ResMed AirSense 10 "Smart Start" Doesn't Shut Down

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
User avatar
Wulfman...
Posts: 6688
Joined: Mon Sep 01, 2014 6:41 pm
Location: Nearest fishing spot

Re: ResMed AirSense 10 "Smart Start" Doesn't Shut Down

Post by Wulfman... » Wed Aug 19, 2015 12:44 pm

palerider wrote:
Wulfman... wrote:
Reuven Gruber wrote:Where in the manual does it require a certain pressure for the feature to work [and which manual are you referring to?]
Page 32.
And, I (like Krelvin) think it's working as it's supposed to.

In the previous generations of the ResMed machines, when a full face mask was selected, it turned off the Smartstart feature.
in really OLD previous machines, like first software revision s9 machines, but every one I've had smart start worked with all mask selections.
Right.
But, then I have never and would never own a ResMed machine to test its features.


Den

.
(5) REMstar Autos w/C-Flex & (6) REMstar Pro 2 CPAPs w/C-Flex - Pressure Setting = 14 cm.
"Passover" Humidification - ResMed Ultra Mirage FF - Encore Pro w/Card Reader & MyEncore software - Chiroflow pillow
User since 05/14/05

User avatar
palerider
Posts: 32300
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: ResMed AirSense 10 "Smart Start" Doesn't Shut Down

Post by palerider » Wed Aug 19, 2015 1:03 pm

Wulfman... wrote:
palerider wrote:in really OLD previous machines, like first software revision s9 machines, but every one I've had smart start worked with all mask selections.
Right.
But, then I have never and would never own a ResMed machine to test its features.
not a problem. I don't have any of the big old respirs, so that's why I leave the advice about them to you cuz I know you know more about 'em than I ever will

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
Wulfman...
Posts: 6688
Joined: Mon Sep 01, 2014 6:41 pm
Location: Nearest fishing spot

Re: ResMed AirSense 10 "Smart Start" Doesn't Shut Down

Post by Wulfman... » Wed Aug 19, 2015 1:12 pm

palerider wrote:
Wulfman... wrote:
palerider wrote:in really OLD previous machines, like first software revision s9 machines, but every one I've had smart start worked with all mask selections.
Right.
But, then I have never and would never own a ResMed machine to test its features.
not a problem. I don't have any of the big old respirs, so that's why I leave the advice about them to you cuz I know you know more about 'em than I ever will
Probably like most forum members......other than the ones I/we own, the amount of knowledge we accumulate is from what we see in the clinical manuals and what users report. When their experiences differ from what the manuals (and other users) say, then we have interesting threads like this one.

Sooooo........in your opinion, is the OP's machine working "properly"?


Den

.
(5) REMstar Autos w/C-Flex & (6) REMstar Pro 2 CPAPs w/C-Flex - Pressure Setting = 14 cm.
"Passover" Humidification - ResMed Ultra Mirage FF - Encore Pro w/Card Reader & MyEncore software - Chiroflow pillow
User since 05/14/05

User avatar
Reuven Gruber
Posts: 39
Joined: Tue Sep 16, 2014 3:29 am
Location: Israel

Re: ResMed AirSense 10 "Smart Start" Doesn't Shut Down

Post by Reuven Gruber » Wed Aug 19, 2015 3:26 pm

My machine cant possibly be considered to be working properly - there is an advertised and documented feature (automatic shut off) that doesn't work

_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Additional Comments: Pressure limits 4.0 to 8.0, ResScan 5.7 Software, Ruby Chin Strap. Untreated AHI 32; With CPAP 1

User avatar
MarylandCPAPer
Posts: 438
Joined: Fri Jan 15, 2010 1:15 pm

Re: ResMed AirSense 10 "Smart Start" Doesn't Shut Down

Post by MarylandCPAPer » Wed Aug 19, 2015 4:40 pm

Reuven Gruber wrote:MarylandCPAPer

Nice cat

Your post shows that you are using the model 10. Does automatic shutoff work for you?
Yes. My pressure range is from 10.6 to 20, so I never use it with a pressure less than 7.

