AirSense 10 units won't run on an inverter? Screwed again?

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
theaftermath
Posts: 5
Joined: Sat Nov 03, 2018 6:18 pm

Re: AirSense 10 units won't run on an inverter? Screwed again?

Post by theaftermath » Sat Nov 03, 2018 6:55 pm

I must be missing something. The DC-DC converters for these are terribly inefficient aren't they? Input is 12v/24v 13A/6.5A with an output of 24v 3.75A!!

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: AirSense 10 units won't run on an inverter? Screwed again?

Post by palerider » Sat Nov 03, 2018 8:18 pm

theaftermath wrote:
Sat Nov 03, 2018 6:55 pm
I must be missing something. The DC-DC converters for these are terribly inefficient aren't they? Input is 12v/24v 13A/6.5A with an output of 24v 3.75A!!
No, they're not. they're very efficient (> 95%)

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
chunkyfrog
Posts: 34545
Joined: Mon Jul 12, 2010 5:10 pm
Location: Nowhere special--this year in particular.

Re: AirSense 10 units won't run on an inverter? Screwed again?

Post by chunkyfrog » Sat Nov 03, 2018 10:49 pm

Electricity--it's all about math.
Basic electrical equations explain it all.
Knowledge is power--especially knowledge of power. :lol:

_________________
Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear
Additional Comments: Airsense 10 Autoset for Her

theaftermath
Posts: 5
Joined: Sat Nov 03, 2018 6:18 pm

Re: AirSense 10 units won't run on an inverter? Screwed again?

Post by theaftermath » Sun Nov 04, 2018 3:50 pm

I guess the reason I'm a bit stumped is because hooking this up to a Killawatt meter it shows about 300wh for about 8hrs of use with humidity and tube turned on but it won't last 6 hours using this converter with 70ah battery? Is the converter going bad?

Snoregone Conclusion
Posts: 210
Joined: Thu Jun 07, 2018 10:28 pm
Location: Issaquah, WA

Re: AirSense 10 units won't run on an inverter? Screwed again?

Post by Snoregone Conclusion » Sun Nov 04, 2018 4:10 pm

theaftermath wrote:
Sun Nov 04, 2018 3:50 pm
I guess the reason I'm a bit stumped is because hooking this up to a Killawatt meter it shows about 300wh for about 8hrs of use with humidity and tube turned on but it won't last 6 hours using this converter with 70ah battery? Is the converter going bad?
Ah is not the same as wh: power (watts) is current (amps) multiplied by voltage, amps only measures current.

_________________
Machine
Additional Comments: 3B Elara FFM
Sleep, sleep monster, sleep!

prodigyplace
Posts: 1795
Joined: Mon Mar 27, 2017 5:45 pm
Location: Central Virginia

Re: AirSense 10 units won't run on an inverter? Screwed again?

Post by prodigyplace » Sun Nov 04, 2018 4:15 pm

Here is the Resmed battery guide that assumes you are using their inverter to convert 12 V DC to 24 V DC.

https://www.resmed.com/us/dam/documents ... lo_eng.pdf

The amount of current / power used depends on a number of factors. Mask pressure, humidity level, and use of heated hose can all affect how long a battery will last.

_________________
Machine: AirSense 11 Autoset
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Additional Comments: ClimateLineAir Heated Tube & Sleepyhead software
Please visit our sponsor, CPAP.com at https://www.cpap.com/ for all your CPAP needs.

theaftermath
Posts: 5
Joined: Sat Nov 03, 2018 6:18 pm

Re: AirSense 10 units won't run on an inverter? Screwed again?

