viewtopic/t101108/SleepyHead-0981-Testi ... ml#p943707footballgirl13 wrote:Error. Can't work due to VSVCR something or other. I can't paste the error in here. Oh , it says
MSVCR 120.dll is missing from your computer. Try reinstalling the. ..
It said this on BOTH broken and open.
Sleepy Head software?
Re: Sleepy Head software?
Current Settings PS 4.0 over 10.6-18.0 (cmH2O) - Resmed S9 VPAP Auto w/h5i Humidifier - Quattro Air FFM
TNET Sleep Resource Pages - CPAP Machine Database
Put your equip in your Signature - SleepyHead v1.0.0-beta-1
Kevin... alias Krelvin
TNET Sleep Resource Pages - CPAP Machine Database
Put your equip in your Signature - SleepyHead v1.0.0-beta-1
Kevin... alias Krelvin
- footballgirl13
- Posts: 92
- Joined: Tue Jan 27, 2015 1:50 pm
- Location: Inland Empire (California)
Re: Sleepy Head software?
I think I'm giving up. I can't see to get the help I need and I have done all I could. I installed this sleepyhead twice. I shouldn't have to install it for the rd time.
Why does this have to be so damn confusing and complicated? Why can't they make things to open up very easy instead of having to install and then extract files. So many steps just to use software. What's the freaking point of using the software if you can't get to it ?
What a waste of my time. I guess that is what I get using free stuff. It is too complicated. Forget it. I'll never know if my BiPap is doing any good. I'm so done with this stupid sleepyhead stuff. I will not post about this topic every again. I may not come back to this website either. Not sure to any users - just due to me being pissed off with this stupid sleepyhead crap. If it ain't easy I will not use it and if it is not easy I will not spend 5 hours trying to figure it out. This is why we have computer people to do this stuff. They need to make it EASY.
Why does this have to be so damn confusing and complicated? Why can't they make things to open up very easy instead of having to install and then extract files. So many steps just to use software. What's the freaking point of using the software if you can't get to it ?
What a waste of my time. I guess that is what I get using free stuff. It is too complicated. Forget it. I'll never know if my BiPap is doing any good. I'm so done with this stupid sleepyhead stuff. I will not post about this topic every again. I may not come back to this website either. Not sure to any users - just due to me being pissed off with this stupid sleepyhead crap. If it ain't easy I will not use it and if it is not easy I will not spend 5 hours trying to figure it out. This is why we have computer people to do this stuff. They need to make it EASY.
_________________
| Mask: Nuance & Nuance Pro Nasal Pillow CPAP Mask with Gel Nasal Pillows |
| Additional Comments: I don't use software yet, this should be an option in the dropdown box |
3rd sleep study Dec 2014:
1st half of study:
Total AHI: 18.4
RDI: 29.7
RERAs: 11.3
oxygen saturation below 90% during 0.4% of the time
Lowest oxygen: 84%
Weight
spring 2015: 202
fall 2015: 182 MBI 34
spring 2016: 194
fall 2016:
1st half of study:
Total AHI: 18.4
RDI: 29.7
RERAs: 11.3
oxygen saturation below 90% during 0.4% of the time
Lowest oxygen: 84%
Weight
spring 2015: 202
fall 2015: 182 MBI 34
spring 2016: 194
fall 2016:
Re: Sleepy Head software?
If you would read what I read in this thread I pointed you to where to get the missing file and told you which one to get and what to do.
This can be confusing for some people though. How about we let you use Encore Basic instead? It's a little easier to get installed.
EncoreGuide link this will show you examples of what you will see. For a different machine than yours but the basics are the same.
http://www.healthcare.philips.com/pwc_h ... _Guide.pdf
Send me a private message if you want it.
This can be confusing for some people though. How about we let you use Encore Basic instead? It's a little easier to get installed.
EncoreGuide link this will show you examples of what you will see. For a different machine than yours but the basics are the same.
http://www.healthcare.philips.com/pwc_h ... _Guide.pdf
Send me a private message if you want it.