One way to test it, if you can tolerate it, would be to set the pressure up to 8, Put the mask on and let it run for a minute, then take the mask off. That would avoid the 7 cm pressure range issue.

You can change your settings using the information in the previously linked clinician menu. Also, you may want to change the settings on Essentials to Plus, if it is not already, so that you can see more data on your display (just for general reference, not to do with the current issue you have),

Has the auto off feature EVER worked for you on this machine?

_________________
Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear
Additional Comments: Have also used Opus 360, Swift LT for Her, Optilife, and Breeze Nasal Pillows masks.
Started CPAP 12/18/09 Pressure 13. Changed to APAP 1/18/10, Pressure 10-16. 4/2014:Switched to AirFit P10 mask. PR System One REMstar Auto Series 550 with A-Flex for 5.5 years. 7/2015 to present: AirSense 10 AutoSet FOR HER. Range: 10-20

User avatar
Jay Aitchsee
Posts: 2936
Joined: Sun May 22, 2011 12:47 pm
Location: Southwest Florida

Re: ResMed AirSense 10 "Smart Start" Doesn't Shut Down

Post by Jay Aitchsee » Wed Aug 19, 2015 5:15 pm

Reuven Gruber wrote:My machine cant possibly be considered to be working properly - there is an advertised and documented feature (automatic shut off) that doesn't work
Why not take it to the DME and let them check it out?

_________________
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Additional Comments: S9 Auto, P10 mask, P=7.0, EPR3, ResScan 5.3, SleepyHead V1.B2, Windows 10, ZEO, CMS50F, Infrared Video

User avatar
chunkyfrog
Posts: 34417
Joined: Mon Jul 12, 2010 5:10 pm
Location: Nebraska--I am sworn to keep the secret of this paradise.

Re: ResMed AirSense 10 "Smart Start" Doesn't Shut Down

Post by chunkyfrog » Wed Aug 19, 2015 6:02 pm

Reuven Gruber wrote:Chunkyfrog

Your post shows that you are using the AirSense 10. Does automatic shutoff work for you?
It does shut down, but it takes longer than I care to wait. I use a small P10.
I have no trouble reaching over to hit the button, so I do it just to keep the noise from waking the spousal unit.
It sounds like there may be a defect in your machine.
Darlin', just trot it over to the DME. The sooner you do, the sooner it gets fixed--or whatever.

_________________
Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear
Additional Comments: Airsense 10 Autoset for Her

User avatar
palerider
Posts: 32300
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: ResMed AirSense 10 "Smart Start" Doesn't Shut Down

Post by palerider » Wed Aug 19, 2015 6:15 pm

Wulfman... wrote:Sooooo........in your opinion, is the OP's machine working "properly"?
well, it's working "within the parameters established by the manual", no matter what our favorite manual nitpicker wants to say.

easy enough to figure out if it's working right or not, set the min pressure to 8, and see, if it cuts off, then it's probably just not as sensitive as most, if it doesn't, it's got a glitch.

I recently had an issue with a vauto machine because I'd put a AB filter on it, but the setting was 'no" so the ab filter caused enough back pressure that the machine didn't recognize the mask off situation.

I just checked the airsense manual, and it shows that there's an ab filter setting for them. I wonder if the ab filter setting is proper on the OP's machine.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
MarylandCPAPer
Posts: 438
Joined: Fri Jan 15, 2010 1:15 pm

Re: ResMed AirSense 10 "Smart Start" Doesn't Shut Down

Post by MarylandCPAPer » Wed Aug 19, 2015 6:40 pm

palerider wrote:
Wulfman... wrote:Sooooo........in your opinion, is the OP's machine working "properly"?

I recently had an issue with a vauto machine because I'd put a AB filter on it, but the setting was 'no" so the ab filter caused enough back pressure that the machine didn't recognize the mask off situation.

I just checked the airsense manual, and it shows that there's an ab filter setting for them. I wonder if the ab filter setting is proper on the OP's machine.
This is a tangent, but I am wondering how often you have to replace anti-bacterial filters and whether a dr. has to prescribe them for a DME to provide them. I got a regular filter with the A10. Just curious about the AB filters. Interesting that they could affect auto off. Do AB filters require higher pressure settings?

