Bipap not working/ fixed it/ Thx forum

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
User avatar
Pugsy
Posts: 65105
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Bipap not working. working as cpap...HELP

Post by Pugsy » Fri Apr 24, 2015 7:45 am

yaconsult wrote:I'm starting to wonder if I should check that wire on my low hour 550P as opposed to waiting for it to fail? Do we have any idea how common this problem is? If rare, I wouldn't bother. If it happens 50% of the time or more, it would probably be worth it.
While not extremely rare (unless you are one of those it happens to and then it doesn't seem so rare )...I don't think it comes anywhere near the 50% category.
I would be of the opinion to go by the "if it ain't broke don't fix it" rule.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Bipap not working. working as cpap...HELP

Post by palerider » Fri Apr 24, 2015 8:52 am

yaconsult wrote:Nice job figuring it out! So the machine wasn't able to sense what was happening properly - makes sense.

I always advise when doing jobs like this - grab your phone and take lots of pictures as you take things apart. If you don't need them, they are easy to delete. If you do need them, they can be invaluable!

I'm starting to wonder if I should check that wire on my low hour 550P as opposed to waiting for it to fail? Do we have any idea how common this problem is? If rare, I wouldn't bother. If it happens 50% of the time or more, it would probably be worth it. I have lots of experience with electronics so I'm comfortable working on stuff.
it's so easy to pop the cover of the sysone machines, I say, just do it... long shank t20, two screws, gently pull off one knob, and you're in. I'd unplug and replug all the connectors, maybe even take a pencil eraser to the posts to clean them...

call it 'preventative maintenance'.

there's no telling how prevalent this issue is, but it keeps getting repeated.... how many people simply replace their units without ever finding out that's what's wrong?

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

mubiksski
Posts: 8
Joined: Sun Mar 29, 2015 9:40 am

Re: Bipap not working. working as cpap...HELP

Post by mubiksski » Fri Apr 24, 2015 7:49 pm

My Machine was about 4.5 years old. Used most every night. But why fix it if it aint' broken? It is an easy repair thanks to this forum. I would wait. and pictures, which I do as well wouldn't have helped here. It was nothing you could see. I am just glad I got it working. I can't really afford a new one at this point.

JDS74
Posts: 3397
Joined: Sun Jan 23, 2011 2:57 pm
Location: South Carolina

Re: Bipap not working. working as cpap...HELP

Post by JDS74 » Sat Apr 25, 2015 6:42 am

SGearhart wrote:Quick question. You stated you have a P-R 750, I thought the BiPAP model was a 760. With model is the BiPAP?
In case this hasn't been answered before:

The 750 is the 50 series Auto BiPap
The 760 is the 60 series Auto BiPap

They are very similar machines. The 60 series can accept a heated tube heated humidifier while the 50 series cannot.
There are likely other minor differences between the machines but that is the big one.

_________________
Mask: Oracle HC452 Oral CPAP Mask
Humidifier: DreamStation Heated Humidifier
Additional Comments: EverFlo Q 3.0 Liters O2 PR DSX900 ASV
Oracle 452 Lessons Learned Updated
DSX900 AutoSV with HC150 extra humidifier and Hibernite heated hose
Settings: EPAP Min-10.0, EPAP Max-17, PS Min-3, PS Max-10, Max Pressure-20, Rate-Auto, Biflex-1.
Sleepyhead and Encore Pro 2.21.

User avatar
Sir NoddinOff
Posts: 4190
Joined: Mon May 14, 2012 5:30 pm
Location: California

Re: Bipap not working/ fixed it/ Thx forum

Post by Sir NoddinOff » Sat Apr 25, 2015 9:25 am

Glad things worked out. Just a cautionary note on soldering ANYTHING in close proximity or adjacent to a printed circuit board - especially the board connector to various power sources (as in this case): First be sure to clamp an alligator clip (see link below), a hemostat (roach holder, for those with MJ experience) OR one or two of small needle-nose pliers (w/ a sturdy rubber band around the handles) around the metal area to be soldered. Also make them as close to the area soldered as reasonably possible. This trick will provide a heat sink and keep the heat from the soldering iron from going down into the circuit board and damaging the delicate printed wiring.

http://www.instructables.com/id/Save-yo ... heat-sink/

_________________
Mask: AirFit™ F10 Full Face Mask with Headgear
Additional Comments: Sleepyhead software v.0.9.8.1 Open GL and Encore Pro v2.2.
I like my ResMed AirFit F10 FFM - reasonably low leaks for my ASV therapy. I'm currently using a PR S1 AutoSV 960P Advanced. I also keep a ResMed S9 Adapt as backup. I use a heated Hibernite hose. Still rockin' with Win 7 by using GWX to stop Win 10.