Understanding Flow Rate (ASV)

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
User avatar
Uncle Flapp
Posts: 247
Joined: Thu Nov 14, 2013 2:24 pm
Location: Arizona

Re: Understanding Flow Rate (ASV)

Post by Uncle Flapp » Wed Jan 28, 2015 11:01 am

Sleeprider wrote:Your machine is capable of dealing with that issue. Do you have timed breaths enabled?
Yes. Backup rate is set to auto. The ticks in the graph are the machine initiated breaths.

_________________
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Additional Comments: Dreamstation ASV. UPPP in 2007; Untreated AHI 84

User avatar
palerider
Posts: 32308
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Understanding Flow Rate (ASV)

Post by palerider » Wed Jan 28, 2015 11:24 am

Uncle Flapp wrote:Thanks for all the responses.

I see a more telling picture when I enable the Timed Breath flag .... the reason I am on ASV in the first place.
I thin you've found your answer, good uncle,

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
Uncle Flapp
Posts: 247
Joined: Thu Nov 14, 2013 2:24 pm
Location: Arizona

Re: Understanding Flow Rate (ASV)

Post by Uncle Flapp » Wed Jan 28, 2015 12:04 pm

palerider wrote:I thin you've found your answer, good uncle,
It sure looks that way. Now I know the "what". And to Greyghost's point, it very well could be related to REM - the "why".
Unable to let sleeping dogs lie, my only remaining question is in regards to how it impacts my health. I haven't measured my O2 since titration a year ago - it might be worth picking up an oximeter just to satisfy my curiosity.

_________________
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Additional Comments: Dreamstation ASV. UPPP in 2007; Untreated AHI 84

User avatar
palerider
Posts: 32308
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Understanding Flow Rate (ASV)

Post by palerider » Wed Jan 28, 2015 12:58 pm

Uncle Flapp wrote:
palerider wrote:I thin you've found your answer, good uncle,
It sure looks that way. Now I know the "what". And to Greyghost's point, it very well could be related to REM - the "why".
Unable to let sleeping dogs lie, my only remaining question is in regards to how it impacts my health. I haven't measured my O2 since titration a year ago - it might be worth picking up an oximeter just to satisfy my curiosity.
the cms50f and i are probably the nicest of the inexpensive monitors, under 100$ at aliexpress.com if you're not in a big hurry to get them. wrist worn, the I records multiple sessions. both import directly into sleepyhead.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
Uncle Flapp
Posts: 247
Joined: Thu Nov 14, 2013 2:24 pm
Location: Arizona

Re: Understanding Flow Rate (ASV)

Post by Uncle Flapp » Wed Jan 28, 2015 1:59 pm

Thanks, PR! You saved me the legwork of doing more research. I have Amazon Prime and went that route to get the CMS50F. Should be here by Friday.

I appreciate the help!

- Flappy

_________________
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Additional Comments: Dreamstation ASV. UPPP in 2007; Untreated AHI 84

icipher
Posts: 306
Joined: Thu Oct 10, 2013 11:09 am

Re: Understanding Flow Rate (ASV)

Post by icipher » Wed Jan 28, 2015 3:45 pm

Uncle Flapp wrote:Thanks, PR! You saved me the legwork of doing more research. I have Amazon Prime and went that route to get the CMS50F. Should be here by Friday.

I appreciate the help!

- Flappy
I never realized there was a wrist watch version, that is very cool and would probably be more accurate than being on your finger all night.

User avatar
Krelvin
Posts: 1977
Joined: Tue Jun 06, 2006 5:23 pm
Location: Metro Phx Area - Dry Heat!

Re: Understanding Flow Rate (ASV)

Post by Krelvin » Wed Jan 28, 2015 6:11 pm

icipher wrote:I never realized there was a wrist watch version, that is very cool and would probably be more accurate than being on your finger all night.
The wrist is where the monitor attached, the sensor itself is still on your finger.

