Any idea why high AHI when using FFM
Any idea why high AHI when using FFM
I am trying to figure out why a FFM is not working for me (I have tried 2, F10 and Simplus plus the Amara View - not sure if it is considered a FF or something else).
I have been using the Wisp but need a chin guard to keep my mouth closed and have been trying to get away from this. I have the same issue with both FFM and the Amara View; my AHI goes from 1-2 with the wisp to 8 to 14 with the FFM . I am trying to figure out why, the masks all seem pretty comfortable and my leak rate is not higher than with the Wisp but...
Any thoughts?
I have been using the Wisp but need a chin guard to keep my mouth closed and have been trying to get away from this. I have the same issue with both FFM and the Amara View; my AHI goes from 1-2 with the wisp to 8 to 14 with the FFM . I am trying to figure out why, the masks all seem pretty comfortable and my leak rate is not higher than with the Wisp but...
Any thoughts?
_________________
| Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier |
| Mask: Wisp Nasal CPAP Mask with Headgear - Fit Pack |
| Additional Comments: Also Sleepyhead, P10, F10 and Amara View |
Re: Any idea why high AHI when using FFM
Have you looked at a Sleepyhead graph? Please fill in your profile with your equipment. (still sounds like leaks).
_________________
| Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier |
| Mask: Fisher & Paykel Vitera Full Face Mask with Headgear (S, M, or L Cushion) |
| Additional Comments: Back up is a new AS10. |
Re: Any idea why high AHI when using FFM
I have, the last two nights (the nights I went back to a FF mask my leak rate was the lowest ever, doesn't even show on the graph but the AHI went from .55 three nights ago with the Wisp to 8.9 two nights ago and 13.01 last night.LSAT wrote:Have you looked at a Sleepyhead graph? Please fill in your profile with your equipment. (still sounds like leaks).
I don't think I can upload graphs yet, correct?
I believe my profile of equipment is listed; I see it, hoping you do too
_________________
| Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier |
| Mask: Wisp Nasal CPAP Mask with Headgear - Fit Pack |
| Additional Comments: Also Sleepyhead, P10, F10 and Amara View |
Re: Any idea why high AHI when using FFM
How to do screen shots of reports along with examples in this thread
viewtopic/t103468/Need-help-with-screen-shots.html
and more details here ..be sure to read all 3 pages
https://sleep.tnet.com/reference/tips/imgur
What is the AHI primarily made up of on the high AHI nights?
What pressure(s) settings are you using?
Off hand without seeing any reports...if the AHI is primarily OA then I suspect you need more pressure with a full face mask than you do with a nasal mask.
Best to evaluate the reports though.
viewtopic/t103468/Need-help-with-screen-shots.html
and more details here ..be sure to read all 3 pages
https://sleep.tnet.com/reference/tips/imgur
What is the AHI primarily made up of on the high AHI nights?
What pressure(s) settings are you using?
Off hand without seeing any reports...if the AHI is primarily OA then I suspect you need more pressure with a full face mask than you do with a nasal mask.
Best to evaluate the reports though.
_________________
| Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
| Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/ |
I may have to RISE but I refuse to SHINE.
Re: Any idea why high AHI when using FFM
Pugsy, I think you also need to change your mask in the clinical setting to FFM from Nasal mask, don't know if this makes any difference or not with the pressures.Pugsy wrote:How to do screen shots of reports along with examples in this thread
viewtopic/t103468/Need-help-with-screen-shots.html
and more details here ..be sure to read all 3 pages
https://sleep.tnet.com/reference/tips/imgur
What is the AHI primarily made up of on the high AHI nights?
What pressure(s) settings are you using?
Off hand without seeing any reports...if the AHI is primarily OA then I suspect you need more pressure with a full face mask than you do with a nasal mask.
Best to evaluate the reports though.
