my spo2 level decrease dramatically when I raise my arm

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
gemsergio
Posts: 7
Joined: Sun Apr 26, 2015 11:46 am

my spo2 level decrease dramatically when I raise my arm

Post by gemsergio » Sun May 03, 2015 12:31 am

Even when I have my oxymeter at 98 when I am laying down and raise my arm it drops rapidly to 85 even lower, does it mean I have bad circulation? Is it normal??
As soon as I lower my arm the spo2 level goes back to normal 96/98 readings. I am not on a CPAP machine.

gemsergio
Posts: 7
Joined: Sun Apr 26, 2015 11:46 am

Re: my spo2 level decrease dramatically when I raise my arm

Post by gemsergio » Sun May 03, 2015 12:57 am

I forgot to mention that it drops only in the arm that is lifted the other arm apo2 level stays the same

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: my spo2 level decrease dramatically when I raise my arm

Post by palerider » Sun May 03, 2015 1:07 am

gemsergio wrote:Even when I have my oxymeter at 98 when I am laying down and raise my arm it drops rapidly to 85 even lower, does it mean I have bad circulation? Is it normal??
As soon as I lower my arm the spo2 level goes back to normal 96/98 readings. I am not on a CPAP machine.
I don't believe that's normal, but I'll have to check with my oximeter when I get it back from the person I loaned it out to.
I can't answer what that means, though your idea sounds reasonable.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
mralaska
Posts: 63
Joined: Mon Apr 13, 2015 11:45 am

Re: my spo2 level decrease dramatically when I raise my arm

Post by mralaska » Sun May 03, 2015 5:04 am

I would be curious myself whether it could indicate a condition to look into.

I just tried it and my SpO2 reading made no significant change. I did have trouble typing with it on but I think that might be normal.

_________________
Mask: Swift™ FX Nasal Pillow CPAP Mask with Headgear
Additional Comments: PRS1 760 (BiPap Auto set up as APAP)
Happily hosing since 4/19/2015
Do not believe a word I say until Pugsy or PaleRider has had a chance to review!

Cheers,
Paul

User avatar
chunkyfrog
Posts: 34544
Joined: Mon Jul 12, 2010 5:10 pm
Location: Nowhere special--this year in particular.

Re: my spo2 level decrease dramatically when I raise my arm

Post by chunkyfrog » Sun May 03, 2015 9:42 am

I tried it with mine, and no difference with either hand, up or down, all the same.
Are you at risk for lymphedema?

_________________
Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear
Additional Comments: Airsense 10 Autoset for Her

User avatar
Pugsy
Posts: 65194
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: my spo2 level decrease dramatically when I raise my arm

Post by Pugsy » Sun May 03, 2015 9:48 am

I would take the pulse ox to the doctor's office and have them check the spo2 with their machine and arm raised and lowered and at the same time compare it to what my pulse ox showed doing the same thing.

It doesn't sound normal to me and if it happens at the doctors office with their equipment then you will be right there to talk to the doctor about it.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
BlackSpinner
Posts: 9742
Joined: Sat Apr 25, 2009 5:44 pm
Location: Edmonton Alberta
Contact:

Re: my spo2 level decrease dramatically when I raise my arm

Post by BlackSpinner » Sun May 03, 2015 10:22 am

Assuming the meter works, try it on someone else, then get to your doctor asap and get a referral to a cardiologist.
This kind of thing happens to heart patients and it is why they are not allowed to lift their arms over their heads.

_________________
Machine: PR System One REMStar 60 Series Auto CPAP Machine
Additional Comments: Quatro mask for colds & flus S8 elite for back up
71. The lame can ride on horseback, the one-handed drive cattle. The deaf, fight and be useful. To be blind is better than to be burnt on the pyre. No one gets good from a corpse. The Havamal

User avatar
Krelvin
Posts: 1977
Joined: Tue Jun 06, 2006 5:23 pm
Location: Metro Phx Area - Dry Heat!

Re: my spo2 level decrease dramatically when I raise my arm

Post by Krelvin » Sun May 03, 2015 10:30 am

BlackSpinner wrote:This kind of thing happens to heart patients and it is why they are not allowed to lift their arms over their heads.
CHF patient... I asked once why while in the hospital they didn't want me putting my arm over my head and it was because it could mess up the readings on the sensor, not because it was bad for my heart or circulation. More for monitoring reasons than for health reasons.

