AHI recently increased

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
aghauler
Posts: 7
Joined: Fri Oct 16, 2015 11:25 am

AHI recently increased

Post by aghauler » Fri Oct 16, 2015 11:42 am

Equipment:
Respironics System One with humidifier
Wisp Nasal Mask + Chin Strap
Preset CPAP pressure 13cmH2O
Length of use 2.5 years
Average AHI number 2-7

Two weeks ago I had two cardiac events (heart attacks) and am now on Beta Blockers and diuretics while in the recovery stage before I can start Rehab.
When I checked the onboard Data information on my CPAP machine it indicates my AHI has increased to the 16-22 range. Periodic breathing has increased also
to sometimes greater than 45%, Large leak is zero.

My question is why the sudden increase in AHI?

My pulmonologist doesn't believe in AUTOPAP insists the 13 cmH2O is correct.
He will only prescribe melatonin as sleep aid now (tried it and didn't the side effects). Prior to that he only recommended Lunesta which has worked for me in the past.

nanwilson
Posts: 3463
Joined: Tue Apr 13, 2010 10:35 am
Location: Southern Alberta

Re: AHI recently increased

Post by nanwilson » Fri Oct 16, 2015 11:57 am

Two weeks ago 2 heart attacks...... when that happened to me it took some time for my body to settle back down to a reasonable pace. You have had 2 heart attacks, and are taking some pretty heavy heart drugs...give it time .. and don't add another drug to the mix, just try and relax and let your body heal. Once things settle down with your heart and whatever else it has jogged you will SLOWLY get back to your normal self and your ahi will go down.
Good luck with your recovery.
Nan
Started cpap in 2010.. still at it with great results.

User avatar
Julie
Posts: 20051
Joined: Tue Feb 28, 2006 12:58 pm

Re: AHI recently increased

Post by Julie » Fri Oct 16, 2015 12:20 pm

Beta blockers are considered to be counter productive to using Cpap by many.

aghauler
Posts: 7
Joined: Fri Oct 16, 2015 11:25 am

Re: AHI recently increased

Post by aghauler » Fri Oct 16, 2015 12:29 pm

nanwilson wrote:Two weeks ago 2 heart attacks...... when that happened to me it took some time for my body to settle back down to a reasonable pace. You have had 2 heart attacks, and are taking some pretty heavy heart drugs...give it time .. and don't add another drug to the mix, just try and relax and let your body heal. Once things settle down with your heart and whatever else it has jogged you will SLOWLY get back to your normal self and your ahi will go down.
Good luck with your recovery.
Nan
Thanks for the info and encouragement, glad I found this forum.

aghauler
Posts: 7
Joined: Fri Oct 16, 2015 11:25 am

Re: AHI recently increased

Post by aghauler » Fri Oct 16, 2015 12:33 pm

Julie wrote:Beta blockers are considered to be counter productive to using Cpap by many.
Interesting, how so?

Beta blockers post MI are used to decrease the workload on the heart while it heals.
With OSA the CPAP still has to be used.

Thanks for the input.

User avatar
Julie
Posts: 20051
Joined: Tue Feb 28, 2006 12:58 pm

Re: AHI recently increased

Post by Julie » Fri Oct 16, 2015 12:42 pm

Does your MD want you to be on BBs long term?

User avatar
Wulfman...
Posts: 6688
Joined: Mon Sep 01, 2014 6:41 pm
Location: Nearest fishing spot

Re: AHI recently increased

Post by Wulfman... » Fri Oct 16, 2015 1:10 pm

aghauler wrote:Equipment:
Respironics System One with humidifier
Wisp Nasal Mask + Chin Strap
Preset CPAP pressure 13cmH2O
Length of use 2.5 years
Average AHI number 2-7

Two weeks ago I had two cardiac events (heart attacks) and am now on Beta Blockers and diuretics while in the recovery stage before I can start Rehab.
When I checked the onboard Data information on my CPAP machine it indicates my AHI has increased to the 16-22 range. Periodic breathing has increased also to sometimes greater than 45%, Large leak is zero.

My question is why the sudden increase in AHI?

