Re: Looking For Tips On How To Interact With The Doctor

Post by palerider » Thu Jan 15, 2015 1:44 pm

novatom wrote:This machine only tracks and shows compliance data through the on machine display or through the Encore Software.

Hmmm.. is it that last sentence that you were referring to? Because if it is, it's wrong. I AM getting detailed data, I know it via Sleepyhead.
the machine you linked to is several years old, and the bottom of the line model.

when you get home, check the number on the bottom, and then pick the appropriate one from the list for your equipment setting here.
460 PR System One REMStar 60 Series Pro CPAP Machine
560 PR System One REMStar 60 Series Auto CPAP Machine
760 PR System One REMStar 60 Series BiPAP Auto with Bi-Flex
960 PR System One REMStar 60 Series BiPAP autoSV Advanced
660 PR System One REMStar 60 Series BiPAP Pro with Bi-Flex
260 PR System One REMstar 60 Series Plus CPAP Machine with C-Flex

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
Mike208
Posts: 100
Joined: Tue Dec 30, 2014 2:59 pm
Location: Upstate SC.

Re: Looking For Tips On How To Interact With The Doctor

Post by Mike208 » Thu Jan 15, 2015 2:00 pm

Julie wrote:Hi - while Cpap may be making it obvious, is it possible you have issues with dairy or gluten that have not been obvious til now but that Cpap has made so? Worth looking into?
LOL Thanks. Way ahead of you. I have tried gluten, dairy, wheat, the whole shebang. In my situation, a given food will be fine and not bother me. All of a sudden, it does. And then it won't. The only thing I have found that makes it worse is heavily spiced and fried foods. I avoid them as much as possible.

User avatar
Julie
Posts: 20052
Joined: Tue Feb 28, 2006 12:58 pm

Re: Looking For Tips On How To Interact With The Doctor

Post by Julie » Thu Jan 15, 2015 2:31 pm

When's the last time you had an endoscopy?

User avatar
Mike208
Posts: 100
Joined: Tue Dec 30, 2014 2:59 pm
Location: Upstate SC.

Re: Looking For Tips On How To Interact With The Doctor

Post by Mike208 » Thu Jan 15, 2015 2:54 pm

Julie wrote:When's the last time you had an endoscopy?
I had one in 2012. They did a dilation. No serious issues, no hernia was noted.

User avatar
Mike208
Posts: 100
Joined: Tue Dec 30, 2014 2:59 pm
Location: Upstate SC.

Re: Looking For Tips On How To Interact With The Doctor

Post by Mike208 » Thu Jan 15, 2015 2:56 pm

49er wrote:
It's amazing to me how some in the health care profession can be so unwilling to accept that, maybe, just maybe, something they have not ever seen is happening. You're right, it shouldn't matter. He basically told me I was wrong. I believe it's coming to an end with him. I have not heard back yet but I have a feeling he will say there is nothing else that he can do, I have failed CPAP, and refer me back to the ENT. That's what I think is going to happen.
Not to sound like a broken record but it is time to see another sleep doctor.
I hate it but you're probably correct. I found this in an older thread and thought it was interesting:




Re: I Can Say One Thing About Aerophagia

Postby SacramentoGranny on Sun Mar 08, 2009 1:21 am
One thing I can say about aerophagia...DON'T GIVE UP TRYING TO FIND A SOLUTION. After almost two years of suffering from what I would call severe aerophagia (due to my hiatal hernia) I finally took things in my own hands. I had been getting the same stupid answers...raise your head (just means the gas comes out but will resume going back in), take GasEx (worthless), etc. I changed masks but all the time it was the straight CPAP machine that was the problem. I read what other patients had done to alleviate the gas and figured an auto-titrating machine might do the trick.

Of course my initial request (through my doctor) for an auto-titrating machine was turned down. I was told I could "grieve" the decision and I did. Their answer was to send me to the HMO's sleep lab (newly expanded), which tested me on several machines and eventually prescribed a RemStar Series M Auto BiPAP. Only one problem -- it was not set on "auto", just BiPAP. The straight BiPAP continued to shove air into my stomach during inhalation so the aerophia persisted.

