Humidifier doesn't last the whole night

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
WindCpap
Posts: 394
Joined: Fri Oct 09, 2015 6:57 pm

Re: Humidifier doesn't last the whole night

Post by WindCpap » Mon Dec 14, 2015 4:18 pm

Wulfman... wrote:
WindCpap wrote:The vent rate is not proportional to pressure. More like vent rate squared is proportional to pressure.
NO! WRONG!

Page 7 of the following manual (for your mask).

http://www.resmed.com/us/dam/documents/ ... er_eng.pdf


Den

.

Look at the curve. It is not a straight line. Proportional means that the relationship is linear, and therefore a straight line.

_________________
MachineMask

WindCpap
Posts: 394
Joined: Fri Oct 09, 2015 6:57 pm

Re: Humidifier doesn't last the whole night

Post by WindCpap » Mon Dec 14, 2015 4:20 pm

OkyDoky wrote:I used propotional as the # 3 definition at dictionary.com
3. of, relating to, or based on proportion; relative.
Every mask has a venting pressure curve flow that shows how it increases.
Here is the P10
Image
Even by this definition, a proportional curve is a straight line. If the curve you show was proportional, the flow rate at 20 cmH20 would be 100 lpm. I am not arguing with you, only correcting an inaccuracy.

Pressure Flow Proportional Prediction More Accurate Prediction based on pressure being roughly proportional to square of flow.
4 20 20 20
8 29 40 28.28427125
12 37 60 34.64101615
16 43 80 40
20 49 100 44.72135955

_________________
MachineMask
Last edited by WindCpap on Mon Dec 14, 2015 4:29 pm, edited 1 time in total.

User avatar
OkyDoky
Posts: 2869
Joined: Mon Aug 25, 2014 5:18 pm

Re: Humidifier doesn't last the whole night

Post by OkyDoky » Mon Dec 14, 2015 4:29 pm

The increases are relative not a mathmatical constant.
ResMed Aircurve 10 VAUTO EPAP 11 IPAP 15 / P10 pillows mask / Sleepyhead Software / Back up & travel machine Respironics 760

WindCpap
Posts: 394
Joined: Fri Oct 09, 2015 6:57 pm

Re: Humidifier doesn't last the whole night

Post by WindCpap » Mon Dec 14, 2015 4:30 pm

The definition of proportional is that the relationship is defined by a mathematical constant if you are talking about numbers.

http://dictionary.reference.com/browse/proportional

Proportional can only be taken as a more loose relationship if you are talking about non-quantifiable things.

_________________
MachineMask

User avatar
Wulfman...
Posts: 6692
Joined: Mon Sep 01, 2014 6:41 pm
Location: Nearest fishing spot

Re: Humidifier doesn't last the whole night

Post by Wulfman... » Mon Dec 14, 2015 4:35 pm

WindCpap wrote:
Wulfman... wrote:
WindCpap wrote:The vent rate is not proportional to pressure. More like vent rate squared is proportional to pressure.
NO! WRONG!

Page 7 of the following manual (for your mask).

http://www.resmed.com/us/dam/documents/ ... er_eng.pdf


Den

.

Look at the curve. It is not a straight line. Proportional means that the relationship is linear, and therefore a straight line.
Doesn't look like much of a "curve" to me. It's "straighter" than "curvier". And, possibly at the higher pressures, there's a "Law of Diminishing Returns" coming into play based on the amount of air that can exit through the vents.


Den

.
(5) REMstar Autos w/C-Flex & (6) REMstar Pro 2 CPAPs w/C-Flex - Pressure Setting = 14 cm.
"Passover" Humidification - ResMed Ultra Mirage FF - Encore Pro w/Card Reader & MyEncore software - Chiroflow pillow
User since 05/14/05

WindCpap
Posts: 394
Joined: Fri Oct 09, 2015 6:57 pm

Re: Humidifier doesn't last the whole night

Post by WindCpap » Mon Dec 14, 2015 5:13 pm

Wulfman... wrote:
WindCpap wrote:
Wulfman... wrote:
WindCpap wrote:The vent rate is not proportional to pressure. More like vent rate squared is proportional to pressure.
NO! WRONG!

Page 7 of the following manual (for your mask).

http://www.resmed.com/us/dam/documents/ ... er_eng.pdf


Den

.

Look at the curve. It is not a straight line. Proportional means that the relationship is linear, and therefore a straight line.
Doesn't look like much of a "curve" to me. It's "straighter" than "curvier". And, possibly at the higher pressures, there's a "Law of Diminishing Returns" coming into play based on the amount of air that can exit through the vents.


Den

.

Got to love how you said I was wrong, I proved with numbers that I am right, and you still say I am wrong. "Law of Diminishing Returns" has absolutely nothing to do with fluid mechanics. At air speeds less than 1/3 the speed of sound, the square of the flow rate is roughly proportional to the pressure difference through most openings. "Straighter" than "curvier"...really.