_________________
Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear
Additional Comments: Have also used Opus 360, Swift LT for Her, Optilife, and Breeze Nasal Pillows masks.
Started CPAP 12/18/09 Pressure 13. Changed to APAP 1/18/10, Pressure 10-16. 4/2014:Switched to AirFit P10 mask. PR System One REMstar Auto Series 550 with A-Flex for 5.5 years. 7/2015 to present: AirSense 10 AutoSet FOR HER. Range: 10-20

User avatar
palerider
Posts: 32300
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: ResMed AirSense 10 "Smart Start" Doesn't Shut Down

Post by palerider » Wed Aug 19, 2015 7:19 pm

MarylandCPAPer wrote:This is a tangent,
no, and no.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
Jay Aitchsee
Posts: 2936
Joined: Sun May 22, 2011 12:47 pm
Location: Southwest Florida

Re: ResMed AirSense 10 "Smart Start" Doesn't Shut Down

Post by Jay Aitchsee » Wed Aug 26, 2015 12:01 pm

Reuven, did you ever get Smart Start to work to your liking?

_________________
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Additional Comments: S9 Auto, P10 mask, P=7.0, EPR3, ResScan 5.3, SleepyHead V1.B2, Windows 10, ZEO, CMS50F, Infrared Video

User avatar
jschumann
Posts: 15
Joined: Thu Oct 15, 2009 6:09 pm

Re: ResMed AirSense 10 "Smart Start" Doesn't Shut Down

Post by jschumann » Tue Apr 30, 2019 2:18 pm

I just got my AirSense 10 and I cannot enable Smart Start. It doesn't respond when selected. It appears to be 'greyed' out. Does another feature or setting need to be activated to make it come on?

Thanks
John

_________________
MachineMask

User avatar
Pugsy
Posts: 64109
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: ResMed AirSense 10 "Smart Start" Doesn't Shut Down

Post by Pugsy » Tue Apr 30, 2019 3:28 pm

jschumann wrote:
Tue Apr 30, 2019 2:18 pm
I just got my AirSense 10 and I cannot enable Smart Start. It doesn't respond when selected. It appears to be 'greyed' out. Does another feature or setting need to be activated to make it come on?

Thanks
John
Exactly which model machine do you have?.
Is there a leak alarm feature? If so...is it turned on? If it is turned on..turn it off and see if you can access SmartStart

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

If you want to try the Eclipse mask and want a special promo code to get a little off the price...send me a private message.

User avatar
jschumann
Posts: 15
Joined: Thu Oct 15, 2009 6:09 pm

Re: ResMed AirSense 10 "Smart Start" Doesn't Shut Down

Post by jschumann » Mon May 06, 2019 2:42 am

Pugsie said:
Exactly which model machine do you have?.
Is there a leak alarm feature? If so...is it turned on? If it is turned on..turn it off and see if you can access SmartStart

I have Aircurve 10 Vauto

Will try leaksense setting

John

_________________
MachineMask

User avatar
Pugsy
Posts: 64109
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: ResMed AirSense 10 "Smart Start" Doesn't Shut Down

Post by Pugsy » Mon May 06, 2019 2:50 am

jschumann wrote:
Mon May 06, 2019 2:42 am
Pugsie said:
Exactly which model machine do you have?.
Is there a leak alarm feature? If so...is it turned on? If it is turned on..turn it off and see if you can access SmartStart

I have Aircurve 10 Vauto

Will try leaksense setting

John

AirSense 10 machines don't have a leak alarm setting to turn off or on but the AirCurve 10 machines do have that setting in the setup menu area.
When the leak alarm setting is turned on the SmartStart auto on/off feature is disabled. You can have one or the other but not both.
It's been that way for at least over 10 years with the ResMed bilevel machines.
The AirCurve 10 is a bilevel machine...

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

If you want to try the Eclipse mask and want a special promo code to get a little off the price...send me a private message.