Using Respironics asv bipap lost weight, apena worse
Re: Using Respironics asv bipap lost weight, apena worse
I am trying to download skeepyhead. It is telling me that my computer could be harmed. Should I continue?
Re: Using Respironics asv bipap lost weight, apena worse
That's just the standard Windows warning to confirm that "do you know where this software has come from, if you don't it could hurt your computer"
Of course we know where SleepyHead is coming and it's totally safe.
Of course we know where SleepyHead is coming and it's totally safe.
_________________
| Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
| Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/ |
I may have to RISE but I refuse to SHINE.
Re: Using Respironics asv bipap lost weight, apena worse
Also wanted to ask about encore. Is this ok to download? or would I have problem?
Re: Using Respironics asv bipap lost weight, apena worse
Thx. I will try those settings tonight. I need some sleep. Will let you know what happened.
Re: Using Respironics asv bipap lost weight, apena worse
I tried to download sleepyhead but I am having a problem. When my son comes in, I will ask him for help so I can post some data.
Re: Using Respironics asv bipap lost weight, apena worse
My son did download sleepyhead. I have to figure out how to send it. In the meanwhile am not seeing many Oa . But I am seeing a lot of ca, h, and vs. I am really confused. I thought I would be showing more Oa.
Re: Using Respironics asv bipap lost weight, apena worse
These links will help you to learn how to post your shots. They give you pictures and guide you through the steps.
https://sleep.tnet.com/resources/sleepyhead/shorganize
https://sleep.tnet.com/resources/sleepy ... screenshot
https://sleep.tnet.com/reference/tips/imgur
https://sleep.tnet.com/resources/sleepyhead/shorganize
https://sleep.tnet.com/resources/sleepy ... screenshot
https://sleep.tnet.com/reference/tips/imgur
_________________
| Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
| Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear |
| Additional Comments: Sleepyhead Software |
ResMed Aircurve 10 VAUTO EPAP 11 IPAP 15 / P10 pillows mask / Sleepyhead Software / Back up & travel machine Respironics 760
Re: Using Respironics asv bipap lost weight, apena worse
Last night with those settings that I was asked to put in, it was terrible. All the areas went out of wack. Until I am able to send you a printout, here are some data from sleepy time last night. Should I set the machine back to the original settings?
Feeling worn out. And I have so much to do today.
Crs: 1.67%
Leak: 0%
Ca : 2.71
Oa: 2.30
H: 5.42
Fl: 1.08
Vs: 0.27
Timed breath: 555.34
Feeling worn out. And I have so much to do today.
Crs: 1.67%
Leak: 0%
Ca : 2.71
Oa: 2.30
H: 5.42
Fl: 1.08
Vs: 0.27
Timed breath: 555.34
Re: Using Respironics asv bipap lost weight, apena worse
How was the aerophagia last night?
Which settings did you use last night...what you reported or what JDS74 suggested?
His EPAP minimum is likely too low since your AHI has climbed with the reduction in EPAP minimum.
If you used your original 8.5 min EPAP last night...try this tonight
EPAP min 9.0
and change PS minimum from 0 to 3
leave everything else the same.
Which settings did you use last night...what you reported or what JDS74 suggested?
His EPAP minimum is likely too low since your AHI has climbed with the reduction in EPAP minimum.
If you used your original 8.5 min EPAP last night...try this tonight
EPAP min 9.0
and change PS minimum from 0 to 3
leave everything else the same.
_________________
| Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
| Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/ |
I may have to RISE but I refuse to SHINE.
Re: Using Respironics asv bipap lost weight, apena worse
I used jds74 settings. The aerophagia was lessened. . Epap min was set to 5. So should I put epap min to 8.5 or 9
Re: Using Respironics asv bipap lost weight, apena worse
Okay...let's try his settings but let's not go to min EPAP of 9 just yet.Boop wrote:I used jds74 settings. The aerophagia was lessened. . Epap min was set to 5. So should I put epap min to 8.5 or 9
How about going from 5.0 min to 6.5 min and let's see what happens with the AHI as well as the aerophagia.
Use his other setting ideas...so just change EPAP min to 6.5.
Let's see if we can find a happy medium between less aerophagia and still deal with the obstructive stuff.
Adding the PS min to go from 0 to 3 or 3.5 means that IPAP can help us out a little bit.
With PS of 0 you weren't getting the full benefit of bilevel both in terms of comfort and therapy.
_________________
| Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
| Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/ |
I may have to RISE but I refuse to SHINE.
Re: Using Respironics asv bipap lost weight, apena worse

Last night data
Re: Using Respironics asv bipap lost weight, apena worse
OK the aerophagia is better. As Pugsy suggests, moving the EPAP up just a little to 6 or 6.5 will help the OSA events without pushing the aerophagia too much. Keeping the PSMin at 3 gives the machine some help in controlling the rest.
The CAs are treated by the machine and al long as the number doesn't get too big is not a worry right now.
The goal is to get the OAs and aerophagia under control so your rest improves then tweaking can be done to control the rest.
Remember, your machine can deal with your events while you are sleeping.
The CAs are treated by the machine and al long as the number doesn't get too big is not a worry right now.
The goal is to get the OAs and aerophagia under control so your rest improves then tweaking can be done to control the rest.
Remember, your machine can deal with your events while you are sleeping.
_________________
| Mask: Oracle HC452 Oral CPAP Mask |
| Humidifier: DreamStation Heated Humidifier |
| Additional Comments: EverFlo Q 3.0 Liters O2 PR DSX900 ASV |
Oracle 452 Lessons Learned Updated
DSX900 AutoSV with HC150 extra humidifier and Hibernite heated hose
Settings: EPAP Min-10.0, EPAP Max-17, PS Min-3, PS Max-10, Max Pressure-20, Rate-Auto, Biflex-1.
Sleepyhead and Encore Pro 2.21.
DSX900 AutoSV with HC150 extra humidifier and Hibernite heated hose
Settings: EPAP Min-10.0, EPAP Max-17, PS Min-3, PS Max-10, Max Pressure-20, Rate-Auto, Biflex-1.
Sleepyhead and Encore Pro 2.21.
Re: Using Respironics asv bipap lost weight, apena worse
What sleeping position were you in the first part of the night up to around 1 AM? Do you know if you were maybe on your back during that time?
_________________
| Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
| Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/ |
I may have to RISE but I refuse to SHINE.