fault 66 on Z1 auto

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
mondray

fault 66 on Z1 auto

Post by mondray » Fri Dec 25, 2015 12:15 pm

I have a message on the Z1 screen this morning : fault 66. If i presse the center button for few seconds a 8.0 appear.
The machine doesn't work anymore.
Anyone can help me?

User avatar
Julie
Posts: 20051
Joined: Tue Feb 28, 2006 12:58 pm

Re: fault 66 on Z1 auto

Post by Julie » Fri Dec 25, 2015 12:21 pm

1-855-436-8729 Where to call for help... got that from a Google search of the machine manual, but 66 wasn't listed (diff. model yr?) tho' others were.

mondray

Re: fault 66 on Z1 auto

Post by mondray » Fri Dec 25, 2015 12:32 pm

No service on Christmas. Thanks for the number i will try tomorrow. A long night to come...

User avatar
wm_hess
Posts: 334
Joined: Sat Oct 16, 2010 2:39 am

Re: fault 66 on Z1 auto

Post by wm_hess » Fri Dec 25, 2015 12:36 pm

Mondray where are you located? If you registered and filled out your profile, someone might be able to provide a backup machine

-Bill

_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier
Mask: AirFit™ F10 Full Face Mask with Headgear
Additional Comments: CMS-50I Wrist Pulse Oximeter, SP02 Review, Sleepyhead

mondray
Posts: 1
Joined: Fri Dec 25, 2015 12:56 pm

Re: fault 66 on Z1 auto

Post by mondray » Fri Dec 25, 2015 12:59 pm

I am in Sutton quebec Canada.
I will have to go back in Montreal. I have a Resmed there.
But the would be an indication of what to do with this error message.
Thank you very much for your concern.

User avatar
TangledHose
Posts: 862
Joined: Wed Mar 13, 2013 11:13 am
Location: Centennial, CO

Re: fault 66 on Z1 auto

Post by TangledHose » Fri Dec 25, 2015 2:09 pm

mondray wrote:I have a message on the Z1 screen this morning : fault 66. If i presse the center button for few seconds a 8.0 appear.
The machine doesn't work anymore.
Anyone can help me?


If you were doing the firmware update......this was in the instructions for the update:

"The upgrade will take 1 minute and in the process you may see “Fault 66” on the Z1 screen. Ignore this message as it will clear when you power cycle after the upgrade is complete."

_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier
Mask: Simplus Full Face CPAP Mask with Headgear
Additional Comments: Also own F&P Icon Auto for backup and travel; and a Transcend when packing space is limited
Link to Sleepyhead:
http://sleepfiles.com/SH2/

bill-e
Posts: 152
Joined: Fri Dec 08, 2006 1:28 pm
Location: New Hampshire

Re: fault 66 on Z1 auto

Post by bill-e » Sun Dec 27, 2015 5:08 pm

TangledHose wrote:
mondray wrote:I have a message on the Z1 screen this morning : fault 66. If i presse the center button for few seconds a 8.0 appear.
The machine doesn't work anymore.
Anyone can help me?


If you were doing the firmware update......this was in the instructions for the update:

"The upgrade will take 1 minute and in the process you may see “Fault 66” on the Z1 screen. Ignore this message as it will clear when you power cycle after the upgrade is complete."
I got the "Fault 66" during the upgrade and it cleared as stated.

_________________
Mask: Eson™ Nasal CPAP Mask with Headgear
Additional Comments: Human Design Medical Z1 Auto with HME's for travel, setting 14-20 no ramp