My pulmonologist doesn't believe in AUTOPAP insists the 13 cmH2O is correct.
He will only prescribe melatonin as sleep aid now (tried it and didn't the side effects). Prior to that he only recommended Lunesta which has worked for me in the past.
Which specific model do you have? Is it a data-capable machine? It appears to be from the numbers you posted.
Are you using software (like Sleepyhead or Encore) to SEE what's going on at night? If not, why not? You NEED to. It would help answer your questions.

And, due to your heart conditions, your pulmonologist has it exactly right.


Den

.
(5) REMstar Autos w/C-Flex & (6) REMstar Pro 2 CPAPs w/C-Flex - Pressure Setting = 14 cm.
"Passover" Humidification - ResMed Ultra Mirage FF - Encore Pro w/Card Reader & MyEncore software - Chiroflow pillow
User since 05/14/05

aghauler
Posts: 7
Joined: Fri Oct 16, 2015 11:25 am

Re: AHI recently increased

Post by aghauler » Mon Oct 19, 2015 2:47 pm

Julie wrote:Does your MD want you to be on BBs long term?
Undetermined at this point, may only be during the recovery - healing phase.

aghauler
Posts: 7
Joined: Fri Oct 16, 2015 11:25 am

Re: AHI recently increased

Post by aghauler » Mon Oct 19, 2015 3:04 pm

Wulfman... wrote:
aghauler wrote:Equipment:
Respironics System One with humidifier
Wisp Nasal Mask + Chin Strap
Preset CPAP pressure 13cmH2O
Length of use 2.5 years
Average AHI number 2-7

Two weeks ago I had two cardiac events (heart attacks) and am now on Beta Blockers and diuretics while in the recovery stage before I can start Rehab.
When I checked the onboard Data information on my CPAP machine it indicates my AHI has increased to the 16-22 range. Periodic breathing has increased also to sometimes greater than 45%, Large leak is zero.

My question is why the sudden increase in AHI?

My pulmonologist doesn't believe in AUTOPAP insists the 13 cmH2O is correct.
He will only prescribe melatonin as sleep aid now (tried it and didn't the side effects). Prior to that he only recommended Lunesta which has worked for me in the past.
Which specific model do you have? Is it a data-capable machine? It appears to be from the numbers you posted.
Are you using software (like Sleepyhead or Encore) to SEE what's going on at night? If not, why not? You NEED to. It would help answer your questions.

And, due to your heart conditions, your pulmonologist has it exactly right.


Den

.
This CPAP does have a data card. Didn't know anything about the available software, and had wondered about it's availability. Also thought if I printed the data on the data card it would erase the info so it wouldn't be available to anyone else.
Pulmonologist and staff are very controlling and not very empathetic to my situation, this was before any heart issues showed up. The "too bad you have to drive 3 hours to see us, don't try any other mask, I'm holier than thou staff attitude" has turned me off to them and I've currently turned my CPAP issues over to my Primary Care Provider. The local vendors don't like dealing with this particular Sleep Lab group due to their dictatorial attitude.
Also I should add I've been a Respiratory Therapist for 45 years, so know "something" about BiPaPs used clinically, but not too much regarding CPAP machines as that is the Night Shift realm. I've setup and managed numerous BiPAPs on Day shift, never have setup a CPAP on Day shift though.

User avatar
Wulfman...
Posts: 6688
Joined: Mon Sep 01, 2014 6:41 pm
Location: Nearest fishing spot

Re: AHI recently increased

Post by Wulfman... » Mon Oct 19, 2015 3:25 pm

aghauler wrote:
Wulfman... wrote:
aghauler wrote:Equipment:
Respironics System One with humidifier
Wisp Nasal Mask + Chin Strap
Preset CPAP pressure 13cmH2O
Length of use 2.5 years
Average AHI number 2-7

Two weeks ago I had two cardiac events (heart attacks) and am now on Beta Blockers and diuretics while in the recovery stage before I can start Rehab.
When I checked the onboard Data information on my CPAP machine it indicates my AHI has increased to the 16-22 range. Periodic breathing has increased also to sometimes greater than 45%, Large leak is zero.

My question is why the sudden increase in AHI?

