Adjust pressure REMstar Plus C-Flex

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
Mattkcc
Posts: 3
Joined: Fri Dec 12, 2014 6:54 pm

Adjust pressure REMstar Plus C-Flex

Post by Mattkcc » Fri Dec 12, 2014 7:00 pm

I need to reduce the pressure on my Remstar Plus C-Flex. I found a site that showed how to do it for a ton of cpap's but not my model. The reason I need to change is when my allergies are bad I need to change from a nasal pillow to a face mask.
Thanks, Matt

User avatar
palerider
Posts: 32308
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Adjust pressure REMstar Plus C-Flex

Post by palerider » Fri Dec 12, 2014 7:17 pm

your throat needs the same pressure whether it's full faced or nasal.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
RicaLynn
Posts: 526
Joined: Fri Aug 15, 2014 4:23 pm
Location: Western MT

Re: Adjust pressure REMstar Plus C-Flex

Post by RicaLynn » Fri Dec 12, 2014 8:10 pm

What PR means is that you don't need to adjust the pressure to suit the mask. You may, however, need to change the mask setting (I don't have a Respironics, I have a ResMed, and it has a setting to adapt to the mask type)

_________________
Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear
Additional Comments: Backup/travel unit is an identical S9 AutoSet for Her w/Eson nasal mask
Last edited by RicaLynn on Fri Dec 12, 2014 8:21 pm, edited 1 time in total.

User avatar
kaiasgram
Posts: 3569
Joined: Sat Jun 02, 2012 2:08 pm
Location: Northern California

Re: Adjust pressure REMstar Plus C-Flex

Post by kaiasgram » Fri Dec 12, 2014 8:16 pm

You can obtain the clinical manual from the Apnea Board website -- use the link in my signature box below.

_________________
Machine: AirSense 10 AutoSet with Heated Humidifer + Aifit N30i Nasal Mask Bundle
Additional Comments: SleepyHead-now-OSCAR software on Mac OSX Ventura

Mattkcc
Posts: 3
Joined: Fri Dec 12, 2014 6:54 pm

Re: Adjust pressure REMstar Plus C-Flex

Post by Mattkcc » Fri Dec 12, 2014 9:37 pm

RicaLynn wrote:What PR means is that you don't need to adjust the pressure to suit the mask. You may, however, need to change the mask setting (I don't have a Respironics, I have a ResMed, and it has a setting to adapt to the mask type)
My mask doesn't have adjustable settings, my specialist advised me at my last appointment to decrease the pressure two psi when I use it with the nasal pillow and up two psi when I use the mask. He had showed me how to do it on a cpap in the office however I couldn't get it to work when I got home. I ordered the manual on the site so I should be good to go. The current psi is just to strong when I use the nasal pillow but fine when I use a mask. I prefer the nasal pillow due to not being able to find a mask that works well without excessive leaking especially around the eyes. I've tried on a bunch of masks took two different ones home. They seem great in the office but at home in bed not so good. Sleeping on my side doesn't help matters, I have never been able to sleep on my back.

User avatar
OkyDoky
Posts: 2869
Joined: Mon Aug 25, 2014 5:18 pm

Re: Adjust pressure REMstar Plus C-Flex

Post by OkyDoky » Fri Dec 12, 2014 10:01 pm

Respironics do not have mask settings like Resmed but they have resistance settings. These settings correlate to Respironic masks but it is more difficult to determine for ResMed masks. If your tech told you to change a number with different masks this may be the number they were talking about. I don't change the number when I switch because it makes little difference.

If you are talking about your pressure settings I would question that without data measurement. What pressure are you on?
ResMed Aircurve 10 VAUTO EPAP 11 IPAP 15 / P10 pillows mask / Sleepyhead Software / Back up & travel machine Respironics 760

Mattkcc
Posts: 3
Joined: Fri Dec 12, 2014 6:54 pm

Re: Adjust pressure REMstar Plus C-Flex

Post by Mattkcc » Fri Dec 12, 2014 10:24 pm

I'm on 11 psi, my mask is a Simplus and has no adjustment. 11 psi is fine with the mask but with the pillow it's just to strong for me, my doctor recommended to try 9 psi.

