Sleepyhead data showing summary only

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
dazza
Posts: 33
Joined: Mon Feb 23, 2015 2:47 am

Sleepyhead data showing summary only

Post by dazza » Sat Nov 14, 2015 4:30 am

Hi , for the last week i cant seem to get any graphs ect on my sleepyhead data,it is only showing summary only ? There is no info for leaks ,pressure ect.The only data its showing is AHI and time usage.
Im using resmed 10 auto,could it have anything to do with the sd card getting full.Their is about 262 days of data on it.

cheers Dazza

_________________
MachineMask

User avatar
Pugsy
Posts: 65276
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Sleepyhead data showing summary only

Post by Pugsy » Sat Nov 14, 2015 7:12 am

Is the SD card correctly inserted during the entire night?
Are you turning the machine totally off right when you first stop the therapy upon waking up...like unplugging from power source (not the thing to do because the machine needs a little bit of time to close out the files it is writing to the SD card)?

It has nothing to do with the amount of data on the SD card.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
Last edited by Pugsy on Sat Nov 14, 2015 8:43 am, edited 1 time in total.
I may have to RISE but I refuse to SHINE.

dazza
Posts: 33
Joined: Mon Feb 23, 2015 2:47 am

Re: Sleepyhead data showing summary only

Post by dazza » Sat Nov 14, 2015 8:10 am

Pugsy wrote:Is the SD card correctly inserted during the entire night?
Are you turning the machine totally off right when you first stop the therapy upon waking up...like unplugging from power source?

It has nothing to do with the amount of data on the SD card.
Thanks for your input Pugsy,,,the card reader was the problem so all good now thanks.

_________________
MachineMask