My pulmonologist doesn't believe in AUTOPAP insists the 13 cmH2O is correct.
He will only prescribe melatonin as sleep aid now (tried it and didn't the side effects). Prior to that he only recommended Lunesta which has worked for me in the past.
Which specific model do you have? Is it a data-capable machine? It appears to be from the numbers you posted.
Are you using software (like Sleepyhead or Encore) to SEE what's going on at night? If not, why not? You NEED to. It would help answer your questions.

And, due to your heart conditions, your pulmonologist has it exactly right.


Den

.
This CPAP does have a data card. Didn't know anything about the available software, and had wondered about it's availability. Also thought if I printed the data on the data card it would erase the info so it wouldn't be available to anyone else.
Pulmonologist and staff are very controlling and not very empathetic to my situation, this was before any heart issues showed up. The "too bad you have to drive 3 hours to see us, don't try any other mask, I'm holier than thou staff attitude" has turned me off to them and I've currently turned my CPAP issues over to my Primary Care Provider. The local vendors don't like dealing with this particular Sleep Lab group due to their dictatorial attitude.
Also I should add I've been a Respiratory Therapist for 45 years, so know "something" about BiPaPs used clinically, but not too much regarding CPAP machines as that is the Night Shift realm. I've setup and managed numerous BiPAPs on Day shift, never have setup a CPAP on Day shift though.
Let's try again........
WHICH SPECIFIC model of machine do you have?
There are a few low-end machines that don't collect data but still have data cards in them.
You may have to peek at the bottom of your machine to see the model number, but if it's "450" or higher, it's fully data-capable. If it's lower than that, it's not. Model names that are data capable are "Pro" or "Auto".

See if this link will help.

https://www.cpap.com/cpap-machines/phil ... s.php#:::1::


Den

.
(5) REMstar Autos w/C-Flex & (6) REMstar Pro 2 CPAPs w/C-Flex - Pressure Setting = 14 cm.
"Passover" Humidification - ResMed Ultra Mirage FF - Encore Pro w/Card Reader & MyEncore software - Chiroflow pillow
User since 05/14/05

aghauler
Posts: 7
Joined: Fri Oct 16, 2015 11:25 am

Re: AHI recently increased

Post by aghauler » Mon Oct 19, 2015 3:32 pm

Wulfman... wrote:
aghauler wrote:
Wulfman... wrote:
aghauler wrote:Equipment:
Respironics System One with humidifier
Wisp Nasal Mask + Chin Strap
Preset CPAP pressure 13cmH2O
Length of use 2.5 years
Average AHI number 2-7

Two weeks ago I had two cardiac events (heart attacks) and am now on Beta Blockers and diuretics while in the recovery stage before I can start Rehab.
When I checked the onboard Data information on my CPAP machine it indicates my AHI has increased to the 16-22 range. Periodic breathing has increased also to sometimes greater than 45%, Large leak is zero.

My question is why the sudden increase in AHI?

My pulmonologist doesn't believe in AUTOPAP insists the 13 cmH2O is correct.
He will only prescribe melatonin as sleep aid now (tried it and didn't the side effects). Prior to that he only recommended Lunesta which has worked for me in the past.
Which specific model do you have? Is it a data-capable machine? It appears to be from the numbers you posted.
Are you using software (like Sleepyhead or Encore) to SEE what's going on at night? If not, why not? You NEED to. It would help answer your questions.

And, due to your heart conditions, your pulmonologist has it exactly right.


Den

.
This CPAP does have a data card. Didn't know anything about the available software, and had wondered about it's availability. Also thought if I printed the data on the data card it would erase the info so it wouldn't be available to anyone else.
Pulmonologist and staff are very controlling and not very empathetic to my situation, this was before any heart issues showed up. The "too bad you have to drive 3 hours to see us, don't try any other mask, I'm holier than thou staff attitude" has turned me off to them and I've currently turned my CPAP issues over to my Primary Care Provider. The local vendors don't like dealing with this particular Sleep Lab group due to their dictatorial attitude.
Also I should add I've been a Respiratory Therapist for 45 years, so know "something" about BiPaPs used clinically, but not too much regarding CPAP machines as that is the Night Shift realm. I've setup and managed numerous BiPAPs on Day shift, never have setup a CPAP on Day shift though.
Let's try again........
WHICH SPECIFIC model of machine do you have?
There are a few low-end machines that don't collect data but still have data cards in them.
You may have to peek at the bottom of your machine to see the model number, but if it's "450" or higher, it's fully data-capable. If it's lower than that, it's not. Model names that are data capable are "Pro" or "Auto".