Image
Last edited by Krelvin on Wed Jan 28, 2015 6:25 pm, edited 2 times in total.
Current Settings PS 4.0 over 10.6-18.0 (cmH2O) - Resmed S9 VPAP Auto w/h5i Humidifier - Quattro Air FFM
TNET Sleep Resource Pages - CPAP Machine Database
Put your equip in your Signature - SleepyHead v1.0.0-beta-1
Kevin... alias Krelvin

User avatar
palerider
Posts: 32308
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Understanding Flow Rate (ASV)

Post by palerider » Wed Jan 28, 2015 6:22 pm

Krelvin wrote:
icipher wrote:I never realized there was a wrist watch version, that is very cool and would probably be more accurate than being on your finger all night.
The wrist is where the monitor attached, the sensor itself is still on your finger.
it's better because the sensor is much lighter and stays on the finger better...

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
Morbius
Posts: 1040
Joined: Wed Jan 28, 2015 7:03 pm

Re: Understanding Flow Rate (ASV)

Post by Morbius » Wed Jan 28, 2015 7:06 pm

Uncle Flapp wrote:
palerider wrote:I thin you've found your answer, good uncle,
It sure looks that way. Now I know the "what".
Well, not if those are post-arousal centrals. If that's the case, what you've done is wipe out the compensation mechanism and created a problem, not solved it.

User avatar
palerider
Posts: 32308
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Understanding Flow Rate (ASV)

Post by palerider » Wed Jan 28, 2015 8:11 pm

Morbius wrote:
Uncle Flapp wrote:
palerider wrote:I thin you've found your answer, good uncle,
It sure looks that way. Now I know the "what".
Well, not if those are post-arousal centrals. If that's the case, what you've done is wipe out the compensation mechanism and created a problem, not solved it.
cite references please, mysterious person using his first post to make some unhelpful comment.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
Morbius
Posts: 1040
Joined: Wed Jan 28, 2015 7:03 pm

Re: Understanding Flow Rate (ASV)

Post by Morbius » Wed Jan 28, 2015 8:15 pm

palerider wrote:
Morbius wrote:
Uncle Flapp wrote:
palerider wrote:I thin you've found your answer, good uncle,
It sure looks that way. Now I know the "what".
Well, not if those are post-arousal centrals. If that's the case, what you've done is wipe out the compensation mechanism and created a problem, not solved it.
cite references please, mysterious person using his first post to make some unhelpful comment.
Or what?

User avatar
palerider
Posts: 32308
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Understanding Flow Rate (ASV)

Post by palerider » Wed Jan 28, 2015 9:04 pm

Morbius wrote:
palerider wrote:
Morbius wrote:
Uncle Flapp wrote:
palerider wrote:I thin you've found your answer, good uncle,
It sure looks that way. Now I know the "what".
Well, not if those are post-arousal centrals. If that's the case, what you've done is wipe out the compensation mechanism and created a problem, not solved it.
cite references please, mysterious person using his first post to make some unhelpful comment.
Or what?
or we'll ignore your noise.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
Morbius
Posts: 1040
Joined: Wed Jan 28, 2015 7:03 pm

Re: Understanding Flow Rate (ASV)

Post by Morbius » Wed Jan 28, 2015 9:38 pm

palerider wrote:or we'll ignore your noise.
OK.

User avatar
Julie
Posts: 20051
Joined: Tue Feb 28, 2006 12:58 pm

Re: Understanding Flow Rate (ASV)

Post by Julie » Wed Jan 28, 2015 11:22 pm

Do you have an oximeter you can wear and check in the am to see what took place overnight?
Last edited by Julie on Thu Jan 29, 2015 4:26 am, edited 1 time in total.

User avatar
palerider
Posts: 32308
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Understanding Flow Rate (ASV)

Post by palerider » Thu Jan 29, 2015 12:06 am

Julie wrote:Do you have an oximeter you can wear overnight and check in the am to see what took place overnight?
he's ordered one should have it in a couple days.... yay for data and less guessing!

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.