ResMed AirSense 10 "Smart Start" Doesn't Shut Down
- Reuven Gruber
- Posts: 39
- Joined: Tue Sep 16, 2014 3:29 am
- Location: Israel
ResMed AirSense 10 "Smart Start" Doesn't Shut Down
With smart start enabled the machine starts automatically but when I remove my mask (full face mask) the machine should automatically shut down but it keeps running. Any Ideas?
			_________________
| Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier | 
| Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear | 
| Additional Comments: Pressure limits 4.0 to 8.0, ResScan 5.7 Software, Ruby Chin Strap. Untreated AHI 32; With CPAP 1 | 
Re: ResMed AirSense 10 "Smart Start" Doesn't Shut Down
Define running.  It's it blowing lots of air or is it doing its cool down where it blows at a very low amount while the humidifier cools down.  If the latter, it is supposed to do that.  Even if you turn it off at the switch it will fyi the same.
If it continues to blow at a high rate something is not right.
			
			
									
									If it continues to blow at a high rate something is not right.
Current Settings PS 4.0 over 10.6-18.0 (cmH2O) - Resmed S9 VPAP Auto w/h5i Humidifier - Quattro Air FFM
TNET Sleep Resource Pages - CPAP Machine Database
Put your equip in your Signature - SleepyHead v1.0.0-beta-1
Kevin... alias Krelvin
						TNET Sleep Resource Pages - CPAP Machine Database
Put your equip in your Signature - SleepyHead v1.0.0-beta-1
Kevin... alias Krelvin
Re: ResMed AirSense 10 "Smart Start" Doesn't Shut Down
I've seen that happen when the mask and tubing settings aren't correct.Reuven Gruber wrote:With smart start enabled the machine starts automatically but when I remove my mask (full face mask) the machine should automatically shut down but it keeps running. Any Ideas?
Get OSCAR 
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
						Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
- Wulfman...
- Posts: 6688
- Joined: Mon Sep 01, 2014 6:41 pm
- Location: Nearest fishing spot
Re: ResMed AirSense 10 "Smart Start" Doesn't Shut Down
From the manual:Reuven Gruber wrote:With smart start enabled the machine starts automatically but when I remove my mask (full face mask) the machine should automatically shut down but it keeps running. Any Ideas?
SmartStart is enabled, but the device does not automatically stop when the patient removes their mask
Incompatible mask being used. Only use equipment recommended by ResMed.
Contact ResMed or see http://www.resmed.com for more information.
If the patient is using a nasal pillows mask with set pressure less than 7 cm H2O, SmartStart will not work and should be
disabled.
Den
.
(5) REMstar Autos w/C-Flex & (6) REMstar Pro 2 CPAPs w/C-Flex - Pressure Setting = 14 cm. 
"Passover" Humidification - ResMed Ultra Mirage FF - Encore Pro w/Card Reader & MyEncore software - Chiroflow pillow
User since 05/14/05
						"Passover" Humidification - ResMed Ultra Mirage FF - Encore Pro w/Card Reader & MyEncore software - Chiroflow pillow
User since 05/14/05
- Jay Aitchsee
- Posts: 2936
- Joined: Sun May 22, 2011 12:47 pm
- Location: Southwest Florida
Re: ResMed AirSense 10 "Smart Start" Doesn't Shut Down
Edit: I guess we covered that, didn't we Den?Clinician's Manual wrote:SmartStart is enabled, but the device does not automatically stop when the patient removes their mask
- Incompatible mask being used. Only use equipment recommended by ResMed.
Contact ResMed or see http://www.resmed.com for more
information.
If the patient is using a nasal pillows mask with set pressure
less than 7 cm H2O, SmartStart will not work and should be
disabled.
I'm betting on this: If you are using the humidifier the blower will run for approx. 20 minutes (at a reduced rate) after quitting therapy to prevent condensation in the hose.
_________________
| Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear | 
| Additional Comments: S9 Auto, P10 mask, P=7.0, EPR3, ResScan 5.3, SleepyHead V1.B2, Windows 10, ZEO, CMS50F, Infrared Video | 
- Reuven Gruber
- Posts: 39
- Joined: Tue Sep 16, 2014 3:29 am
- Location: Israel
Re: ResMed AirSense 10 "Smart Start" Doesn't Shut Down
Krelvin - - It blows hard. I let it go for 1 minute and 10 seconds and it was still blowing so I turned it off manually
Palerider - Settings for the mask and tubing are correct
			Palerider - Settings for the mask and tubing are correct
_________________
| Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier | 
| Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear | 
| Additional Comments: Pressure limits 4.0 to 8.0, ResScan 5.7 Software, Ruby Chin Strap. Untreated AHI 32; With CPAP 1 | 
- Reuven Gruber
- Posts: 39
- Joined: Tue Sep 16, 2014 3:29 am
- Location: Israel
Re: ResMed AirSense 10 "Smart Start" Doesn't Shut Down
Mask is ResMed F10 and tubing is ResMed Climateline
			_________________
| Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier | 
| Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear | 
| Additional Comments: Pressure limits 4.0 to 8.0, ResScan 5.7 Software, Ruby Chin Strap. Untreated AHI 32; With CPAP 1 | 
Re: ResMed AirSense 10 "Smart Start" Doesn't Shut Down
What pressure do you normally use?
Next time it happens look on the LCD screen to see what pressure it is blowing at.
			Next time it happens look on the LCD screen to see what pressure it is blowing at.