_________________
| Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier |
| Mask: AirFit™ F10 Full Face Mask with Headgear |
| Additional Comments: FlexiFit HC431 FFM with Headgear |
Re: Any idea why high AHI when using FFM
Light bulb just went off, I completely forgot about that. Will do it now before I forget again.cancun wrote:Pugsy, I think you also need to change your mask in the clinical setting to FFM from Nasal mask, don't know if this makes any difference or not with the pressures.Pugsy wrote:How to do screen shots of reports along with examples in this thread
viewtopic/t103468/Need-help-with-screen-shots.html
and more details here ..be sure to read all 3 pages
https://sleep.tnet.com/reference/tips/imgur
What is the AHI primarily made up of on the high AHI nights?
What pressure(s) settings are you using?
Off hand without seeing any reports...if the AHI is primarily OA then I suspect you need more pressure with a full face mask than you do with a nasal mask.
Best to evaluate the reports though.
_________________
| Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier |
| Mask: Wisp Nasal CPAP Mask with Headgear - Fit Pack |
| Additional Comments: Also Sleepyhead, P10, F10 and Amara View |
Re: Any idea why high AHI when using FFM
Pugsy wrote:How to do screen shots of reports along with examples in this thread
viewtopic/t103468/Need-help-with-screen-shots.html
and more details here ..be sure to read all 3 pages
https://sleep.tnet.com/reference/tips/imgur
What is the AHI primarily made up of on the high AHI nights?
What pressure(s) settings are you using?
Off hand without seeing any reports...if the AHI is primarily OA then I suspect you need more pressure with a full face mask than you do with a nasal mask.
Best to evaluate the reports though.
Will read these and see if I can upload; I thought you had to have 25 posts or something like that
_________________
| Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier |
| Mask: Wisp Nasal CPAP Mask with Headgear - Fit Pack |
| Additional Comments: Also Sleepyhead, P10, F10 and Amara View |
Re: Any idea why high AHI when using FFM
The mask selection choice in the ResMed machines is primarily for establishing a baseline for figuring excess leaks. To have the incorrect mask choice might hide a little excess leak but it won't hide a lot and unless the leak was marginal to begin with I doubt it makes much difference.cancun wrote:Pugsy, I think you also need to change your mask in the clinical setting to FFM from Nasal mask, don't know if this makes any difference or not with the pressures.
The mask choice does NOT affect pressure though...X amount of pressure is X amount of pressure no matter what the mask choice.
The most common cause for higher AHI (that is made up of OAs and hyponeas) when comparing a nasal mask to a FFM is that the FFM needs a little more pressure to do the same job. Don't know why because in theory X amount of pressure should yield the same results no matter which mask but in real life the theory doesn't hold up. Now it might have something to do with the lower straps of some FFM maybe affecting the jaw position...that's a maybe.
All I know is that I have seen lots and lots of reports where people need different pressures to get the same job done between nasal masks and FFmasks...and full face is always needing the higher.
Now if the higher AHI is made up of mainly centrals...we have a different discussion.
_________________
| Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
| Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/ |
I may have to RISE but I refuse to SHINE.
Re: Any idea why high AHI when using FFM
Full face masks typically have a little higher vent rate. If by chance the ResMed machine thinks you are still using a nasal mask it is not going to make the leak report all that different but if anything it will over estimate the excess leak because it would be still working under the assumption a nasal mask was being used which would have a lower vent rate to start with.
Incorrect mask choice would be very unlikely to be the cause of the higher AHI...at worst the leak rate might be off a little but that's about it.
Incorrect mask choice would be very unlikely to be the cause of the higher AHI...at worst the leak rate might be off a little but that's about it.
_________________
| Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
| Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/ |
I may have to RISE but I refuse to SHINE.
Re: Any idea why high AHI when using FFM
Will upload when I get home tonight but here is the makeup of last nightPugsy wrote:Now if the higher AHI is made up of mainly centrals...we have a different discussion.cancun wrote:Pugsy, I think you also need to change your mask in the clinical setting to FFM from Nasal mask, don't know if this makes any difference or not with the pressures.
