Need advice

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
bruskee
Posts: 14
Joined: Wed Nov 12, 2014 12:26 pm

Need advice

Post by bruskee » Thu Jan 08, 2015 1:37 pm

Been lurking around here for a while learning as much as possible. Got my apap and been using it for the past 17 nights. The dr wanted me to have cpap with a pressure of 10 but I told him I wanted apap (thanks to advise I read here) and he wrote me up with a pressure of 6-12. I've been creeping up slowly on the low number and have been at the 11-14 for the past 3 nights with my erp at 1. I'm sleeping fine but I'm still having Ca's and Oa's and still feel the foggy head and as tired as before I started. Do you think I should bump it up more to clear the Oa's or give it time.??

Thanks for all the info you have posted here, I'd of been blind without it !

Here's last nights report.

Image
Last edited by bruskee on Thu Jan 08, 2015 2:28 pm, edited 1 time in total.

VictorC
Posts: 62
Joined: Thu Oct 16, 2014 3:54 pm
Location: Belleville, IL

Re: Need advice

Post by VictorC » Thu Jan 08, 2015 1:50 pm

I wouldn't worry about your CAs. The number of them is quite low, and more pressure will not help with them, since they are not caused by an obstruction.

_________________
Machine: PR System One REMStar 60 Series Auto CPAP Machine
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Additional Comments: Settings: APAP at 10.5-14cm H2O, AFLEX @ 2, Humidifier @ 2, Hose Temperature @ 3, No Ramp

bruskee
Posts: 14
Joined: Wed Nov 12, 2014 12:26 pm

Re: Need advice

Post by bruskee » Thu Jan 08, 2015 2:48 pm

Thanks, I was asking if I should bump it up to alleviate the Obstructive ?

User avatar
palerider
Posts: 32308
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Need advice

Post by palerider » Thu Jan 08, 2015 3:12 pm

bruskee wrote:Thanks, I was asking if I should bump it up to alleviate the Obstructive ?
unfortunately, when you shrink the screenshot down so it looks pretty on the forum, it becomes fairly useless for actually trying to figure out what's going on.

based on the pressure curve, (though I can't see how well it tracks with OAs) I'd say 'yes' to your question, just keep an eye on the CAs and make sure they don't start increasing more and more as the pressure is raised.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
Julie
Posts: 20051
Joined: Tue Feb 28, 2006 12:58 pm

Re: Need advice

Post by Julie » Thu Jan 08, 2015 3:19 pm

There's no leak rate on that graph, so can't see if you're losing air, possibly mouth breathing, so you might want to post those results.

bruskee
Posts: 14
Joined: Wed Nov 12, 2014 12:26 pm

Re: Need advice

Post by bruskee » Thu Jan 08, 2015 3:29 pm

Palerider thank you, and what's the easiest way to get the correct size for the forum ? Julie there were no Leaks last night, I'll post that pic when I get home, thank you ..

User avatar
Wulfman...
Posts: 6694
Joined: Mon Sep 01, 2014 6:41 pm
Location: Nearest fishing spot

Re: Need advice

Post by Wulfman... » Thu Jan 08, 2015 5:07 pm

I would suggest increasing the minimum by 1 cm. and decreasing the maximum by 1 cm.


Den

.
(5) REMstar Autos w/C-Flex & (6) REMstar Pro 2 CPAPs w/C-Flex - Pressure Setting = 14 cm.
"Passover" Humidification - ResMed Ultra Mirage FF - Encore Pro w/Card Reader & MyEncore software - Chiroflow pillow
User since 05/14/05

User avatar
palerider
Posts: 32308
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Need advice

Post by palerider » Thu Jan 08, 2015 5:21 pm

bruskee wrote:Palerider thank you, and what's the easiest way to get the correct size for the forum ?
the *best* way I've found is to upload the full size screen shot (there's a guide floating around on the forum created by someone having a period of lunacy where they suggest such things as pasting into word and resizing and other bs) to imgur.com. then imgur gives you a linked bbcode string to paste into the forum here.

there's a trick with imgur that I like to do.

