Contec CMS50I Pulse Oximeter

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
User avatar
coolbranch
Posts: 172
Joined: Sat Jul 12, 2014 9:59 am
Location: SE Alabama

Contec CMS50I Pulse Oximeter

Post by coolbranch » Wed Mar 04, 2015 6:02 pm

I am attempting to upload last nights recorded data to Sleepyhead. I received the CMS50I yesterday. I should be able to import from SPO2 software but it is giving me a parsing error for last night's recorded data.

I would prefer to import directly from the device in to Sleepyhead, but the CMS50I does not have an upload selection in the menu.

I would be very thankful if someone has this pulse oximeter and can help.

_________________
Machine: AirSense 11 Autoset
Mask: ResMed AirFit F30i Full Face CPAP Mask with Headgear
Additional Comments: OSCAR v1.4.0
I LOVE OSCAR

User avatar
M'ohms
Posts: 516
Joined: Fri Aug 09, 2013 2:44 pm
Location: Georgetown, Indiana

Re: Contec CMS50I Pulse Oximeter

Post by M'ohms » Wed Mar 04, 2015 6:07 pm

I have the same pulse-ox as you, and I have no problems loading the data directly into sleepyhead from the pulse-ox. After you upload you overnight data from your xPap, you can import your pulse-ox data from Sleepyhead by clicking the oximeter icon in the right hand column. This opens a page that asks you how you want to import the data. Choose the "import from device" option.

_________________
Mask: Quattro™ FX Full Face CPAP Mask with Headgear
Additional Comments: Supplemental oxygen at 2.5 lpm
AutoSV 960 with heated hose. Settings: EPAP Min-12, EPAP Max-17, PS Min-5.5, PS Max-13, Max Pressure-25, Rate-Auto, Rise Time 1. Use Sleepyhead and Encore Pro.

User avatar
coolbranch
Posts: 172
Joined: Sat Jul 12, 2014 9:59 am
Location: SE Alabama

Re: Contec CMS50I Pulse Oximeter

Post by coolbranch » Wed Mar 04, 2015 6:34 pm

Thanks M'ohms. Dummy me did not select the correct device from SH's drop down menu. I still had the CMS50e selected. I did what you said and the import started and finished. Funny thing though, no data showed up in the statistics or graphs. I deleted the device data and will try again tonight. I assume when I start the APAP machine I simply turn on record in the record menu of the CMS50I, right?

_________________
Machine: AirSense 11 Autoset
Mask: ResMed AirFit F30i Full Face CPAP Mask with Headgear
Additional Comments: OSCAR v1.4.0
I LOVE OSCAR

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Contec CMS50I Pulse Oximeter

Post by palerider » Wed Mar 04, 2015 7:27 pm

coolbranch wrote:Thanks M'ohms. Dummy me did not select the correct device from SH's drop down menu. I still had the CMS50e selected. I did what you said and the import started and finished. Funny thing though, no data showed up in the statistics or graphs. I deleted the device data and will try again tonight. I assume when I start the APAP machine I simply turn on record in the record menu of the CMS50I, right?
the I has a clock, and sleepyhead will sync that to your computer time, as long as you have the time offset properly set up for your prs1 drifting clock, then sleepyhead will sync the data from the i with your cpap data.

make a note of exactly what time you start the machine tonight, so you can get the cpap time offset correct.

then, yes, start recording when you go to bed, and stop when you're back awake.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
SGearhart
Posts: 434
Joined: Fri Nov 07, 2014 8:54 am
Location: Georgia

Re: Contec CMS50I Pulse Oximeter

Post by SGearhart » Wed Mar 04, 2015 7:37 pm

I start my 50I at the same time I start my S9. When I import the data I select the use the CPAP time option. If you go to bed after midnight, at least on a S9, your O2 import will show that it occurred on the current date and the CPAP data will be in the previous date. Even thought you select the CPAP time option, SH show no CPAP data for the current date to sync with. You have to select the previous date from the calendar to get to the CPAP data to show up.

_________________
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: CMS-50F & 50IW - SleepyHead 1.0.0-Beta

User avatar
coolbranch
Posts: 172
Joined: Sat Jul 12, 2014 9:59 am
Location: SE Alabama

Re: Contec CMS50I Pulse Oximeter

Post by coolbranch » Wed Mar 04, 2015 8:37 pm

10-4 on the great advice. I hope to follow up tomorrow and post a successful outcome here. Thanks again.

_________________
Machine: AirSense 11 Autoset
Mask: ResMed AirFit F30i Full Face CPAP Mask with Headgear
Additional Comments: OSCAR v1.4.0
I LOVE OSCAR

User avatar
coolbranch
Posts: 172
Joined: Sat Jul 12, 2014 9:59 am
Location: SE Alabama

Re: Contec CMS50I Pulse Oximeter

Post by coolbranch » Thu Mar 05, 2015 8:19 am

Thanks to you all, my oximeter CMS50I imported successfully into SH this morning. Looks like I have about a 5 minute drift forward on the PSR1 machine though. I will use computer (gov't) time tonight to calculate the exact time drift adjustment and enter into SH. This morning I used "match to cpap start time" to line up the pulse oximeter data on the SH graphs. Seems there is a lag on the CMS50I from when I select record to when the recording actually starts.

_________________
Machine: AirSense 11 Autoset
Mask: ResMed AirFit F30i Full Face CPAP Mask with Headgear
Additional Comments: OSCAR v1.4.0
I LOVE OSCAR

User avatar
ElvishKnight
Posts: 72
Joined: Thu Feb 12, 2015 3:34 pm
Location: Cedar Park Texas

Re: Contec CMS50I Pulse Oximeter

Post by ElvishKnight » Fri Mar 06, 2015 3:50 am

I feel like if I wore a oximeter to bed it would be off in 2 seconds with the way I sleep.