_________________
| Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
| Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/ |
I may have to RISE but I refuse to SHINE.
- Tatooed Lady
- Posts: 984
- Joined: Sat Jun 07, 2014 6:18 pm
- Location: Central Wisconsin
Re: Sleepy Head software?
Enough. ENough. ENOUGH!!! Look, you asked a question, and got SEVERAL FORUM MEMBERS (INCLUDING PR) trying like hell to help you figure out your computer woes. And then when they didn't hand you the frigging Staples EZ BUTTON, you got mad, and whined how no one was HELPING you! If you're so computer challenged that you can't follow PUGSY's attempts to simplify things, you REALLY need to find someone IN PERSON to help you! Apparently, you don't take direction well. I asked for help when trying to get SH up and running, but my GOD, I didn't insult the folks who were trying to dumb things down for me!
And to try to guilt people, saying you won't use anything if it's not easy...honey, LIFE ain't easy.
I need a drink.
And to try to guilt people, saying you won't use anything if it's not easy...honey, LIFE ain't easy.
I need a drink.
_________________
| Mask: Swift™ FX For Her Nasal Pillow CPAP Mask with Headgear |
| Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control |
| Additional Comments: Precious and POW are very very good to me. |
As Bette Davis famously said, “Old age ain’t for sissies.”
I'm with the band.
So.Many.TOYS!
I'm with the band.
So.Many.TOYS!
- footballgirl13
- Posts: 92
- Joined: Tue Jan 27, 2015 1:50 pm
- Location: Inland Empire (California)
Re: Sleepy Head software?
Hope it is a good stiff drink. So you say I insult others that try to help me? But you cope with me (and I'm online no less) and your coping tool is - drinking? You need a drink. And I'm online. Is not this calling the kettle black? I was more frustrated with the software not working/not extracting right/not installing right and much less frustrated with what anyone posted. I don't need a drink. I'm going to bed.Tatooed Lady wrote:Enough. ENough. ENOUGH!!!
I need a drink.
_________________
| Mask: Nuance & Nuance Pro Nasal Pillow CPAP Mask with Gel Nasal Pillows |
| Additional Comments: I don't use software yet, this should be an option in the dropdown box |
3rd sleep study Dec 2014:
1st half of study:
Total AHI: 18.4
RDI: 29.7
RERAs: 11.3
oxygen saturation below 90% during 0.4% of the time
Lowest oxygen: 84%
Weight
spring 2015: 202
fall 2015: 182 MBI 34
spring 2016: 194
fall 2016:
1st half of study:
Total AHI: 18.4
RDI: 29.7
RERAs: 11.3
oxygen saturation below 90% during 0.4% of the time
Lowest oxygen: 84%
Weight
spring 2015: 202
fall 2015: 182 MBI 34
spring 2016: 194
fall 2016:
- Tatooed Lady
- Posts: 984
- Joined: Sat Jun 07, 2014 6:18 pm
- Location: Central Wisconsin
Re: Sleepy Head software?
My coping tool is telling you that you INSULTED people who were helping you. Actually, I DON'T drink, but damn, if I did, I'd have tied one on by now with all the back and forth DRIVEL.
Follow directions or don't. But holy crap...yeah, not much else to be said, as things might not be easy enough to slog through.
Follow directions or don't. But holy crap...yeah, not much else to be said, as things might not be easy enough to slog through.
_________________
| Mask: Swift™ FX For Her Nasal Pillow CPAP Mask with Headgear |
| Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control |
| Additional Comments: Precious and POW are very very good to me. |
As Bette Davis famously said, “Old age ain’t for sissies.”
I'm with the band.
So.Many.TOYS!
I'm with the band.
So.Many.TOYS!
Re: Sleepy Head software?
In the useful links in my signature line mentioned above, there's a help on downloading and installing SleepyHead.
http://www.apneaboard.com/wiki/index.ph ... Head_Notes
It lists
"DLL Errors
If you get a message something like "MSVCR120.dll is missing," you need to install the "Microsoft Visual C++ 2010 Redistributable Package." http://www.microsoft.com/en-us/download ... x?id=40784
Explanation: A file needed by the SleepyHead program was left out by accident for some versions. The Microsoft package above installs this DLL on your Windows computer.
