6 days at pressure 14

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
FranckDG
Posts: 41
Joined: Mon Apr 06, 2015 5:18 pm

6 days at pressure 14

Post by FranckDG » Thu May 07, 2015 11:25 am

I received my first CPAP machine last Friday (a brick). I should be receiving a brand new S9 Autoset today with nasal pillow mask and get rid of that brick.

Anyway... My experience so far with the brick locked at pressure 14 with full face mask has been as follow:
- Got used to the mask after 2-3 nights
- Woke up because of leaks when reaching pressure 14 almost every hour for the first 3 days then it got better (I learned how to better adjust my mask). Still waking up every 2-3 hours though...
- Once I got over these 2 issues, I started running into the bigger issue: aerophagia. I'm just getting inflated like a balloon all night and at first it was just bloating, burping and farting galore all morning but for the past 3 days it's also been explosive diarrhea from 5am (at which point I have to stop the machine) until about 10am. For reference, I typically hit the sack around 11:30pm so at 5am, I've been on the machine for about 5.5 hours.

Has anybody else here experienced this?

I am getting started on the Autoset tonight with a lower setting (I'll start with 8 with a special pillow to keep me on my side). I saw my doctor 2 days ago and she agreed that pressure 14 was probably too high for me (my AHI is 17.7) and that if I can stay on my side, I can probably use a lower setting.

For people who have experienced these symptoms, do they improve with a lower setting? Does the type of mask matter? I guess I'll find out soon but just asking...

Thanks!

User avatar
postitnote
Posts: 922
Joined: Sat Feb 14, 2015 4:36 pm

Re: 6 days at pressure 14

Post by postitnote » Thu May 07, 2015 1:10 pm

There has been a stomach virus going around. You sure you don't have that?
Morbius, are you bored?

User avatar
LSAT
Posts: 13395
Joined: Sun Nov 16, 2008 10:11 am
Location: SE Wisconsin

Re: 6 days at pressure 14

Post by LSAT » Thu May 07, 2015 4:01 pm

FranckDG wrote:I received my first CPAP machine last Friday (a brick). I should be receiving a brand new S9 Autoset today with nasal pillow mask and get rid of that brick.

Anyway... My experience so far with the brick locked at pressure 14 with full face mask has been as follow:
- Got used to the mask after 2-3 nights
- Woke up because of leaks when reaching pressure 14 almost every hour for the first 3 days then it got better (I learned how to better adjust my mask). Still waking up every 2-3 hours though...
- Once I got over these 2 issues, I started running into the bigger issue: aerophagia. I'm just getting inflated like a balloon all night and at first it was just bloating, burping and farting galore all morning but for the past 3 days it's also been explosive diarrhea from 5am (at which point I have to stop the machine) until about 10am. For reference, I typically hit the sack around 11:30pm so at 5am, I've been on the machine for about 5.5 hours.

Has anybody else here experienced this?

I am getting started on the Autoset tonight with a lower setting (I'll start with 8 with a special pillow to keep me on my side). I saw my doctor 2 days ago and she agreed that pressure 14 was probably too high for me (my AHI is 17.7) and that if I can stay on my side, I can probably use a lower setting.

For people who have experienced these symptoms, do they improve with a lower setting? Does the type of mask matter? I guess I'll find out soon but just asking...

Thanks!
If your pressure is set at 14, you should reach that pressure in a matter of minutes and it stays there. I don't understand when you say that " when the pressure reaches 14 you wake up because of leaks". If your AHI is 17 at a pressure of 14, normally it would go up at a lower pressure...Time will tell.

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: 6 days at pressure 14

Post by palerider » Thu May 07, 2015 6:15 pm

LSAT wrote: If your pressure is set at 14, you should reach that pressure in a matter of minutes and it stays there. I don't understand
ramp.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

FranckDG
Posts: 41
Joined: Mon Apr 06, 2015 5:18 pm

Re: 6 days at pressure 14

Post by FranckDG » Thu May 07, 2015 7:00 pm

postitnote wrote:There has been a stomach virus going around. You sure you don't have that?
I guess it's possible:) There's been something going around the office... Maybe it gets accentuated by the aerophagia symptoms

FranckDG
Posts: 41
Joined: Mon Apr 06, 2015 5:18 pm

Re: 6 days at pressure 14

Post by FranckDG » Thu May 07, 2015 7:01 pm

LSAT wrote:
FranckDG wrote:I received my first CPAP machine last Friday (a brick). I should be receiving a brand new S9 Autoset today with nasal pillow mask and get rid of that brick.

Anyway... My experience so far with the brick locked at pressure 14 with full face mask has been as follow:
- Got used to the mask after 2-3 nights
- Woke up because of leaks when reaching pressure 14 almost every hour for the first 3 days then it got better (I learned how to better adjust my mask). Still waking up every 2-3 hours though...
- Once I got over these 2 issues, I started running into the bigger issue: aerophagia. I'm just getting inflated like a balloon all night and at first it was just bloating, burping and farting galore all morning but for the past 3 days it's also been explosive diarrhea from 5am (at which point I have to stop the machine) until about 10am. For reference, I typically hit the sack around 11:30pm so at 5am, I've been on the machine for about 5.5 hours.