So back to the Sleep Lab I went to request another machine. This time an auto titrating machine was prescribed (similar to the one I had originally requested last fall) but DME drug their heels. In the meantime I noted the "auto" on my machine and realized (DUH) that WAS an auto-titrating machine. After some research and purchase of the programming directions (isn't EBAY wonderful) I set the thing on auto and have been in heaven ever since. My min and max settings were the same as the straight CPAP (even went to 12 instead of 11 to insure I was getting relief). Now that I can also monitor my progress I have noted that my AHI average is 1.4 (was diagnosed at 36 before treatment). After waking up every one to two hours before using the "new" machine I am beginning to sleep for four and five hours at a time...and hopefully will soon be asleep all night.

I did fax the Sleep Lab with the programmed settings so they would be aware of what I had done. Having heard nothing in the meantime I assume that I am headed in the right direction.

From where I see things now, the best treatment for aerophagia is an auto-titrating machine, preferably with BiPAP (worked best for me,anyway).

User avatar
Mike208
Posts: 100
Joined: Tue Dec 30, 2014 2:59 pm
Location: Upstate SC.

Re: Looking For Tips On How To Interact With The Doctor

Post by Mike208 » Thu Jan 15, 2015 3:16 pm

Just got a call from the sleep center. She said my machine has APAP capabilities and they thought it might help. So, she went in and made the changes via the modem and we'll see what happens. She said the machine will start out at 4 and adjust as I need it after I'm asleep. Sound feasible?

Sleeprider
Posts: 1562
Joined: Tue May 06, 2008 5:57 pm
Location: Murrysville, PA

Re: Looking For Tips On How To Interact With The Doctor

Post by Sleeprider » Thu Jan 15, 2015 3:23 pm

Mike208 wrote:Just got a call from the sleep center. She said my machine has APAP capabilities and they thought it might help. So, she went in and made the changes via the modem and we'll see what happens. She said the machine will start out at 4 and adjust as I need it after I'm asleep. Sound feasible?
She turned on ramp, and put your machine in auto-mode. So your new settings are 4.0 for minimum pressure, and we don't know what they programmed for max pressure. Most likely they left the ramp set a 45 minute default. If you feel starved for air or experience high initial AHIs, you will want to raise the minimum pressure and/or shorten the duration of ramp.

As you can see, they are guessing at what to do, which is a great equalizer. You now know they may know less than you do.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Additional Comments: Sleepyhead software. Just changed from PRS1 BiPAP Auto DS760TS

User avatar
Mike208
Posts: 100
Joined: Tue Dec 30, 2014 2:59 pm
Location: Upstate SC.

Re: Looking For Tips On How To Interact With The Doctor

Post by Mike208 » Thu Jan 15, 2015 3:34 pm

Sleeprider wrote:
Mike208 wrote:Just got a call from the sleep center. She said my machine has APAP capabilities and they thought it might help. So, she went in and made the changes via the modem and we'll see what happens. She said the machine will start out at 4 and adjust as I need it after I'm asleep. Sound feasible?
She turned on ramp, and put your machine in auto-mode. So your new settings are 4.0 for minimum pressure, and we don't know what they programmed for max pressure. Most likely they left the ramp set a 45 minute default. If you feel starved for air or experience high initial AHIs, you will want to raise the minimum pressure and/or shorten the duration of ramp.

As you can see, they are guessing at what to do, which is a great equalizer. You now know they may know less than you do.
The max pressure is 10. Sorry, forgot that part. Yep, they're guessing. I could tell that in her voice. She wasn't quite as confident as before.

User avatar
Julie
Posts: 20052
Joined: Tue Feb 28, 2006 12:58 pm

Re: Looking For Tips On How To Interact With The Doctor

Post by Julie » Thu Jan 15, 2015 3:40 pm

4 is the default low machine setting and no one on earth can breathe at it... 6 or 7 being the first reasonable low pressure to start with... plus if you use the ramp, never mind at 45 mins(!) you won't be being treated at all as there's nowhere for the ramp to reach if 4 is the setting entered! Plus 45 mins (even if your low setting were e.g. 12) means you're not getting full treatment for all that time, which is why 90% of us turn the ramp off in the first week of using Cpap.

User avatar
Mike208
Posts: 100
Joined: Tue Dec 30, 2014 2:59 pm
Location: Upstate SC.

