Finally starting CPAP with a few questions....

Post by Twelve12 » Wed Apr 08, 2015 1:48 pm

Today is the day I received CPAP machine/mask (added to signature). My path to this point has been as follows: emailed doctor that I would like to proceed with sleep study per small discussion I initiated about suspecting OSA; home sleep study done early Mar; results received late Mar - just quick message from my doctor that I do have "moderate sleep apnea" and she referred me to DME; spoke with DME late Mar and chose machine I wanted; appt today in which machine received and some basic sleep study info.

The sleep study indicates AHI is 26. DME said CPAP pressure set for 7. I asked about adjusting settings and she said I can't do it and that they will if it needs to be done. I asked how 7 is the number arrived at for pressure. She didn't really seem to know.

In general, I think the DME wasn't greatest with info. Honestly, I am not convinced she knows all that much about OSA, CPAP, etc. I have a lot to learn and figure out. I looked up how I can change machine settings (even though she said I can't) and found that info. I just want the ability to do this - and I want to know how/why the fixed pressure is decided if I did not have titration study. Also, when I was learning about the choice of machines, I thought an advantage to the auto machine is that it will adjust a bit through out the night between set parameters but I don't know that this is the case with a fixed setting. So -- should I start out using CPAP as is with fixed pressure of 7? I was thinking maybe 7 is just a standard setting and they go up or down from there after seeing data.

User avatar
sc0ttt
Posts: 396
Joined: Sat Apr 26, 2014 10:05 am
Location: East Bay San Francisco

Re: Finally starting CPAP with a few questions....

Post by sc0ttt » Wed Apr 08, 2015 1:56 pm

I thought an advantage to the auto machine is that it will adjust a bit through out the night between set parameters but I don't know that this is the case with a fixed setting. So -- should I start out using CPAP as is with fixed pressure of 7? I was thinking maybe 7 is just a standard setting and they go up or down from there after seeing data.
Yeah, no reason not to put it in AUTO ... might be a reason to adjust the min and max after a while.

Might also be a reason not to get in a fight with your DME if they're dealing with your insurance company... practice your response for when they ask you about changing your settings.

_________________
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Additional Comments: 14-cm wg, Model 460, Serial Number: P10175579 4E22 SleepyHead v0.9.6 (testing)

User avatar
Julie
Posts: 20051
Joined: Tue Feb 28, 2006 12:58 pm

Re: Finally starting CPAP with a few questions....

Post by Julie » Wed Apr 08, 2015 1:58 pm

Once it's in auto mode, you could set a high pressure of e.g. 12 to see how things go for a while - I hope you're using SleepyHead to monitor them and give you something to go by if you decide to make changes.

User avatar
Twelve12
Posts: 17
Joined: Sun Mar 29, 2015 5:29 am

Re: Finally starting CPAP with a few questions....

Post by Twelve12 » Wed Apr 08, 2015 2:09 pm

sc0ttt wrote:
I thought an advantage to the auto machine is that it will adjust a bit through out the night between set parameters but I don't know that this is the case with a fixed setting. So -- should I start out using CPAP as is with fixed pressure of 7? I was thinking maybe 7 is just a standard setting and they go up or down from there after seeing data.
Yeah, no reason not to put it in AUTO ... might be a reason to adjust the min and max after a while.

Might also be a reason not to get in a fight with your DME if they're dealing with your insurance company... practice your response for when they ask you about changing your settings.
Yeah, I am not going to change just for sake of changing.

I don't have any insurance requirements/monitoring, but DME will know I have changed settings.

User avatar
Twelve12
Posts: 17
Joined: Sun Mar 29, 2015 5:29 am

Re: Finally starting CPAP with a few questions....

Post by Twelve12 » Wed Apr 08, 2015 2:15 pm

Julie wrote:Once it's in auto mode, you could set a high pressure of e.g. 12 to see how things go for a while - I hope you're using SleepyHead to monitor them and give you something to go by if you decide to make changes.
That is my plan - to use SleepyHead to monitor and then figure out adjustments.

I am curious how my pressure was determined. I think 7 must be standard. Dunno. Though I won't change things just to change them, I would like optimal settings. That will take some time based on data after using. I guess this is the downside of not have titration study.

User avatar
sc0ttt
Posts: 396
Joined: Sat Apr 26, 2014 10:05 am
Location: East Bay San Francisco

Re: Finally starting CPAP with a few questions....

Post by sc0ttt » Wed Apr 08, 2015 2:45 pm

Twelve12 wrote:
Julie wrote: I am curious how my pressure was determined. I think 7 must be standard. Dunno.
Your home test was probably done with an auto machine that determined 7 was the right setting. Just starting everyone with 7cm is such a bad idea I don't think anyone would propose it.

But again, there's no reason NOT to put your machine in auto - go ahead. If 7 constant is your right number, the machine will provide it.

_________________
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Additional Comments: 14-cm wg, Model 460, Serial Number: P10175579 4E22 SleepyHead v0.9.6 (testing)

User avatar
Julie
Posts: 20051
Joined: Tue Feb 28, 2006 12:58 pm

Re: Finally starting CPAP with a few questions....

Post by Julie » Wed Apr 08, 2015 3:04 pm

FYI, I think 10 is more 'popular', but each to their own.

User avatar
ChicagoGranny
Posts: 15516
Joined: Sun Jan 29, 2012 1:43 pm
Location: USA

Re: Finally starting CPAP with a few questions....

Post by ChicagoGranny » Wed Apr 08, 2015 4:55 pm

Twelve12 wrote: I asked how 7 is the number arrived at for pressure. She didn't really seem to know.
This is a question for the doctor or his staff, not the DME. The DME just gets a prescription for the machine and pressure setting from the doc. Often DMEs don't have a clue - they just provide the model machine/pressure setting that the doctor wrote on the prescription.

You don't mention having a titration study (http://www.sleepeducation.com/treatment ... y/overview). So maybe the doctor did just take a starting guess at the pressure setting?
Julie wrote:Once it's in auto mode, you could set a high pressure of e.g. 12
You must mean "max pressure setting"? Don't confuse the beginners Julie.
"It's not the number of breaths we take, it's the number of moments that take our breath away."

Cuando cuentes cuentos, cuenta cuántas cuentos cuentas.

User avatar
palerider
Posts: 32300
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Finally starting CPAP with a few questions....

Post by palerider » Wed Apr 08, 2015 4:59 pm

Twelve12 wrote:[ but DME will know I have changed settings.
so? it's none of their business what you do with your machine, since you're not subject to insurance oversight.