Pugsy - need help with SleepyHead/TNET/imgur

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
User avatar
andycole
Posts: 149
Joined: Thu Nov 19, 2015 11:19 am
Location: Holyoke, MA

Re: Pugsy - need help with SleepyHead

Post by andycole » Thu Dec 03, 2015 1:18 pm

have appt at sleep medicine 12/17 and am trying to get graphs and such ready for them although they have access to ResMed. We have to solve this leak issue even if it means trying total-face mask and getting them to send order. DME (LinCare) handles respironics who appear to be the only total-face supplier. After the AirFit F10, the Amara View was total waste of time (leaked the worst) and now trying Quattro which seems to me to be identical to my F10 except for headgear. Did get chinstrap but RT didn't like it, so haven't tried it.
Computer people just called and they need power supply so have to sign off now.

_________________
Mask: ResMed AirFit™ F20 Full Face CPAP Mask with Headgear
Additional Comments: Machine: ResMed AirCurve 10 ASV with Heated Humidifier/Slimline Hose. Pressures: 5/10

User avatar
archangle
Posts: 9293
Joined: Sun Mar 27, 2011 11:55 am

Re: Pugsy - need help with SleepyHead

Post by archangle » Thu Dec 03, 2015 5:00 pm

andycole wrote:Archangle - Lenovo P500 with win81.
The manual wasn't very clear, but I suspect that's one of the ones where you have to press the "Fn" key before you hit the F1-F12 keys in order to the the function keys instead of the things like display brightness and volume controls.

_________________
Mask: Swift™ FX Nasal Pillow CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: Also SleepyHead, PRS1 Auto, Respironics Auto M series, Legacy Auto, and Legacy Plus
Please enter your equipment in your profile so we can help you.
Click here for information on the most common alternative to CPAP.
If it's midnight and a DME tells you it's dark outside, go and check for yourself.

Useful Links.

User avatar
andycole
Posts: 149
Joined: Thu Nov 19, 2015 11:19 am
Location: Holyoke, MA

Re: Pugsy - need help with SleepyHead

Post by andycole » Thu Dec 03, 2015 6:28 pm

Archangle - turns out this is a known problem and solution is to reinstall older driver. Printef off the instructions and dropped at shop, Would have expecte to hear from them by now.

_________________
Mask: ResMed AirFit™ F20 Full Face CPAP Mask with Headgear
Additional Comments: Machine: ResMed AirCurve 10 ASV with Heated Humidifier/Slimline Hose. Pressures: 5/10

User avatar
andycole
Posts: 149
Joined: Thu Nov 19, 2015 11:19 am
Location: Holyoke, MA

Re: Pugsy - need help with SleepyHead

Post by andycole » Thu Dec 03, 2015 7:03 pm

Pugsy - is this correct. imgur needs to update their documentation as it doesn't match the screens.

Image

_________________
Mask: ResMed AirFit™ F20 Full Face CPAP Mask with Headgear
Additional Comments: Machine: ResMed AirCurve 10 ASV with Heated Humidifier/Slimline Hose. Pressures: 5/10

User avatar
andycole
Posts: 149
Joined: Thu Nov 19, 2015 11:19 am
Location: Holyoke, MA

Re: Pugsy - need help with SleepyHead

Post by andycole » Thu Dec 03, 2015 7:29 pm

actually, it is TNET who needs to update their screenshots of imgur. I contacted them. On page 3 the screenshot for "getting the code" does not match so the load I did may not be too useful.

_________________
Mask: ResMed AirFit™ F20 Full Face CPAP Mask with Headgear
Additional Comments: Machine: ResMed AirCurve 10 ASV with Heated Humidifier/Slimline Hose. Pressures: 5/10

User avatar
OkyDoky
Posts: 2869
Joined: Mon Aug 25, 2014 5:18 pm

Re: Pugsy - need help with SleepyHead/TNET/imgur

Post by OkyDoky » Thu Dec 03, 2015 7:43 pm

Imgur changed where you have to log in to get the options. They were available before on the initial page but not now. Here is a thread where I was trying to figure why I didn't have the screen that I once had. viewtopic.php?f=1&t=108815&st=0&sk=t&sd=a
I added the link that I forgot to post.
Last edited by OkyDoky on Thu Dec 03, 2015 10:36 pm, edited 1 time in total.
ResMed Aircurve 10 VAUTO EPAP 11 IPAP 15 / P10 pillows mask / Sleepyhead Software / Back up & travel machine Respironics 760

User avatar
andycole
Posts: 149
Joined: Thu Nov 19, 2015 11:19 am
Location: Holyoke, MA

Re: Pugsy - need help with SleepyHead/TNET/imgur

Post by andycole » Thu Dec 03, 2015 8:22 pm

tnx OkyDoky - I'm not sure that what is posted in this thread is what Pugsy wants. We're looking right now at my leak rate issue. My prior posting "mask keeps popping into my mouth' could be the whole thing and I have appt with RT to review AGAIN. This time I want supporting info!!!!