_________________
MachineMask

User avatar
OkyDoky
Posts: 2869
Joined: Mon Aug 25, 2014 5:18 pm

Re: Humidifier doesn't last the whole night

Post by OkyDoky » Mon Dec 14, 2015 5:27 pm

Would it have made you happier if I had used the word relational instead of proportional?
ResMed Aircurve 10 VAUTO EPAP 11 IPAP 15 / P10 pillows mask / Sleepyhead Software / Back up & travel machine Respironics 760

User avatar
ChicagoGranny
Posts: 15534
Joined: Sun Jan 29, 2012 1:43 pm
Location: USA

Re: Humidifier doesn't last the whole night

Post by ChicagoGranny » Mon Dec 14, 2015 5:42 pm

WindCpap wrote:The vent rate is not proportional to pressure. More like vent rate squared is proportional to pressure.
I agree with your statement. Over the range in the example, the function is far from a straight line. A simple function that best fits is,

Y = 9.1906(X^0.5577)
correlation = 0.9998

where X = CPAP pressure in cm H2O, and Y = mask design vent flow in lpm.

Someone check my math.
Last edited by ChicagoGranny on Mon Dec 14, 2015 5:53 pm, edited 1 time in total.
"It's not the number of breaths we take, it's the number of moments that take our breath away."

Cuando cuentes cuentos, cuenta cuántas cuentos cuentas.

User avatar
MichaelWD64
Posts: 64
Joined: Sat Dec 12, 2015 6:40 pm

Re: Humidifier doesn't last the whole night

Post by MichaelWD64 » Mon Dec 14, 2015 5:49 pm

Thank you, all!

A lot to think about and try here.


MWD

_________________
Mask: Nuance & Nuance Pro Nasal Pillow CPAP Mask with Gel Nasal Pillows
Additional Comments: Machine: Resmed Aircurve 10 ASVauto bilevel with heated humidifier
Got on CPAP 2015, AHI 147. Much better now. Have central & obstructive apnea. Smoked 30 years, quit when got on O2 & CPAP. I'm not on O2 during day anymore, only when I sleep, into CPAP machine. Ahhhh....oxygen! Refreshing!

User avatar
Wulfman...
Posts: 6692
Joined: Mon Sep 01, 2014 6:41 pm
Location: Nearest fishing spot

Re: Humidifier doesn't last the whole night

Post by Wulfman... » Mon Dec 14, 2015 5:50 pm

WindCpap wrote:
Wulfman... wrote:
WindCpap wrote:
Wulfman... wrote:
WindCpap wrote:The vent rate is not proportional to pressure. More like vent rate squared is proportional to pressure.
NO! WRONG!

Page 7 of the following manual (for your mask).

http://www.resmed.com/us/dam/documents/ ... er_eng.pdf


Den

.

Look at the curve. It is not a straight line. Proportional means that the relationship is linear, and therefore a straight line.
Doesn't look like much of a "curve" to me. It's "straighter" than "curvier". And, possibly at the higher pressures, there's a "Law of Diminishing Returns" coming into play based on the amount of air that can exit through the vents.


Den

.

Got to love how you said I was wrong, I proved with numbers that I am right, and you still say I am wrong. "Law of Diminishing Returns" has absolutely nothing to do with fluid mechanics. At air speeds less than 1/3 the speed of sound, the square of the flow rate is roughly proportional to the pressure difference through most openings. "Straighter" than "curvier"...really.
Yep. I'm sure that if you made the vent holes bigger you could get more airflow going out of them at higher pressures and that line would flatten out even more.
But, as the pressures get higher, there's only about so much that can exit the mask vent.
Besides, those are just some "ideal" numbers from their labratories to compare "reality" (mask LEAKAGE) to.
Having looked at different mask "vent flow rate" charts, there really isn't much difference in them across the board.
If I'm using a pressure of 12 or 13 and my sleep reports show "about 40", that's close enough for me. If I'm a little over or under them.......oh, well.......each night is different.


Den

.
(5) REMstar Autos w/C-Flex & (6) REMstar Pro 2 CPAPs w/C-Flex - Pressure Setting = 14 cm.
"Passover" Humidification - ResMed Ultra Mirage FF - Encore Pro w/Card Reader & MyEncore software - Chiroflow pillow
User since 05/14/05

User avatar
Wulfman...
Posts: 6692
Joined: Mon Sep 01, 2014 6:41 pm
Location: Nearest fishing spot

Re: Humidifier doesn't last the whole night

Post by Wulfman... » Mon Dec 14, 2015 5:56 pm

And, if a person is using a range of pressures it can be a real challenge......determining the average pressures and how much actual "leakage" is over and above the "vent flow rate".
After a certain number of years of use and monitoring, you just look at the numbers and go "that's close enough" (or not).