Re: Looking For Tips On How To Interact With The Doctor

Post by Mike208 » Thu Jan 15, 2015 3:44 pm

Julie wrote:4 is the default low machine setting and no one on earth can breathe at it... 6 or 7 being the first reasonable low pressure to start with... plus if you use the ramp, never mind at 45 mins(!) you won't be being treated at all as there's nowhere for the ramp to reach if 4 is the setting entered! Plus 45 mins (even if your low setting were e.g. 12) means you're not getting full treatment for all that time, which is why 90% of us turn the ramp off in the first week of using Cpap.
You are correct. And that's why I expect the AHI to go way up. Maybe. We'll see. The other night, when I had the pressure at 10 the AHI was well over 40. Last night the machine was just one number lower and the AHI was low.

HoseCrusher
Posts: 2744
Joined: Tue Oct 12, 2010 6:42 pm

Re: Looking For Tips On How To Interact With The Doctor

Post by HoseCrusher » Thu Jan 15, 2015 6:16 pm

Offering a suggestion...

Set your minimum pressure to 6 and adjust the ramp to 5 - 10 minutes. 5 minutes works great if you fall asleep immediately. 10 minutes if you don't.

If you leave it at 4 with a 45 minute ramp, I would suggest you mask up, turn the machine on, then read a few chapters in a book so the machine can get up to a pressure that you can breath at...

_________________
Mask: Brevida™ Nasal Pillow CPAP Mask with Headgear
Additional Comments: Machine is an AirSense 10 AutoSet For Her with Heated Humidifier.
SpO2 96+% and holding...

User avatar
Mike208
Posts: 100
Joined: Tue Dec 30, 2014 2:59 pm
Location: Upstate SC.

Re: Looking For Tips On How To Interact With The Doctor

Post by Mike208 » Thu Jan 15, 2015 7:09 pm

I'll try it where it's set first and if I can't breathe I may adjust it. Thanks.

User avatar
archangle
Posts: 9293
Joined: Sun Mar 27, 2011 11:55 am

Re: Looking For Tips On How To Interact With The Doctor

Post by archangle » Fri Jan 16, 2015 3:44 am

OkyDoky wrote:If that is the machine, it states it is a 250 which if you read at the bottom states only tracks AHI and compliance data which means no detailed waveforms. So check the number tonight when you get home.
The 250 machine does not record any AHI data, only settings and the times you used it.

_________________
Mask: Swift™ FX Nasal Pillow CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: Also SleepyHead, PRS1 Auto, Respironics Auto M series, Legacy Auto, and Legacy Plus
Please enter your equipment in your profile so we can help you.
Click here for information on the most common alternative to CPAP.
If it's midnight and a DME tells you it's dark outside, go and check for yourself.

Useful Links.

User avatar
novatom
Posts: 78
Joined: Mon Dec 29, 2014 10:41 am

Re: Looking For Tips On How To Interact With The Doctor

Post by novatom » Fri Jan 16, 2015 7:02 am

I checked the bottom of my machine last night and this is the version:

460 PR System One REMStar 60 Series Pro CPAP Machine
Actually it says "REMStar Pro, DOM"
REF 460P
The date on it is 2014-10-14

The top of the machine says REMStar Pro C-Flex+ System One

When I google it I come up with this description:
http://www.cpapsupplyusa.com/PR-System- ... -CPAP.aspx
And it says "Respironics has re-designed the REMstar Pro 60 Series"

I guess this must be the updated version. I wonder what DOM means (data only machine?)

Anyways, I will now update my profile. Thanks for your help!

_________________
Mask: Nuance & Nuance Pro Nasal Pillow CPAP Mask with Gel Nasal Pillows
Additional Comments: 460 PR System One REMStar 60 Series Pro CPAP Machine

User avatar
OkyDoky
Posts: 2870
Joined: Mon Aug 25, 2014 5:18 pm

Re: Looking For Tips On How To Interact With The Doctor

Post by OkyDoky » Fri Jan 16, 2015 9:20 am

archangle wrote:
OkyDoky wrote:If that is the machine, it states it is a 250 which if you read at the bottom states only tracks AHI and compliance data which means no detailed waveforms. So check the number tonight when you get home.
The 250 machine does not record any AHI data, only settings and the times you used it.
Thanks for correction. I even read that statement, but since they were getting AHI, I read AHI into the statement. Does the AHI even show on the display window?
ResMed Aircurve 10 VAUTO EPAP 11 IPAP 15 / P10 pillows mask / Sleepyhead Software / Back up & travel machine Respironics 760