Respironics BiPap Pro Stops intermittently
Re: Respironics BiPap Pro Stops intermittently
have you tried popping the cover off and reseating the electrical and pneumatic connections?dbillow wrote:Well the intermittent stopping issue continued last night. I noticed that if I set the humidifier on 1 or 2 it happens less. I also noticed that the AC adapter was very warm this morning, almost hot. Could it be the motherboard going bad? Also, is there anyway to update the firmware on this unit? Thanks