Switch to Airsense 10 Autoset -- Mask Leaks!

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
Tallen234
Posts: 58
Joined: Thu Jul 17, 2008 5:05 pm

Switch to Airsense 10 Autoset -- Mask Leaks!

Post by Tallen234 » Wed Oct 21, 2015 12:31 pm

My Resmed S9 Autoset finally kicked the bucket, so I "upgraded" to an Airsense 10 Autoset. Using the same mask as before (F&P 431), I am now experiencing airleaks. Even when doing to Mask Fit setting, I can't get a tight seal. Even if I hold the mask to my face to make the tightest seal I can, it still gives me the frowny face. With my S9 and the same mask, I would get the occasionally leak, but most mornings I would wake up to a smiley face. Any ideas?

_________________
Mask

PoolQ
Posts: 498
Joined: Sat Jul 11, 2015 10:41 am
Location: Silicon Valley

Re: Switch to Airsense 10 Autoset -- Mask Leaks!

Post by PoolQ » Wed Oct 21, 2015 12:46 pm

just a thought, is the mask type set correctly? Are you actually feeling/hearing leaks or does the machine just say there is a leak? Depending on your answer, could it be a hose connection leak?
Sleeping MUCH better now

Physician
Posts: 693
Joined: Tue Jun 29, 2010 5:42 pm
Location: West Coast USA

Re: Switch to Airsense 10 Autoset -- Mask Leaks!

Post by Physician » Wed Oct 21, 2015 2:53 pm

Tallen234 wrote:My Resmed S9 Autoset finally kicked the bucket, so I "upgraded" to an Airsense 10 Autoset. Using the same mask as before (F&P 431), I am now experiencing airleaks. Even when doing to Mask Fit setting, I can't get a tight seal. Even if I hold the mask to my face to make the tightest seal I can, it still gives me the frowny face. With my S9 and the same mask, I would get the occasionally leak, but most mornings I would wake up to a smiley face. Any ideas?
Frowning face, if correct, is a large leak. What was the actual leak in L/min. ?

Are your settings the same as before? EPR the same ? Have you determined from where the leak is arising ?

A bit off topic:

What in the S9 died ?

Why didn't you get it repaired ?

SewTired
Posts: 1736
Joined: Thu Apr 16, 2015 8:33 am
Location: Minneapolis area

Re: Switch to Airsense 10 Autoset -- Mask Leaks!

Post by SewTired » Thu Oct 22, 2015 12:39 pm

My brother got a new bipap after 7 years and even though the pressure settings are the same, he gets noticeably more air from the new bipap than the old. He was able to actually cut his settings back from a top rate by 2 settings. Could be that your unit was failing over a long time or simply wasn't holding pressure to expected mark.

_________________
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Additional Comments: Sleepyhead
Diabetes 2, RLS & bradycardia
Airsense For Her; Settings: range 8-12, Airfit P10 (M)

User avatar
Wulfman...
Posts: 6688
Joined: Mon Sep 01, 2014 6:41 pm
Location: Nearest fishing spot

Re: Switch to Airsense 10 Autoset -- Mask Leaks!

Post by Wulfman... » Thu Oct 22, 2015 1:56 pm

Tallen234 wrote:My Resmed S9 Autoset finally kicked the bucket, so I "upgraded" to an Airsense 10 Autoset. Using the same mask as before (F&P 431), I am now experiencing airleaks. Even when doing to Mask Fit setting, I can't get a tight seal. Even if I hold the mask to my face to make the tightest seal I can, it still gives me the frowny face. With my S9 and the same mask, I would get the occasionally leak, but most mornings I would wake up to a smiley face. Any ideas?
Please update your profile with the correct information. Your profile still shows a humidifier from an S8 machine.
Check all of your settings in the new machine and also list them in your profile. Sounds like the settings between the S9 and A10 are different.

And, as was also asked.......how did your S9 "kick the bucket"? Is it still under warranty?
The folks at: http://www.acbio.com do authorized factory repairs to ResMed and Philips/Respironics equipment.


Den

.
(5) REMstar Autos w/C-Flex & (6) REMstar Pro 2 CPAPs w/C-Flex - Pressure Setting = 14 cm.
"Passover" Humidification - ResMed Ultra Mirage FF - Encore Pro w/Card Reader & MyEncore software - Chiroflow pillow
User since 05/14/05

User avatar
Pugsy
Posts: 65169
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Switch to Airsense 10 Autoset -- Mask Leaks!

Post by Pugsy » Thu Oct 22, 2015 2:02 pm

If you can't get a smiley face for leaks even when masking up and you know it isn't leaking then I would look elsewhere in the circuit for potential leaks...like a gasket seal on the humidifier or lid not closed or something along those lines and also check the new hose to make sure it doesn't have a hole in it somewhere.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.