Question AHI 0.0

Post by antimess » Thu Jun 25, 2015 5:35 am

I was diagnosed 6 months ago with OSA. I was and am still overweight, with a BMI greater than 30. Since being diagnosed, I've lost 15 lbs. Initially I would register an AHI between 1-2. Now when I read my data in Sleepyhead most nights I am registering an AHI of 0.0 (4-5 nights a week). Some nights it does go up to about 0.40. I am starting to wonder if the weight loss "cured" my OSA. I follow up with my sleep doctor this coming January and plan to continue to wear the mask nightly and focus on continued weight loss. Just curious what other more experienced individuals opinions are. Thanks for reading.

User avatar
OSAHell
Posts: 143
Joined: Wed Jun 17, 2015 8:56 am

Re: Question AHI 0.0

Post by OSAHell » Thu Jun 25, 2015 6:12 am

Hi antimess,

Congrats with your weight loss. Others might pinch in with their own experience and research but it's probably unlikely that weight loss cured your OSA, you'll have to go to an other sleep study to confirm that. From what I've read in some case loss weight can lower your OSA but rarely cure it. But it's far from perfect science and some people's OSA even worsen when they loss weight but that seem pretty rare. And to make things more interesting OSA might be the cause of weight gain for some people, lack off good sleep can make you more hungry!

What you can do is put some Sleepyhead graph here so we can see what's happening. You might be able to lower your pressure a bit, that can be more confortable, especially if you are on high pressure.

What's your pressure setting right now?

_________________
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: SleepyHead on Mac OSX, Resmed S9 VPAP Adapt (36037), EPAP 5 fixed, PS 4-10

User avatar
Julie
Posts: 20051
Joined: Tue Feb 28, 2006 12:58 pm

Re: Question AHI 0.0

Post by Julie » Thu Jun 25, 2015 6:16 am

Hi - you're obviously being well treated, and congrats on the wt loss!

However, you can't assume anything, OSA is rarely 'cured', but if you want to know if you still need Cpap, maybe weight a bit until you've lost even more wt and then have a new sleep study - it's the only valid way to know what's what. Having a low AHI is great, but all it means is that you're well treated, but doesn't address e.g. 02 desats (ever use a recording oximeter overnight?). And of course you could always go for at week without Cpap (I'm not saying you should...) and see how things go before having a new study done.

antimess
Posts: 10
Joined: Mon Dec 15, 2014 6:52 am
Location: Midwest

Re: Question AHI 0.0

Post by antimess » Thu Jun 25, 2015 6:18 am

I'll post some graphs tonight when I get home. Right now my pressure is wide open, 5.0-17.0. I would need to look again to be sure, but my pressure typically stays below 11 most nights.

I haven't used an o2 oximeter over night. I'll look into that too. Thanks!

User avatar
OSAHell
Posts: 143
Joined: Wed Jun 17, 2015 8:56 am

Re: Question AHI 0.0

Post by OSAHell » Thu Jun 25, 2015 6:36 am

Well, if your at 5-17 and your pressure "stays below 11". It probably mean that your pressure go near 11. If that the case you certainly still have OSA and like Julie says probably means that you're well treated. Now if you were saying that your pressure is flat line at 5, that would be an other story... But even then, even 5 or 4 of a pressure can treat some OSA. We'll check your graph tonight, make sure to look on the forum how to do this and what kind of infos are the most helpful to show (sorry don't have the link...).

_________________
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: SleepyHead on Mac OSX, Resmed S9 VPAP Adapt (36037), EPAP 5 fixed, PS 4-10

User avatar
Pugsy
Posts: 65261
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Question AHI 0.0

Post by Pugsy » Thu Jun 25, 2015 6:44 am

OSAHell wrote:Well, if your at 5-17 and your pressure "stays below 11". It probably mean that your pressure go near 11. If that the case you certainly still have OSA and like Julie says probably means that you're well treated. Now if you were saying that your pressure is flat line at 5, that would be an other story... But even then, even 5 or 4 of a pressure can treat some OSA. We'll check your graph tonight, make sure to look on the forum how to do this and what kind of infos are the most helpful to show (sorry don't have the link...).
Ditto to what OSAHell says. If the pressure is increasing then it is fighting something. If you didn't have OSA still going on it wouldn't have anything to fight and wouldn't be doing any changing.