_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier
Mask: Quattro™ FX Full Face CPAP Mask with Headgear
Additional Comments: Been on CPAP for 10 years. No Ramp. No EPR. Minimum pressure setting 8. I use a ClimateLineAir Heated Tube. Software: Resmed myAir & SleepyHead.
If he had a name, it's long since been lost. He would have been known only as a servant of evil. One of a number. One of Nine.

User avatar
SGearhart
Posts: 434
Joined: Fri Nov 07, 2014 8:54 am
Location: Georgia

Re: Contec CMS50I Pulse Oximeter

Post by SGearhart » Fri Mar 06, 2015 8:58 am

coolbranch wrote:Thanks to you all, my oximeter CMS50I imported successfully into SH this morning. Looks like I have about a 5 minute drift forward on the PSR1 machine though. I will use computer (gov't) time tonight to calculate the exact time drift adjustment and enter into SH. This morning I used "match to cpap start time" to line up the pulse oximeter data on the SH graphs. Seems there is a lag on the CMS50I from when I select record to when the recording actually starts.
The SpO2 Assistant software can synchronize the 50I's time to your computer. Make sure that you sync you computer's time before you do this.
To synchronize the time do the following:
1. 50I plugged in to the computer and turned on.
2. Start SpO2 Assistant software.
3. Select the connect icon (looks like a RS-232 port).
4. Don't select the "View Device Stored Data" checkbox.
5. Select the "Connect" button. Note: If your speakers are on be prepared for an annoying pulse alarm.
6. Go to Options->Synchronize Device Time (lower portion of the options list)
7. Exit - SpO2 Assist will prompt you to save the real time file - select "No"; it will then exit.

I am uncertain how your unit adjusts its time. All I tell you is how I sync the Resmed with the 50I in hope that you can adjust this accordingly to your unit.
I turn on the 50I and go to the watch view (1 second press of the power button). I then enter the setup on my CPAP and go to the time setting. I adjust the time where it is 1 minute ahead of the 50I's time without pressing the enter button. I then wait until the time changes on the 50I and then I press the enter button on the CPAP. This puts me within a few seconds of each other.

I hope this information will get your time differences closer.

_________________
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: CMS-50F & 50IW - SleepyHead 1.0.0-Beta

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Contec CMS50I Pulse Oximeter

Post by palerider » Fri Mar 06, 2015 9:20 am

SGearhart wrote:The SpO2 Assistant software can synchronize the 50I's time to your computer.
sleepyhead sets the time on the 50i also, unless you uncheck the "set device date/time" box on the import screen.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

InsomniacGuy
Posts: 87
Joined: Sun Jan 25, 2015 4:14 pm

Re: Contec CMS50I Pulse Oximeter

Post by InsomniacGuy » Fri Mar 06, 2015 9:49 am

I called a US store selling the CMS 50F, and they said they don't offer the CMS 50I because it's not FDA approved. Is that true? With the online FDA database, I searched various forms of other Contec devices that are FDA approved and must not be using the correct format because no positive results were returned.

Is FDA approval even that important for this? Is the only way to purchase the 50I via Aliexpress? Those sellers on Aliexpress list the device as having FDA approval and the CE mark, but I find it odd that there are no US sellers for the 50I.

Thanks.

User avatar
coolbranch
Posts: 172
Joined: Sat Jul 12, 2014 9:59 am
Location: SE Alabama

Re: Contec CMS50I Pulse Oximeter

Post by coolbranch » Fri Mar 06, 2015 12:20 pm

Thanks for your detailed instruction SGearhart. I needed a great way to sync.

_________________
Machine: AirSense 11 Autoset
Mask: ResMed AirFit F30i Full Face CPAP Mask with Headgear
Additional Comments: OSCAR v1.4.0
I LOVE OSCAR

User avatar
coolbranch
Posts: 172
Joined: Sat Jul 12, 2014 9:59 am
Location: SE Alabama

Re: Contec CMS50I Pulse Oximeter

Post by coolbranch » Fri Mar 06, 2015 1:03 pm

Thanks palerider. Can I assume that SH is using computer clock time? If so, I see my prs1 auto's time is 6 min & 50 sec fast.

_________________
Machine: AirSense 11 Autoset
Mask: ResMed AirFit F30i Full Face CPAP Mask with Headgear
Additional Comments: OSCAR v1.4.0
I LOVE OSCAR

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Contec CMS50I Pulse Oximeter

Post by palerider » Fri Mar 06, 2015 3:59 pm

coolbranch wrote:Thanks palerider. Can I assume that SH is using computer clock time? If so, I see my prs1 auto's time is 6 min & 50 sec fast.
that's all it can use

the prs1, unfortunately, is set at the factory/repair depot, and then it just kinda drifts off further and further from correct as time goes by.

SH syncs the oximeter to computer time, and then uses the offset to sync the data from the prs1

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
coolbranch
Posts: 172
Joined: Sat Jul 12, 2014 9:59 am
Location: SE Alabama

Re: Contec CMS50I Pulse Oximeter (New Issue)

Post by coolbranch » Sat Mar 07, 2015 9:26 am

I started CPAP and CMS50I after midnight last night. The CPAP session imported as March 6...it was actually March 7. When trying to import ox/pulse data from the device, there are no CPAP sessions showing up and the calendar is blank. I tried importing from SPO2 program file and get a parsing error. Can't read file. Any ideas as to what is going on?

_________________
Machine: AirSense 11 Autoset
Mask: ResMed AirFit F30i Full Face CPAP Mask with Headgear
Additional Comments: OSCAR v1.4.0
I LOVE OSCAR