User avatar
OkyDoky
Posts: 2869
Joined: Mon Aug 25, 2014 5:18 pm

Re: Adjust pressure REMstar Plus C-Flex

Post by OkyDoky » Fri Dec 12, 2014 10:33 pm

kaiasgram wrote:You can obtain the clinical manual from the Apnea Board website -- use the link in my signature box below.
Go to the bottom of kaisagram's post where there is a link to Apnea Board. Click on it. There you can find and download the clinical manual but you will need to know your model number which is on the bottom of the blower. Be sure the humidifer is removed before you turn your blower over.

Also, the mask adjustments we are talking about are not on the mask but on the machine.
ResMed Aircurve 10 VAUTO EPAP 11 IPAP 15 / P10 pillows mask / Sleepyhead Software / Back up & travel machine Respironics 760

tan
Posts: 565
Joined: Sat Aug 30, 2014 11:58 am

Re: Adjust pressure REMstar Plus C-Flex

Post by tan » Fri Dec 12, 2014 10:53 pm

Mattkcc wrote:I'm on 11 psi, my mask is a Simplus and has no adjustment. 11 psi is fine with the mask but with the pillow it's just to strong for me, my doctor recommended to try 9 psi.
It is not psi, it is cm of H2O

FredToo
Posts: 10
Joined: Sun Dec 14, 2014 7:18 pm

Re: Adjust pressure REMstar Plus C-Flex

Post by FredToo » Sun Dec 14, 2014 9:55 pm

I'm not sure why people are being resistant to answering your very simple question, but I will bite.

There are two basic setting systems on the REMstar...one is for the user (can change the settings daily if you like), and the other is for the vendor (for initial one time setup).

Several of the options are duplicated in both symptoms...such as whether to use the ramping option, or auto on/off, etc. But the air flow setting is not supposed to be available to the patient...this would be similar to having the pharmacy change your dosage on a prescription. I say horse pookey! It's your machine and you need to know how to make the adjustments that need making. But I would advise that you make only minor adjustments to this setting, and I bet after trying a couple of settings you will likely return to where you are now.

Ok , off my soapbox and time for the simple answer. There are only two buttons on the top of your machine...the main on/off button wheel, and the ramp button. Turn the wheel to "Setup" and then press both buttons simultaneously for about 5 seconds...release when you will hear the beep. You will see several new setup options that are not available to you normally. Some of them you will want to read about in the manual as to what they are.

After you have made your selections, turn the wheel back to "Therapy" and press the wheel as you always do. The new options should be working. If you made mistakes, go back and undo them.

Its really very simple, so good luck,
FredToo

User avatar
Wulfman...
Posts: 6689
Joined: Mon Sep 01, 2014 6:41 pm
Location: Nearest fishing spot

Re: Adjust pressure REMstar Plus C-Flex

Post by Wulfman... » Mon Dec 15, 2014 12:25 am

Mattkcc wrote:I need to reduce the pressure on my Remstar Plus C-Flex. I found a site that showed how to do it for a ton of cpap's but not my model. The reason I need to change is when my allergies are bad I need to change from a nasal pillow to a face mask.
Thanks, Matt
You probably need the instructions I posted in this link long ago. One set is for the Plus model with C-Flex and the other is for the Plus model without C-Flex.

viewtopic/t102640/viewtopic.php?f=1&t=2 ... th#p227091


Den

.
(5) REMstar Autos w/C-Flex & (6) REMstar Pro 2 CPAPs w/C-Flex - Pressure Setting = 14 cm.
"Passover" Humidification - ResMed Ultra Mirage FF - Encore Pro w/Card Reader & MyEncore software - Chiroflow pillow
User since 05/14/05

User avatar
Tatooed Lady
Posts: 984
Joined: Sat Jun 07, 2014 6:18 pm
Location: Central Wisconsin

Re: Adjust pressure REMstar Plus C-Flex

Post by Tatooed Lady » Mon Dec 15, 2014 5:56 am

FredToo wrote:I'm not sure why people are being resistant to answering your very simple question, but I will bite.

There are two basic setting systems on the REMstar...one is for the user (can change the settings daily if you like), and the other is for the vendor (for initial one time setup).