Here's the link https://sleep.tnet.com/reference/tips/imgur
and a thread where there are examples of what we like to see
viewtopic/t103468/Need-help-with-screen-shots.html

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Question AHI 0.0

Post by palerider » Thu Jun 25, 2015 11:20 am

antimess wrote: Right now my pressure is wide open, 5.0-17.0. I would need to look again to be sure, but my pressure typically stays below 11 most nights.
wide open is 4-20, for what it's worth.

and, as pointed out, your 'pressure stays below 11' is very telling.

your situation here is much like saying "I put on these glasses and now I can see everything sharply! are my eyesight problems cured?"

no, they're not cured, they're just being treated effectively.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

antimess
Posts: 10
Joined: Mon Dec 15, 2014 6:52 am
Location: Midwest

Re: Question AHI 0.0

Post by antimess » Thu Jun 25, 2015 8:19 pm

Thanks for the information, that makes sense to me. For what it's worth here's one of my recent nights.
Image

User avatar
kaiasgram
Posts: 3569
Joined: Sat Jun 02, 2012 2:08 pm
Location: Northern California

Re: Question AHI 0.0

Post by kaiasgram » Thu Jun 25, 2015 11:37 pm

antimess, I'd be interested in seeing the Flow Limit and Snore graphs for that same night, posted right underneath the Pressure graph. ResMed machines respond pretty aggressively to snores and flow limitations. I don't have a snore line to speak of but I do get flow limitations and the lines in my FL graph and Pressure graph rise and fall together. If you stack those graphs and look at them you'll likely be able to see where/when your machine is doing its job.

_________________
Machine: AirSense 10 AutoSet with Heated Humidifer + Aifit N30i Nasal Mask Bundle
Additional Comments: SleepyHead-now-OSCAR software on Mac OSX Ventura

User avatar
Sheriff Buford
Posts: 4111
Joined: Mon Aug 09, 2010 8:01 am
Location: Kingwood, Texas

Re: Question AHI 0.0

Post by Sheriff Buford » Fri Jun 26, 2015 12:03 am

I get a zero ahi 2 or 3 times a month. Enjoy it... you are successfully treating your disease! I will say that I normally don't feel as well, but that's me.

Sheriff

User avatar
zoocrewphoto
Posts: 3732
Joined: Mon Apr 30, 2012 10:34 pm
Location: Seatac, WA

Re: Question AHI 0.0

Post by zoocrewphoto » Fri Jun 26, 2015 12:58 am

Keep in mind that the ahi you are seeing is your treated ahi. If your sleep apnea were cured, you would see an ahi of zero a pressure line that was straight across at the lowest setting. Your pressure does go up and down, so the machine is doing a great job of preventing events and keeping you well treated.

If you were to set your machine to a straight 5, you would see some events, and there would be even more if you were being monitored without using a cpap machine.

_________________
Mask: Quattro™ FX Full Face CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: Resmed S9 autoset pressure range 11-17
Who would have thought it would be this challenging to sleep and breathe at the same time?

User avatar
archangle
Posts: 9293
Joined: Sun Mar 27, 2011 11:55 am

Re: Question AHI 0.0

Post by archangle » Fri Jun 26, 2015 1:10 am

zoocrewphoto wrote:Keep in mind that the ahi you are seeing is your treated ahi. If your sleep apnea were cured, you would see an ahi of zero a pressure line that was straight across at the lowest setting. Your pressure does go up and down, so the machine is doing a great job of preventing events and keeping you well treated.

If you were to set your machine to a straight 5, you would see some events, and there would be even more if you were being monitored without using a cpap machine.
You know, it would be interesting to put a wide open auto machine on a "normal" person and see if it sticks at 4, or if it goes up and down a bit.

_________________
Mask: Swift™ FX Nasal Pillow CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: Also SleepyHead, PRS1 Auto, Respironics Auto M series, Legacy Auto, and Legacy Plus
Please enter your equipment in your profile so we can help you.
Click here for information on the most common alternative to CPAP.
If it's midnight and a DME tells you it's dark outside, go and check for yourself.

Useful Links.

User avatar
Pugsy
Posts: 65261
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Question AHI 0.0

Post by Pugsy » Fri Jun 26, 2015 5:08 am

archangle wrote:You know, it would be interesting to put a wide open auto machine on a "normal" person and see if it sticks at 4, or if it goes up and down a bit.
I got it almost done on my sister. A while back I tested my machine (was the M series back then) and while we couldn't use 4 cm because she said I was trying to kill her by suffocation and we had to use 6 cm starting point...other than the test pressure probes the pressure never moved off 6 cm.
I don't think she has OSA. Her main complaint was fatigue. She wore the mask 3 nights. While it wouldn't be impossible for her to have OSA that is totally eradicated with 6 cm pressure I doubt that happened.
Her fatigue went totally away with a change in blood pressure meds.
Now her husband (who has the typical snoring/gasping symptoms) on the other hand...his pressure line went to 9 cm and stayed there. He refused to follow up though. "Not going to wear one of those damn things all night"

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
OSAHell
Posts: 143
Joined: Wed Jun 17, 2015 8:56 am

Re: Question AHI 0.0

Post by OSAHell » Fri Jun 26, 2015 8:37 am

zoocrewphoto pretty much wrote what I was about to write this morning after a quick look at your graph last night. I don't see any point in changing anything now if most of your graph's nights look like this one.

_________________
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: SleepyHead on Mac OSX, Resmed S9 VPAP Adapt (36037), EPAP 5 fixed, PS 4-10