new mask spiked events per hour?

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
wrath
Posts: 7
Joined: Sun Aug 16, 2015 8:17 am

new mask spiked events per hour?

Post by wrath » Sun Aug 16, 2015 8:24 am

I switched from a nasal mask with chin strap to a full nose & mouth mask this week.
I use a Resmed AirSense 10
The next day, & for the next 9 days after that, the events per hour tripled.
Is this related to the different style mask?
Doe the pressure need to be adjusted since the volume of the new mask is so much more than the old mask?
Nothing else was changed except the mask.

Thanks

User avatar
Pugsy
Posts: 65120
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: new mask spiked events per hour?

Post by Pugsy » Sun Aug 16, 2015 8:32 am

Sometimes people find out that more pressure is needed when using a full face mask as opposed to nasal mask. Not everyone though.
"Events tripled" doesn't really tell us much because we don't know which events tripled and how much was the "tripling"...was it tripled going from 1 to 3 or 10 to 30?
What tripled? If it were centrals then we wouldn't increase the pressure.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

wrath
Posts: 7
Joined: Sun Aug 16, 2015 8:17 am

Re: new mask spiked events per hour?

Post by wrath » Sun Aug 16, 2015 8:37 am

The events per hour posted on the MyAir dashboard went from under 5/hr to over 15/hr on average since I switched to the new mask.
I like the new mask because the chin strap I used with the old mask would always fall off.
I'm not sure what you mean by "central"
I appreciate the reply & help.

User avatar
Pugsy
Posts: 65120
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: new mask spiked events per hour?

Post by Pugsy » Sun Aug 16, 2015 8:46 am

There's a lot more to the data available than what you see on the MyAir website.
AHI is composed of 3 types of events.
Central apneas (where the airway is open but there's no air flow)
Obstructive apneas (where the tissues in the airway are blocking air flow)
Hyponeas (where the tissues in the airway are partially blocking the air flow)

I would suggest that you get the software so you can see in more detail what is really going on.
https://sleep.tnet.com/resources/sleepyhead

When you have time please add your equipment to your profile like I have mine.
Right now I know you have a ResMed machine because you mention MyAir and I know that means Resmed.

AHI going from 5 to 15...need to figure out for sure what is happening and deal with it. Most likely more pressure is needed but a more in depth evaluation is needed to know what to do and how to go about it.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

wrath
Posts: 7
Joined: Sun Aug 16, 2015 8:17 am

Re: new mask spiked events per hour?

Post by wrath » Sun Aug 16, 2015 10:34 am

Thank you for the link. I will download it & analyze the data when I have a bit more time.

wrath
Posts: 7
Joined: Sun Aug 16, 2015 8:17 am

Re: new mask spiked events per hour?

Post by wrath » Mon Aug 17, 2015 10:26 am

I have downloaded the software & data from the SD card. I don't know what I am looking at.

A screen shot of the stats is available, but I don't see the command to attach a file to this reply.

Thanks.

SewTired
Posts: 1737
Joined: Thu Apr 16, 2015 8:33 am
Location: Minneapolis area

Re: new mask spiked events per hour?

Post by SewTired » Mon Aug 17, 2015 10:30 am

Check the settings on your machine. Go to Mask and check whether it is set to pillows, full face or nasal. It changes the amount of pressure depending on which type of mask you are using. If this doesn't resolve the problem, you likely need a different mask. While you can see variations when you first switch, after NINE days, that tells me that the mask doesn't fit you very well.

_________________
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Additional Comments: Sleepyhead
Diabetes 2, RLS & bradycardia
Airsense For Her; Settings: range 8-12, Airfit P10 (M)

User avatar
Pugsy
Posts: 65120
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: new mask spiked events per hour?

Post by Pugsy » Mon Aug 17, 2015 10:42 am

The mask selection choice on the ResMed machines does NOT change the pressure on the ResMed machines.
Instead the mask selection choice mainly affects how the large leak is calculated.

How to post images of the software reports example here and with examples
viewtopic/t103468/Need-help-with-screen-shots.html
and addition instructions here...be sure to follow all the links
https://sleep.tnet.com/reference/tips/imgur

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

wrath
Posts: 7
Joined: Sun Aug 16, 2015 8:17 am

Re: new mask spiked events per hour?

Post by wrath » Mon Aug 17, 2015 11:26 am

Thanks SewTired. Just changed it from nasal to full face. Ran a mask fit & it said Good with a green happy face.
Will see how tonight goes.
I appreciate the help.

wrath
Posts: 7
Joined: Sun Aug 16, 2015 8:17 am

Re: new mask spiked events per hour?

Post by wrath » Mon Aug 17, 2015 11:45 am

Thanks Pugsy. This my first attempt at Imgur

[img][img]http://i.imgur.com/acMoWNFl.png[/img]/img]

[img][img]http://i.imgur.com/xiuiS9Xl.png[/img]/img]

this is the last night with old mask:

[img][img]http://i.imgur.com/xYN3eipl.png[/img]/img]

On the pie chart, I am seeing a big difference in OA, CA, & H.

I appreciate your patience.

User avatar
Krelvin
Posts: 1977
Joined: Tue Jun 06, 2006 5:23 pm
Location: Metro Phx Area - Dry Heat!

Re: new mask spiked events per hour?

Post by Krelvin » Mon Aug 17, 2015 2:19 pm

Your formatting is wrong.
Image
Image
this is the last night with old mask:
Image
Current Settings PS 4.0 over 10.6-18.0 (cmH2O) - Resmed S9 VPAP Auto w/h5i Humidifier - Quattro Air FFM
TNET Sleep Resource Pages - CPAP Machine Database
Put your equip in your Signature - SleepyHead v1.0.0-beta-1
Kevin... alias Krelvin

wrath
Posts: 7
Joined: Sun Aug 16, 2015 8:17 am

Re: new mask spiked events per hour?

Post by wrath » Tue Aug 18, 2015 8:02 am

Based on Pugsy's comments I went back to the nasal mask w/chin strap last night.
Wow, what a difference in the data:
OA: from 35 obstructive events to 0
CA: from about 22% to about 80%

thanks for the help, looks like the mask is the issue.