when connecting the humidifier it has came up with a fault regarding the power pack main unit checked and ok so the ticking is coming from the humidifier rendering it useless when connected to main unit It works without the humidifier
any ideas as it has been working fine until now
s9 humidifier ticking
Re: s9 humidifier ticking
Can you take it to where you got it (both machine and humidifier) so that you can
1..try the blower with a new humidifier
2..try that humidifier with a different blower
At this point there's no way to know if the fault is in the circuit in the blower or the fault is in the humidifier and while you could just buy another humidifier to check it out ...you might end up with 2 humidifiers and the problem be in the blower circuitry somehow.
1..try the blower with a new humidifier
2..try that humidifier with a different blower
At this point there's no way to know if the fault is in the circuit in the blower or the fault is in the humidifier and while you could just buy another humidifier to check it out ...you might end up with 2 humidifiers and the problem be in the blower circuitry somehow.
_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/ |
I may have to RISE but I refuse to SHINE.
Re: s9 humidifier ticking
i am using it without the humidifier and it works ok so the problem is as soon as i connect the humidifier
i have sent the unit back but they say that the humidifier would be unrepairable so i will persist without the humidifier for the time being
and maybe invest in a new s10 unit in the near future
just wondered if anyone has already had this problem as res med say they have not
i have sent the unit back but they say that the humidifier would be unrepairable so i will persist without the humidifier for the time being
and maybe invest in a new s10 unit in the near future
just wondered if anyone has already had this problem as res med say they have not
Re: s9 humidifier ticking
you can probably pick up a spare humidifier at secondwindcpap for less $$$Guest wrote:i am using it without the humidifier and it works ok so the problem is as soon as i connect the humidifier
i have sent the unit back but they say that the humidifier would be unrepairable so i will persist without the humidifier for the time being
and maybe invest in a new s10 unit in the near future
just wondered if anyone has already had this problem as res med say they have not
Get OSCAR
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
- chunkyfrog
- Posts: 34545
- Joined: Mon Jul 12, 2010 5:10 pm
- Location: Nowhere special--this year in particular.
Re: s9 humidifier ticking
Do what Pugsy said. The problem could still be in either module.
_________________
Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear |
Additional Comments: Airsense 10 Autoset for Her |