Has anybody else here experienced this?

I am getting started on the Autoset tonight with a lower setting (I'll start with 8 with a special pillow to keep me on my side). I saw my doctor 2 days ago and she agreed that pressure 14 was probably too high for me (my AHI is 17.7) and that if I can stay on my side, I can probably use a lower setting.

For people who have experienced these symptoms, do they improve with a lower setting? Does the type of mask matter? I guess I'll find out soon but just asking...

Thanks!
If your pressure is set at 14, you should reach that pressure in a matter of minutes and it stays there. I don't understand when you say that " when the pressure reaches 14 you wake up because of leaks". If your AHI is 17 at a pressure of 14, normally it would go up at a lower pressure...Time will tell.
The machine I had was set to 14 but you could press a button to set it to 4 and it would gradually get to 14 in 45 minutes at which point I would be asleep.

FranckDG
Posts: 41
Joined: Mon Apr 06, 2015 5:18 pm

Re: 6 days at pressure 14

Post by FranckDG » Thu May 07, 2015 7:02 pm

LSAT wrote:
FranckDG wrote:I received my first CPAP machine last Friday (a brick). I should be receiving a brand new S9 Autoset today with nasal pillow mask and get rid of that brick.

Anyway... My experience so far with the brick locked at pressure 14 with full face mask has been as follow:
- Got used to the mask after 2-3 nights
- Woke up because of leaks when reaching pressure 14 almost every hour for the first 3 days then it got better (I learned how to better adjust my mask). Still waking up every 2-3 hours though...
- Once I got over these 2 issues, I started running into the bigger issue: aerophagia. I'm just getting inflated like a balloon all night and at first it was just bloating, burping and farting galore all morning but for the past 3 days it's also been explosive diarrhea from 5am (at which point I have to stop the machine) until about 10am. For reference, I typically hit the sack around 11:30pm so at 5am, I've been on the machine for about 5.5 hours.

Has anybody else here experienced this?

I am getting started on the Autoset tonight with a lower setting (I'll start with 8 with a special pillow to keep me on my side). I saw my doctor 2 days ago and she agreed that pressure 14 was probably too high for me (my AHI is 17.7) and that if I can stay on my side, I can probably use a lower setting.

For people who have experienced these symptoms, do they improve with a lower setting? Does the type of mask matter? I guess I'll find out soon but just asking...

Thanks!
If your pressure is set at 14, you should reach that pressure in a matter of minutes and it stays there. I don't understand when you say that " when the pressure reaches 14 you wake up because of leaks". If your AHI is 17 at a pressure of 14, normally it would go up at a lower pressure...Time will tell.
My AHI is at 17.7 with no CPAP machine.

User avatar
Papit
Posts: 844
Joined: Sun Dec 18, 2011 4:47 pm
Location: USA CA

Re: 6 days at pressure 14

Post by Papit » Thu May 07, 2015 7:06 pm

FranckDG wrote:. . .For people who have experienced these symptoms, do they improve with a lower setting? . . .Thanks!
You bet! You should feel a lot better on the lower setting. I'm surprised your DME didn't reset the machine at lower pressure based on your last visit with your doc.

_________________
Mask: DreamWear Nasal CPAP Mask with Headgear
Additional Comments: Machine: AirCurve 10 ASV (37043), Software:ResScan 5.7.0.9477, SleepyHead V1.00BETA2, Oximeter:CMS-50i
Machine: AirCurve 10 ASV, Mask: AirFit N30i
PulseOx Data-transfer to OSCAR4-23-12http://tinyurl.com/nzd64gu
Wireless SD Card Data-transfer to OSCAR 8-14-15http://tiny.cc/z1kv8x

User avatar
BlackSpinner
Posts: 9742
Joined: Sat Apr 25, 2009 5:44 pm
Location: Edmonton Alberta
Contact:

Re: 6 days at pressure 14

Post by BlackSpinner » Fri May 08, 2015 10:24 am

Code: Select all

 I saw my doctor 2 days ago and she agreed that pressure 14 was probably too high for me (my AHI is 17.7) and that if I can stay on my side, I can probably use a lower setting.
Just for your information - AHI and pressure have no relationship with each other. AHI is just an average of the number of times your breathing stops, or tries to, per hour. The pressure you need to stop it from doing that depends on your physical needs. You can have an AHI of 70+ and still only need a pressure of 5 - 7.

_________________
Machine: PR System One REMStar 60 Series Auto CPAP Machine
Additional Comments: Quatro mask for colds & flus S8 elite for back up
71. The lame can ride on horseback, the one-handed drive cattle. The deaf, fight and be useful. To be blind is better than to be burnt on the pyre. No one gets good from a corpse. The Havamal

cranblarry
Posts: 3
Joined: Thu Aug 23, 2018 9:10 am

Re: 6 days at pressure 14

Post by cranblarry » Thu Aug 23, 2018 9:55 am

My husband is a new CPAP user. He developed diarrhea about 4 days in. As an experiment, he has not used his mask for 6 days and the diarrhea has not stopped. I read somewhere that the esophageal sphincter can get irritated and not function properly, letting air in. Considering the dental device, and starting our second month of Singing for Snorers.