Several of the options are duplicated in both symptoms...such as whether to use the ramping option, or auto on/off, etc. But the air flow setting is not supposed to be available to the patient...this would be similar to having the pharmacy change your dosage on a prescription. I say horse pookey! It's your machine and you need to know how to make the adjustments that need making. But I would advise that you make only minor adjustments to this setting, and I bet after trying a couple of settings you will likely return to where you are now.

Ok , off my soapbox and time for the simple answer. There are only two buttons on the top of your machine...the main on/off button wheel, and the ramp button. Turn the wheel to "Setup" and then press both buttons simultaneously for about 5 seconds...release when you will hear the beep. You will see several new setup options that are not available to you normally. Some of them you will want to read about in the manual as to what they are.

After you have made your selections, turn the wheel back to "Therapy" and press the wheel as you always do. The new options should be working. If you made mistakes, go back and undo them.

Its really very simple, so good luck,
FredToo
I don't think they're being tight with the info for kicks. More likely the reasoning posted right off didn't sound like the OP is aware of how pressure works in different masks. It looked goofy to me, too...but what do I know?

_________________
Mask: Swift™ FX For Her Nasal Pillow CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: Precious and POW are very very good to me.
As Bette Davis famously said, “Old age ain’t for sissies.”
I'm with the band.
So.Many.TOYS!

User avatar
OkyDoky
Posts: 2869
Joined: Mon Aug 25, 2014 5:18 pm

Re: Adjust pressure REMstar Plus C-Flex

Post by OkyDoky » Mon Dec 15, 2014 2:03 pm

Tatooed Lady wrote:
FredToo wrote:I'm not sure why people are being resistant to answering your very simple question, but I will bite.

There are two basic setting systems on the REMstar...one is for the user (can change the settings daily if you like), and the other is for the vendor (for initial one time setup).

Several of the options are duplicated in both symptoms...such as whether to use the ramping option, or auto on/off, etc. But the air flow setting is not supposed to be available to the patient...this would be similar to having the pharmacy change your dosage on a prescription. I say horse pookey! It's your machine and you need to know how to make the adjustments that need making. But I would advise that you make only minor adjustments to this setting, and I bet after trying a couple of settings you will likely return to where you are now.

Ok , off my soapbox and time for the simple answer. There are only two buttons on the top of your machine...the main on/off button wheel, and the ramp button. Turn the wheel to "Setup" and then press both buttons simultaneously for about 5 seconds...release when you will hear the beep. You will see several new setup options that are not available to you normally. Some of them you will want to read about in the manual as to what they are.

After you have made your selections, turn the wheel back to "Therapy" and press the wheel as you always do. The new options should be working. If you made mistakes, go back and undo them.

Its really very simple, so good luck,
FredToo
I don't think they're being tight with the info for kicks. More likely the reasoning posted right off didn't sound like the OP is aware of how pressure works in different masks. It looked goofy to me, too...but what do I know?
Yes, thanks Tatooed Lady. I just wanted to be sure they new what they were changing and why, especially when the question related more to masks and there was no mentioning of data monitoring. The Clinician's Manuel has the info.
ResMed Aircurve 10 VAUTO EPAP 11 IPAP 15 / P10 pillows mask / Sleepyhead Software / Back up & travel machine Respironics 760

Zetwitch5368
Posts: 1
Joined: Sun Feb 21, 2021 1:27 pm

Re: Adjust pressure REMstar Plus C-Flex

Post by Zetwitch5368 » Sun Feb 21, 2021 1:31 pm

kaiasgram wrote:
Fri Dec 12, 2014 8:16 pm
You can obtain the clinical manual from the Apnea Board website -- use the link in my signature box below.
I see no link. Can you provide one?

User avatar
Pugsy
Posts: 65400
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Adjust pressure REMstar Plus C-Flex

Post by Pugsy » Sun Feb 21, 2021 1:44 pm

Zetwitch5368 wrote:
Sun Feb 21, 2021 1:31 pm
kaiasgram wrote:
Fri Dec 12, 2014 8:16 pm
You can obtain the clinical manual from the Apnea Board website -- use the link in my signature box below.
I see no link. Can you provide one?
https://www.apneaboard.com/adjust-cpap- ... tup-manual

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.