_________________
Mask: ResMed AirFit™ F20 Full Face CPAP Mask with Headgear
Additional Comments: Machine: ResMed AirCurve 10 ASV with Heated Humidifier/Slimline Hose. Pressures: 5/10

User avatar
OkyDoky
Posts: 2869
Joined: Mon Aug 25, 2014 5:18 pm

Re: Pugsy - need help with SleepyHead/TNET/imgur

Post by OkyDoky » Thu Dec 03, 2015 8:31 pm

I was just commenting on your mentioned discrepancy from TNET and Imgur.
ResMed Aircurve 10 VAUTO EPAP 11 IPAP 15 / P10 pillows mask / Sleepyhead Software / Back up & travel machine Respironics 760

User avatar
Pugsy
Posts: 65412
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Pugsy - need help with SleepyHead/TNET/imgur

Post by Pugsy » Thu Dec 03, 2015 8:32 pm

The leaks are far from horrible on this report. It's easy to the the big spikes which probably is related to major mask movement..maybe even re-positioning. The big leak right at the end of the night I can't see that one due to the fact you have highlighted part of the night (that's the light and dark areas you see on the events graph) so the 4:00 leak isn't showing on the leak graph.
If you notice your graphs below the events graph the times don't match up with the times on the events graph and that's because you have highlighted part of the night instead of showing all the night.

Are these leaks waking you up?

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
andycole
Posts: 149
Joined: Thu Nov 19, 2015 11:19 am
Location: Holyoke, MA

Re: Pugsy - need help with SleepyHead/TNET/imgur

Post by andycole » Thu Dec 03, 2015 8:39 pm

OkyDokey -I have contacted both imgur and TNET about the documentation issue. Got a message from imgur that this has been brought up in their forum and they agree TNET needs to update. BUT they did not explain how to post so I took a shot at it and that is my graph from today but don't know if Pugsy will be able to see everything.

Pugsy, glad to see you got something. I didn't do any highlighting, just gave the download a try. Will try again tomorrow. I think there is a major leak right at the end which wakes me up. Also just realized the left side looks like yesterday while the right side should be today. OOPS-I'm dumb - date is from start of treatment which was last night - SORRY.

_________________
Mask: ResMed AirFit™ F20 Full Face CPAP Mask with Headgear
Additional Comments: Machine: ResMed AirCurve 10 ASV with Heated Humidifier/Slimline Hose. Pressures: 5/10

User avatar
Pugsy
Posts: 65412
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Pugsy - need help with SleepyHead/TNET/imgur

Post by Pugsy » Thu Dec 03, 2015 8:46 pm

You did the highlighting my accident. Probably accidentally clicked on a space on the graph somewhere.
Not a bit deal. You will get the hang of it soon enough.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
andycole
Posts: 149
Joined: Thu Nov 19, 2015 11:19 am
Location: Holyoke, MA

Re: Pugsy - need help with SleepyHead/TNET/imgur

Post by andycole » Thu Dec 03, 2015 9:10 pm

think I haven't worked so hard since I was involuntarily retired in May. Between the laptop, this new desktop (almost ready for hard usage), SH, TNET, and imgur, I'm pooped. Time for beddy bye and try the Quattro again. See you tomorrow - have a good night, and thanks for being so helpful.

_________________
Mask: ResMed AirFit™ F20 Full Face CPAP Mask with Headgear
Additional Comments: Machine: ResMed AirCurve 10 ASV with Heated Humidifier/Slimline Hose. Pressures: 5/10

User avatar
andycole
Posts: 149
Joined: Thu Nov 19, 2015 11:19 am
Location: Holyoke, MA

Re: Pugsy - need help with SleepyHead/TNET/imgur

Post by andycole » Thu Dec 03, 2015 9:34 pm

Pugsy - tired but couldn't wait.

Image

_________________
Mask: ResMed AirFit™ F20 Full Face CPAP Mask with Headgear
Additional Comments: Machine: ResMed AirCurve 10 ASV with Heated Humidifier/Slimline Hose. Pressures: 5/10

User avatar
andycole
Posts: 149
Joined: Thu Nov 19, 2015 11:19 am
Location: Holyoke, MA

Re: Pugsy - need help with SleepyHead/TNET/imgur

Post by andycole » Fri Dec 04, 2015 10:29 am

Pugsy - there has to be something wrong with this. I DISTINCTLY remember waking up several times last night. BTW: the imgur piece seems easier than originally shown in the TNET document. Are you able to see everything clearly?

Image

_________________
Mask: ResMed AirFit™ F20 Full Face CPAP Mask with Headgear
Additional Comments: Machine: ResMed AirCurve 10 ASV with Heated Humidifier/Slimline Hose. Pressures: 5/10

User avatar
OkyDoky
Posts: 2869
Joined: Mon Aug 25, 2014 5:18 pm

Re: Pugsy - need help with SleepyHead/TNET/imgur

Post by OkyDoky » Fri Dec 04, 2015 10:52 am

Andy, TNET is maintained by one of our members Krevlin on a volunteer basis. He has provided a much needed resource. What happened is that Imgur changed on us.
Your posting can be seen as is by opening it in another window. The large thumbnail that we liked now has to be accessed this way:

You have to sign in and then click on your ID in the upper right corner. It will then give you options and you click on images. This will take you to your images. Then you click on the image you want and it will open up the image with the full menu. At the bottom of this menu will be the Linked BB Code and the sizes. You choose the Large Thumbnail then copy the BB Code to your post.

The difference is that in order to get this menu you now have to open a free account and sign in when before it was on the first page.

This is the thread link where I was trying to clear this up for myself. viewtopic.php?f=1&t=108815&st=0&sk=t&sd=a
ResMed Aircurve 10 VAUTO EPAP 11 IPAP 15 / P10 pillows mask / Sleepyhead Software / Back up & travel machine Respironics 760