While it might also be an example of a possible medical condition that needs to be looked at, it by itself should not cause a medical issue.
Current Settings PS 4.0 over 10.6-18.0 (cmH2O) - Resmed S9 VPAP Auto w/h5i Humidifier - Quattro Air FFM
TNET Sleep Resource Pages - CPAP Machine Database
Put your equip in your Signature - SleepyHead v1.0.0-beta-1
Kevin... alias Krelvin

gemsergio
Posts: 7
Joined: Sun Apr 26, 2015 11:46 am

Re: my spo2 level decrease dramatically when I raise my arm

Post by gemsergio » Sun May 03, 2015 10:51 am

I think its mainly that there is not enough oxygen pumped to the extremities, I have no sign of lymphedema, the oxymeter seems to work since the other hand stays at 97 while the other one goes down to 87 for example.I noticed a couple of points difference as well with my wife but nothing major and as fast as me.

HoseCrusher
Posts: 2744
Joined: Tue Oct 12, 2010 6:42 pm

Re: my spo2 level decrease dramatically when I raise my arm

Post by HoseCrusher » Sun May 03, 2015 11:26 am

When you raise your arm up the pulse signal in your finger is reduced. This throws the oximeter readings off as it struggles to increase amplification to regain the weaker pulse signal. This is usually a step process and can take several minutes to normalize.

If you raise your arm up and hold it there for something like 10 minutes you should see the SpO2 readings come in close to what they were with your arm laying next to you. It may be a couple of points lower but should be close.

The main variables that govern how fast the oximeter returns to normal readings include blood pressure, resting pulse rate, and the aggressiveness of the algorithm in trying to recover from the reduced pulse signal.

_________________
Mask: Brevida™ Nasal Pillow CPAP Mask with Headgear
Additional Comments: Machine is an AirSense 10 AutoSet For Her with Heated Humidifier.
SpO2 96+% and holding...

borgready
Posts: 267
Joined: Mon Sep 27, 2010 2:41 pm
Location: VA, USA

Re: my spo2 level decrease dramatically when I raise my arm

Post by borgready » Sun May 03, 2015 11:49 am

If you hold your arm up, how long does it take for it to go numb? Going numb is a good sign of reduced blood flow.

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: my spo2 level decrease dramatically when I raise my arm

Post by palerider » Sun May 03, 2015 12:49 pm

borgready wrote:If you hold your arm up, how long does it take for it to go numb? Going numb is a good sign of reduced blood flow.
citation please... because the typical cause of going "numb" is a NERVE issue.

now, if OP's hand starts appearing cyanotic, that would be a blood issue.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

borgready
Posts: 267
Joined: Mon Sep 27, 2010 2:41 pm
Location: VA, USA

Re: my spo2 level decrease dramatically when I raise my arm

Post by borgready » Mon May 04, 2015 10:18 am

palerider wrote:
borgready wrote:If you hold your arm up, how long does it take for it to go numb? Going numb is a good sign of reduced blood flow.
citation please... because the typical cause of going "numb" is a NERVE issue.

now, if OP's hand starts appearing cyanotic, that would be a blood issue.
Numbness from nerve issues are more permanent and numbness from low or no blood flow issues are temporary as long as blood flow is restored. Cheap oximeters don't read low blood flow very well. My pulse oximeter goes out all the time when my finger or hand goes numb. The readings will drop off. My citation comes from the LIFE"S EXPERIENCE JOURNAL.

PALERIDER have you been thinking about which of my theories pisses you off the most. Its up for discussion.

User avatar
Krelvin
Posts: 1977
Joined: Tue Jun 06, 2006 5:23 pm
Location: Metro Phx Area - Dry Heat!

Re: my spo2 level decrease dramatically when I raise my arm

Post by Krelvin » Mon May 04, 2015 10:25 am

http://www.abc.net.au/health/talkinghea ... 513662.htm

The "nurse" who was telling me to lower my arms didn't know why other than it was "bad for my heart" when I asked why. I then asked my cardiologist when he showed up later that day, he laughed said it had nothing to do with the heart, more with what it did with monitoring the O2 levels. BTW, I only had one arm raised it wasn't the one with the sensor on it.

I like playing with the O2 sensor reading when in the hospital and have nothing else to do. I can concentrate and make the readings go up and down. At 88%, the alarm goes off and the nurse comes running. BUT if the alarm goes off on the CPAP nobody shows up. funny.
Current Settings PS 4.0 over 10.6-18.0 (cmH2O) - Resmed S9 VPAP Auto w/h5i Humidifier - Quattro Air FFM
TNET Sleep Resource Pages - CPAP Machine Database
Put your equip in your Signature - SleepyHead v1.0.0-beta-1
Kevin... alias Krelvin

HoseCrusher
Posts: 2744
Joined: Tue Oct 12, 2010 6:42 pm

Re: my spo2 level decrease dramatically when I raise my arm

Post by HoseCrusher » Mon May 04, 2015 10:48 am

The "obvious" solution is that when your CPAP alarm goes off, raise your arm...

_________________
Mask: Brevida™ Nasal Pillow CPAP Mask with Headgear
Additional Comments: Machine is an AirSense 10 AutoSet For Her with Heated Humidifier.
SpO2 96+% and holding...