Post by theaftermath » Sun Nov 04, 2018 4:43 pm

Snoregone Conclusion wrote:
Sun Nov 04, 2018 4:10 pm
theaftermath wrote:
Sun Nov 04, 2018 3:50 pm
I guess the reason I'm a bit stumped is because hooking this up to a Killawatt meter it shows about 300wh for about 8hrs of use with humidity and tube turned on but it won't last 6 hours using this converter with 70ah battery? Is the converter going bad?
Ah is not the same as wh: power (watts) is current (amps) multiplied by voltage, amps only measures current.
I do understand they are different, just showing my results. When plugged into AC the results showed that the machine would easily run off of that same 70ah setup that wouldn't run for 6 hours with the DC-DC converter.

theaftermath
Posts: 5
Joined: Sat Nov 03, 2018 6:18 pm

Re: AirSense 10 units won't run on an inverter? Screwed again?

Post by theaftermath » Sun Nov 04, 2018 4:55 pm

prodigyplace wrote:
Sun Nov 04, 2018 4:15 pm
Here is the Resmed battery guide that assumes you are using their inverter to convert 12 V DC to 24 V DC.

https://www.resmed.com/us/dam/documents ... lo_eng.pdf

The amount of current / power used depends on a number of factors. Mask pressure, humidity level, and use of heated hose can all affect how long a battery will last.
I did see this and it suggests a 45ah battery for the settings used on the machine but from the kw meter results when plugged into AC compared to not being able to get a full night out of a 70ah battery when using their DC converter something is off?

theaftermath
Posts: 5
Joined: Sat Nov 03, 2018 6:18 pm

Re: AirSense 10 units won't run on an inverter? Screwed again?

Post by theaftermath » Sun Nov 04, 2018 5:07 pm

prodigyplace wrote:
Sun Nov 04, 2018 4:15 pm
Here is the Resmed battery guide that assumes you are using their inverter to convert 12 V DC to 24 V DC.

https://www.resmed.com/us/dam/documents ... lo_eng.pdf

The amount of current / power used depends on a number of factors. Mask pressure, humidity level, and use of heated hose can all affect how long a battery will last.
The settings of the machine were the same but I don't know about surrounding air humidity or temperature differences.

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: AirSense 10 units won't run on an inverter? Screwed again?

Post by palerider » Sun Nov 04, 2018 6:33 pm

theaftermath wrote:
Sun Nov 04, 2018 3:50 pm
I guess the reason I'm a bit stumped is because hooking this up to a Killawatt meter it shows about 300wh for about 8hrs of use with humidity and tube turned on but it won't last 6 hours using this converter with 70ah battery? Is the converter going bad?
More likely your battery is bad,

You can get five nights with a resmed converter and a 35ah battery... *without humidification*.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
CapnLoki
Posts: 1162
Joined: Tue Jun 25, 2013 6:26 pm
Location: North East

Re: AirSense 10 units won't run on an inverter? Screwed again?

Post by CapnLoki » Mon Nov 05, 2018 1:57 pm

palerider wrote:
Sun Nov 04, 2018 6:33 pm
theaftermath wrote:
Sun Nov 04, 2018 3:50 pm
I guess the reason I'm a bit stumped is because hooking this up to a Killawatt meter it shows about 300wh for about 8hrs of use with humidity and tube turned on but it won't last 6 hours using this converter with 70ah battery? Is the converter going bad?
More likely your battery is bad,
I agree - a 70ah battery should have about 840 watt-hours. The testing with my Philips 560 was around 20ah with moderate humidity/hose settiings. The ResMed is a bit less efficient.
palerider wrote:
Sun Nov 04, 2018 6:33 pm
You can get five nights with a resmed converter and a 35ah battery... *without humidification*.
Five nights might be a stretch - that's 7ah a night with no reserve.

_________________
Machine: DreamStation Auto CPAP Machine
Mask: Quattro™ Air Full Face Mask with Headgear
Humidifier: DreamStation Heated Humidifier
Additional Comments: Pressure 9-20, average ~9.5; often use battery power while off-grid
Hark, how hard he fetches breath . . .  Act II, Scene IV, King Henry IV Part I, William Shakespeare
Choosing a Battery thread: http://www.cpaptalk.com/viewtopic/t1140 ... ttery.html