It may not always be "MSVCR120.dll", but it will be something similar starting with "MSV" and ending with ".dll"."
http://www.apneaboard.com/wiki/index.ph ... Head_Notes
It lists
"DLL Errors
If you get a message something like "MSVCR120.dll is missing," you need to install the "Microsoft Visual C++ 2010 Redistributable Package." http://www.microsoft.com/en-us/download ... x?id=40784
Explanation: A file needed by the SleepyHead program was left out by accident for some versions. The Microsoft package above installs this DLL on your Windows computer.
It may not always be "MSVCR120.dll", but it will be something similar starting with "MSV" and ending with ".dll"."
_________________
| Mask: Swift™ FX Nasal Pillow CPAP Mask with Headgear |
| Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control |
| Additional Comments: Also SleepyHead, PRS1 Auto, Respironics Auto M series, Legacy Auto, and Legacy Plus |
Please enter your equipment in your profile so we can help you.
Click here for information on the most common alternative to CPAP.
If it's midnight and a DME tells you it's dark outside, go and check for yourself.
Useful Links.
Click here for information on the most common alternative to CPAP.
If it's midnight and a DME tells you it's dark outside, go and check for yourself.
Useful Links.
- Chairman Meow
- Posts: 53
- Joined: Thu Jan 22, 2015 4:13 pm
- Location: Sunshine State
Re: Sleepy Head software?
I feel kind of bad for resurrecting this topic, but I just want it to be clear that I've actually done all of the things suggested here. I was able to download the latest testing version of SleepyHead (and the earlier version), opened the compressed file, found the EXE file, got a .dll error message, downloaded the C++ package, and now I realize that my .dll error is not the MSV one - my .dll error message says "The program can't start because Qt5Core.dll is missing from your computer. Try reinstalling the program to fix this problem"
I'm not even close to a computer whiz... but I understand the basics, and even have a neighborhood child on standby in case she's needed (I'm totally kidding, she's my kid and she says she doesn't have a clue what the Qt5Core.dll is) - but I'm not sure if the program I'm supposed to 'reinstall' is the Sleepyhead program (which I have tried to 'reinstall several times now) or do I need to find this Qt5Core.dll?
Can someone help? I really want to play with graphs!!!
I'm not even close to a computer whiz... but I understand the basics, and even have a neighborhood child on standby in case she's needed (I'm totally kidding, she's my kid and she says she doesn't have a clue what the Qt5Core.dll is) - but I'm not sure if the program I'm supposed to 'reinstall' is the Sleepyhead program (which I have tried to 'reinstall several times now) or do I need to find this Qt5Core.dll?
Can someone help? I really want to play with graphs!!!
_________________
| Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear |
| Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control |
| Additional Comments: Untreated AHI of 124.9 - Fixed pressure of 12 - Sleepyhead software |
Britt~
Re: Sleepy Head software?
This happened to me when I tried to open it inside the zip file. Be sure and extract all the files from the 7 Zip into another folder. Then open that folder and click on the open or broken GL.
_________________
| Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
| Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear |
| Additional Comments: Sleepyhead Software |
ResMed Aircurve 10 VAUTO EPAP 11 IPAP 15 / P10 pillows mask / Sleepyhead Software / Back up & travel machine Respironics 760
- Chairman Meow
- Posts: 53
- Joined: Thu Jan 22, 2015 4:13 pm
- Location: Sunshine State
Re: Sleepy Head software?
I have to extract all the files to a separate folder? I was just extracting the .exe file and the only choice for where to put it was in my download file.
OMG it worked.
Waluigi - I can't believe it. Thank you so so so much. I was in despair... that I was never going to figure it out.
Thank you!
OMG it worked.
Waluigi - I can't believe it. Thank you so so so much. I was in despair... that I was never going to figure it out.