CA 6.99
Unclassified 0.00
OA 3.37
H 2.65
RERA 3.37
Large Leak .004%
For reference, my last night on Wisp (3 nights ago)
CA .37
Unclassified 0.00
OA .18
H 0.00
RERA 0.00
Large Leak 2.78%
_________________
| Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier |
| Mask: Wisp Nasal CPAP Mask with Headgear - Fit Pack |
| Additional Comments: Also Sleepyhead, P10, F10 and Amara View |
Re: Any idea why high AHI when using FFM
Posting of images or links to images (or anything else for that matter) is NOT tied to post count.
All you have to be is logged in under your member ID.
People who post and are not logged in are the ones who can't post links to anything.
All you have to be is logged in under your member ID.
People who post and are not logged in are the ones who can't post links to anything.
_________________
| Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
| Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/ |
I may have to RISE but I refuse to SHINE.
Re: Any idea why high AHI when using FFM
The CAs are Clear Airway events or Centrals and we don't treat those with more pressure because the airway is already open.flightco wrote: Will upload when I get home tonight but here is the makeup of last night
CA 6.99
Unclassified 0.00
OA 3.37
H 2.65
RERA 3.37
Large Leak .004%
For reference, my last night on Wisp (3 nights ago)
CA .37
Unclassified 0.00
OA .18
H 0.00
RERA 0.00
Large Leak 2.78%
The increase in CAs along with the increase in RERAs makes me wonder if the CAs are related to arousals or wake ups.
Now are those potential wake ups related to the also increase in OAs.....that's a million dollar question.
Need to see exactly when those increased events are happening in relation to the sleep time line.
_________________
| Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
| Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/ |
I may have to RISE but I refuse to SHINE.
Re: Any idea why high AHI when using FFM
Need to see exactly when those increased events are happening in relation to the sleep time line.[/quote]
Will get those uploaded this evening.
After you spoke about higher pressures with a FF mask, I looked and saw that my 95% pressure was 15 which is the max I have it set for; could that be the problem?
Will get those uploaded this evening.
After you spoke about higher pressures with a FF mask, I looked and saw that my 95% pressure was 15 which is the max I have it set for; could that be the problem?
_________________
| Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier |
| Mask: Wisp Nasal CPAP Mask with Headgear - Fit Pack |
| Additional Comments: Also Sleepyhead, P10, F10 and Amara View |
Re: Any idea why high AHI when using FFM
Thanks Pugsy, of course my DME didn't tell me to change my mask when I got the pillow I had just remembered seeing it when I went in to the clinical menu. I couldn't deal with the pillow and went back to the FFM.Pugsy wrote:The mask selection choice in the ResMed machines is primarily for establishing a baseline for figuring excess leaks. To have the incorrect mask choice might hide a little excess leak but it won't hide a lot and unless the leak was marginal to begin with I doubt it makes much difference.cancun wrote:Pugsy, I think you also need to change your mask in the clinical setting to FFM from Nasal mask, don't know if this makes any difference or not with the pressures.
The mask choice does NOT affect pressure though...X amount of pressure is X amount of pressure no matter what the mask choice.
The most common cause for higher AHI (that is made up of OAs and hyponeas) when comparing a nasal mask to a FFM is that the FFM needs a little more pressure to do the same job. Don't know why because in theory X amount of pressure should yield the same results no matter which mask but in real life the theory doesn't hold up. Now it might have something to do with the lower straps of some FFM maybe affecting the jaw position...that's a maybe.
All I know is that I have seen lots and lots of reports where people need different pressures to get the same job done between nasal masks and FFmasks...and full face is always needing the higher.
Now if the higher AHI is made up of mainly centrals...we have a different discussion.
_________________
| Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier |
| Mask: AirFit™ F10 Full Face Mask with Headgear |
| Additional Comments: FlexiFit HC431 FFM with Headgear |
Re: Any idea why high AHI when using FFM