Den

.
(5) REMstar Autos w/C-Flex & (6) REMstar Pro 2 CPAPs w/C-Flex - Pressure Setting = 14 cm.
"Passover" Humidification - ResMed Ultra Mirage FF - Encore Pro w/Card Reader & MyEncore software - Chiroflow pillow
User since 05/14/05

User avatar
ChicagoGranny
Posts: 15534
Joined: Sun Jan 29, 2012 1:43 pm
Location: USA

Re: Humidifier doesn't last the whole night

Post by ChicagoGranny » Mon Dec 14, 2015 5:58 pm

Wulfman... wrote:Besides, those are just some "ideal" numbers from their labratories to compare "reality" (mask LEAKAGE) to.
You do understand the chart is about design vent rates? Your comment indicates you don't - those are not "ideal" vent rates. They are design vent rates.
Wulfman... wrote:Having looked at different mask "vent flow rate" charts, there really isn't much difference in them across the board.
Well, yes, that is because the manufacturers designed the vents that way.
Wulfman... wrote:If I'm using a pressure of 12 or 13 and my sleep reports show "about 40", that's close enough for me. If I'm a little over or under them.......oh, well.......each night is different.
No, each night is not different - the same pressures will yield the same vent rate on any night.

You are confusing vent rate with total leak which is vent rate plus unintentional leak.
"It's not the number of breaths we take, it's the number of moments that take our breath away."

Cuando cuentes cuentos, cuenta cuántas cuentos cuentas.

User avatar
Wulfman...
Posts: 6692
Joined: Mon Sep 01, 2014 6:41 pm
Location: Nearest fishing spot

Re: Humidifier doesn't last the whole night

Post by Wulfman... » Mon Dec 14, 2015 6:06 pm

ChicagoGranny wrote:
Wulfman... wrote:Besides, those are just some "ideal" numbers from their labratories to compare "reality" (mask LEAKAGE) to.
You do understand the chart is about design vent rates? Your comment indicates you don't - those are not "ideal" vent rates. They are design vent rates.
Wulfman... wrote:Having looked at different mask "vent flow rate" charts, there really isn't much difference in them across the board.
Well, yes, that is because the manufacturers designed the vents that way.
Wulfman... wrote:If I'm using a pressure of 12 or 13 and my sleep reports show "about 40", that's close enough for me. If I'm a little over or under them.......oh, well.......each night is different.
No, each night is not different - the same pressures will yield the same vent rate on any night.

You are confusing vent rate with total leak which is vent rate plus unintentional leak.
Right. And, I'm trying to interject some "reality" here. Nobody sleeps the same night after night and the "leak rates" from the reports are going to vary......sometimes greatly. Of course these "vent flow rates" are their "design rates" for us to be able to judge (if we have data-capable machines and use software to monitor our therapy) if we have excessive leaking or not.

Edit:
And, this subject becomes even more convoluted when using ResMed machines which deduct the anticipated vent flow rate based on the mask type entered in the setup menu.


Den

.
(5) REMstar Autos w/C-Flex & (6) REMstar Pro 2 CPAPs w/C-Flex - Pressure Setting = 14 cm.
"Passover" Humidification - ResMed Ultra Mirage FF - Encore Pro w/Card Reader & MyEncore software - Chiroflow pillow
User since 05/14/05

User avatar
archangle
Posts: 9293
Joined: Sun Mar 27, 2011 11:55 am

Re: Humidifier doesn't last the whole night

Post by archangle » Mon Dec 14, 2015 6:55 pm

Proportional: y=mx
Linear: y=mx+b

That's the mathematical definition. Some people may misuse the term or define it poorly in colloquial usage.

The graph is not proportional. The flow at 8 cmH2O is not 2x the flow at 4 cm. The graph is fairly linear, but not quite. I'm tempted to try to fit a curve to the data, but since the airflow has to drop to zero at zero pressure, I think I don't have enough data points at low pressure to mean much.

I would have thought the flow would be proportional to pressure, but apparently not. The flow increases more slowly than proportionally. Apparently air through an orifice is not proportional to pressure. This page is a bit interesting for airflow through an oriface. http://www.aircompressorworks.com/airfl ... fices.html

I'm guessing it's a question of turbulent airflow being nonlinear. I guess I gots myself edumacated today.

It would be neat if it were possible to have the flow rate be constant with pressure, because that would waste less air and water at higher pressures, but I don't see a reasonable way to do that in a mask without having something like a pressure setting on the mask itself.

_________________
Mask: Swift™ FX Nasal Pillow CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: Also SleepyHead, PRS1 Auto, Respironics Auto M series, Legacy Auto, and Legacy Plus
Please enter your equipment in your profile so we can help you.
Click here for information on the most common alternative to CPAP.
If it's midnight and a DME tells you it's dark outside, go and check for yourself.

Useful Links.

WindCpap
Posts: 394
Joined: Fri Oct 09, 2015 6:57 pm

Re: Humidifier doesn't last the whole night

Post by WindCpap » Mon Dec 14, 2015 7:19 pm

ChicagoGranny wrote:
WindCpap wrote:The vent rate is not proportional to pressure. More like vent rate squared is proportional to pressure.
I agree with your statement. Over the range in the example, the function is far from a straight line. A simple function that best fits is,

Y = 9.1906(X^0.5577)
correlation = 0.9998

where X = CPAP pressure in cm H2O, and Y = mask design vent flow in lpm.

Someone check my math.
In other words, the square of the flow is roughly proportional to the pressure (or the square root of the pressure is roughly proportional to the flow).

Out of curiosity, what did you do to get the correlation?

_________________
MachineMask