See if this link will help.

https://www.cpap.com/cpap-machines/phil ... s.php#:::1::


Den

.
Phillps Respironics
REMstar Auto A-Flex
System One

User avatar
OkyDoky
Posts: 2869
Joined: Mon Aug 25, 2014 5:18 pm

Re: AHI recently increased

Post by OkyDoky » Mon Oct 19, 2015 3:32 pm

This site was developed by a member here, Krelvin, and has a lot of info on how to download and use Sleepyhead. Be sure and click on all the links to see how to organize your charts and post, if you want, for further help. https://sleep.tnet.com/resources/sleepyhead
Also if you tell us the REF# on the bottom of your blower, be sure and remove you humidifier before turning upside down, we can point you to where you can get the clinical manual to see how to change settings if you want.
This link may help you identify your machine. https://sleep.tnet.com/cmd
Like Wulfman said some do not record detailed data. So we need the specific model to give better help.
ResMed Aircurve 10 VAUTO EPAP 11 IPAP 15 / P10 pillows mask / Sleepyhead Software / Back up & travel machine Respironics 760

aghauler
Posts: 7
Joined: Fri Oct 16, 2015 11:25 am

Re: AHI recently increased

Post by aghauler » Mon Oct 19, 2015 3:42 pm

OkyDoky wrote:This site was developed by a member here, Krelvin, and has a lot of info on how to download and use Sleepyhead. Be sure and click on all the links to see how to organize your charts and post, if you want, for further help. https://sleep.tnet.com/resources/sleepyhead
Also if you tell us the REF# on the bottom of your blower, be sure and remove you humidifier before turning upside down, we can point you to where you can get the clinical manual to see how to change settings if you want.
This link may help you identify your machine. https://sleep.tnet.com/cmd
Like Wulfman said some do not record detailed data. So we need the specific model to give better help.

Ok looked at the bottom of the CPAP unit:

REMstar AUTO DOM
REF 560P

IP22

User avatar
OkyDoky
Posts: 2869
Joined: Mon Aug 25, 2014 5:18 pm

Re: AHI recently increased

Post by OkyDoky » Mon Oct 19, 2015 3:54 pm

Good! That means it is fully data capable and you can download the SD card to sleepyhead to see what is going on. Sleepyhead does not erase your card but you want to be sure and slide the card lock to lock when you put it in your computer so the computer doesn't write any unwanted files, and the slide it back to write when you put it back in your machine.
If you have any questions about downloading Sleepyhead ask and someone will help. Use my links above.
If you go to this link you can go to the bottom of the page and follow the instructions for Respironics PR System One 60 Series REMstar Auto with A-flex and request the manual by email. http://www.apneaboard.com/adjust-cpap-p ... tup-manual
ResMed Aircurve 10 VAUTO EPAP 11 IPAP 15 / P10 pillows mask / Sleepyhead Software / Back up & travel machine Respironics 760

User avatar
Wulfman...
Posts: 6688
Joined: Mon Sep 01, 2014 6:41 pm
Location: Nearest fishing spot

Re: AHI recently increased

Post by Wulfman... » Mon Oct 19, 2015 4:03 pm

aghauler wrote:Phillps Respironics
REMstar Auto A-Flex
System One
Good. That's a fully data-capable machine.

Now, get the software and take control of YOUR therapy.


Den

.
(5) REMstar Autos w/C-Flex & (6) REMstar Pro 2 CPAPs w/C-Flex - Pressure Setting = 14 cm.
"Passover" Humidification - ResMed Ultra Mirage FF - Encore Pro w/Card Reader & MyEncore software - Chiroflow pillow
User since 05/14/05