Code: Select all

pick the 'linked bbcode entry, and change this:
[url=http://imgur.com/1TScxYz][img]http://i.imgur.com/1TScxYz.png[/img][/url]
to this:
[url=http://imgur.com/1TScxYz.png][img]http://i.imgur.com/1TScxYzl.png[/img][/url]
(add .png after the first link, and change the second one by adding a l *before* the .png)
then you get a large thumbnail pasted in the forum here, and when you click it, you get the full res pic.

if you don't want to go to the trouble of that minor edit, then just use the 'bbcode' link, it'll post a huge pic here, and it's ugly, but people can right click on it and see the whole pic, and be able to read the numbers and such.... much better.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

Guest 2

Re: Need advice

Post by Guest 2 » Thu Jan 08, 2015 5:38 pm

palerider wrote:
bruskee wrote:Thanks, I was asking if I should bump it up to alleviate the Obstructive ?
unfortunately, when you shrink the screenshot down so it looks pretty on the forum, it becomes fairly useless for actually trying to figure out what's going on.

based on the pressure curve, (though I can't see how well it tracks with OAs) I'd say 'yes' to your question, just keep an eye on the CAs and make sure they don't start increasing more and more as the pressure is raised.

You can use your Zoom feature ( control + to zoom in and control - to zoom out ) try it some time It works good !
Sleep Well

bruskee
Posts: 14
Joined: Wed Nov 12, 2014 12:26 pm

Re: Need advice

Post by bruskee » Thu Jan 08, 2015 6:11 pm

Image

Damn Palerider that's pretty sweet... Didn't get the thumbnail but better ! Thx

User avatar
palerider
Posts: 32308
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Need advice

Post by palerider » Thu Jan 08, 2015 6:29 pm

Guest 2 wrote:
palerider wrote:
bruskee wrote:Thanks, I was asking if I should bump it up to alleviate the Obstructive ?
unfortunately, when you shrink the screenshot down so it looks pretty on the forum, it becomes fairly useless for actually trying to figure out what's going on.
You can use your Zoom feature ( control + to zoom in and control - to zoom out ) try it some time It works good !
Sleep Well
believe it or not, drive by snarker, blowing up a low res picture does NOT (no matter what you see on CSI Apneaville, "enhance and bring out detail", it just makes small fuzzy into large fuzzy.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
palerider
Posts: 32308
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Need advice

Post by palerider » Thu Jan 08, 2015 6:34 pm

bruskee wrote:Image

Damn Palerider that's pretty sweet... Didn't get the thumbnail but better ! Thx
well, that's cuz you messed up the lower case ell in the instructions. it should have been this:

Image

hit 'reply' to see what the differences in what you did and what I corrected it to look like (you put .1png, I put l.png inside the img tag.)

next time you post one, turn off the calendar (or hit f12 to take the screenshot) and move the mask pressure graph out of the way so that we can see event flags, flow, pressure, leak, snore, and flow limitations. I think you're either getting snoring or flow limitations (which drive the pressure up rapidly) when you're under around 12cm.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
cathyf
Posts: 515
Joined: Mon Sep 01, 2014 11:56 am

Re: Need advice

Post by cathyf » Thu Jan 08, 2015 7:49 pm

Looking at the pressure graph, it doesn't look like the machine wants to go up that high. On an auto pap, the upper pressure is about preventing the machine from going up there if it wants to, because the high pressure is causing you some problem. If the machine doesn't want to go there, it doesn't really matter what your upper pressure is. One of the reasons for limiting your machine is that higher pressures can cause centrals, and looking at that one graph you have some centrals at pressure peaks. But that could be a coincidence, too, which you could tell by looking over lots of nights.

When you have lower pressures allowing obstructive events, and higher pressures causing centrals, then you have to set your pressures to balance out the two. Otherwise, if you are NOT having problems with high pressures, you need to get your minimum pressure high enough that when your breathing starts going south (for example, you enter REM sleep) you need to be able to work up to the higher pressure in time to keep you asleep and prevent events. And the upper pressure set so that the machine can do what it wants.

Guest 2

Re: Need advice

Post by Guest 2 » Thu Jan 08, 2015 7:57 pm

believe it or not, drive by snarker, blowing up a low res picture does NOT (no matter what you see on CSI Apneaville, "enhance and bring out detail", it just makes small fuzzy into large fuzzy.[/quote]

Wow your a little beside yourself, was only trying to help, by no means did I mean to offend you.

Snarker

bruskee
Posts: 14
Joined: Wed Nov 12, 2014 12:26 pm

Re: Need advice

Post by bruskee » Thu Jan 08, 2015 8:30 pm

Ok Palerider, how's this ??? Appreciate you taking time to explain..

Image