The 50 series machines do not support a heated hose. That started with 60 series. You can get a heated hose that will work great, it just won't come from Philips Respironics unless you upgrade to a 60 series+ machine. Try SleepZone http://www.sleepzone.com.au/sleepzone-h ... ed-states/ It's actually a warmer better hose than the ones made by PR or Resmed.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Additional Comments: Sleepyhead software. Just changed from PRS1 BiPAP Auto DS760TS

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Broken Heater Hose

Post by palerider » Sun Jan 31, 2016 9:25 pm

Sleeprider wrote:
Mopheus wrote:OK the hose is good, it is not getting juice from the unit. I have '50 unit. I will try for a replacement.
The 50 series machines do not support a heated hose.
mopheus is wrong (or a typo) about what type of machine he has, based on the SH page he posted, he's got a 60 series.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

Mopheus
Posts: 75
Joined: Wed Dec 09, 2015 5:16 pm

Re: Broken Heater Hose

Post by Mopheus » Sun Jan 31, 2016 9:57 pm

You are correct, it must be a 60 because the heated hose used to heat, ha.
I removed the water, invert, and I saw garblegoogLe 50. So you are right, must be a 60 if i have good data.
I swear I didn't see the number 60 underneath. Filled up with water now, I will check again.

_________________
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Additional Comments: PR sytem one Aflex. 560 p

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Broken Heater Hose

Post by palerider » Sun Jan 31, 2016 10:38 pm

Mopheus wrote:You are correct, it must be a 60 because the heated hose used to heat, ha.
I removed the water, invert, and I saw garblegoogLe 50. So you are right, must be a 60 if i have good data.
I swear I didn't see the number 60 underneath. Filled up with water now, I will check again.
other than putting the right machine in your sig so the question doesn't keep coming up, there's no need to check.

but if you want to, look under the blower, not the humidifier, at where it says "REF 560" with possibly some other letters.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.