_________________
| Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier | 
| Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/ | 
I may have to RISE but I refuse to SHINE.
						- Jay Aitchsee
- Posts: 2936
- Joined: Sun May 22, 2011 12:47 pm
- Location: Southwest Florida
Re: ResMed AirSense 10 "Smart Start" Doesn't Shut Down
Reuven, a while back you said you were using pillows (as it says in your profile). Are you sure the mask setting of the machine is what you are actually using?Reuven Gruber wrote:Mask is ResMed F10 and tubing is ResMed Climateline
Seriously, I would (and do) just use the on/off button.
_________________
| Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear | 
| Additional Comments: S9 Auto, P10 mask, P=7.0, EPR3, ResScan 5.3, SleepyHead V1.B2, Windows 10, ZEO, CMS50F, Infrared Video | 
- Reuven Gruber
- Posts: 39
- Joined: Tue Sep 16, 2014 3:29 am
- Location: Israel
Re: ResMed AirSense 10 "Smart Start" Doesn't Shut Down
My pressure settings are Min 4.0; Max 10.0
I used to use pillows but now I am using the full face mask - I will update my profile
I do turn it off manually but I really want to get the device working correctly as advertised
			I used to use pillows but now I am using the full face mask - I will update my profile
I do turn it off manually but I really want to get the device working correctly as advertised
_________________
| Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier | 
| Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear | 
| Additional Comments: Pressure limits 4.0 to 8.0, ResScan 5.7 Software, Ruby Chin Strap. Untreated AHI 32; With CPAP 1 | 
Re: ResMed AirSense 10 "Smart Start" Doesn't Shut Down
Can you do an experiment while awake?
Change your minimum pressure from 4 cm to 6 or 7 cm...breathe enough for the machine to come on for maybe a minute...then remove the mask and see if the machine will turn off.
			Change your minimum pressure from 4 cm to 6 or 7 cm...breathe enough for the machine to come on for maybe a minute...then remove the mask and see if the machine will turn off.
_________________
| Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier | 
| Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/ | 
I may have to RISE but I refuse to SHINE.
						- Reuven Gruber
- Posts: 39
- Joined: Tue Sep 16, 2014 3:29 am
- Location: Israel
Re: ResMed AirSense 10 "Smart Start" Doesn't Shut Down
Pugsy
I set min pressure to 7, ran it for 2 minutes, took the mask off and it still didn't shut down
			I set min pressure to 7, ran it for 2 minutes, took the mask off and it still didn't shut down
_________________
| Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier | 
| Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear | 
| Additional Comments: Pressure limits 4.0 to 8.0, ResScan 5.7 Software, Ruby Chin Strap. Untreated AHI 32; With CPAP 1 | 
- Wulfman...
- Posts: 6688
- Joined: Mon Sep 01, 2014 6:41 pm
- Location: Nearest fishing spot
Re: ResMed AirSense 10 "Smart Start" Doesn't Shut Down
The manual says "less than 7 cm". I suspect that it's "blowing hard" trying to keep up whatever pressure it was at when you took the mask off. Even if it was below 7 cm., it would seem to be blowing hard trying to maintain that pressure.Reuven Gruber wrote:Pugsy
I set min pressure to 7, ran it for 2 minutes, took the mask off and it still didn't shut down
Den
.
(5) REMstar Autos w/C-Flex & (6) REMstar Pro 2 CPAPs w/C-Flex - Pressure Setting = 14 cm. 
"Passover" Humidification - ResMed Ultra Mirage FF - Encore Pro w/Card Reader & MyEncore software - Chiroflow pillow
User since 05/14/05
						"Passover" Humidification - ResMed Ultra Mirage FF - Encore Pro w/Card Reader & MyEncore software - Chiroflow pillow
User since 05/14/05
- Wulfman...
- Posts: 6688
- Joined: Mon Sep 01, 2014 6:41 pm
- Location: Nearest fishing spot
Re: ResMed AirSense 10 "Smart Start" Doesn't Shut Down
Is the mask setting in your machine correct?Reuven Gruber wrote:My pressure settings are Min 4.0; Max 10.0
I used to use pillows but now I am using the full face mask - I will update my profile
I do turn it off manually but I really want to get the device working correctly as advertised
Den
.
(5) REMstar Autos w/C-Flex & (6) REMstar Pro 2 CPAPs w/C-Flex - Pressure Setting = 14 cm. 
"Passover" Humidification - ResMed Ultra Mirage FF - Encore Pro w/Card Reader & MyEncore software - Chiroflow pillow
User since 05/14/05
						"Passover" Humidification - ResMed Ultra Mirage FF - Encore Pro w/Card Reader & MyEncore software - Chiroflow pillow
User since 05/14/05
Re: ResMed AirSense 10 "Smart Start" Doesn't Shut Down
Mine shuts off in about 5 seconds, min is never less than 10.6.  Doesn't matter what type of mask, Resmed or F&P.
			
			
									
									Current Settings PS 4.0 over 10.6-18.0 (cmH2O) - Resmed S9 VPAP Auto w/h5i Humidifier - Quattro Air FFM
TNET Sleep Resource Pages - CPAP Machine Database
Put your equip in your Signature - SleepyHead v1.0.0-beta-1
Kevin... alias Krelvin
						TNET Sleep Resource Pages - CPAP Machine Database
Put your equip in your Signature - SleepyHead v1.0.0-beta-1
Kevin... alias Krelvin
 
                 
                         
                         
                         
                         
                         
                         
                         
			 
	