Oh No- Machine not writing to SD card?

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
User avatar
coolbranch
Posts: 176
Joined: Sat Jul 12, 2014 9:59 am
Location: SE Alabama

Oh No- Machine not writing to SD card?

Post by coolbranch » Wed Dec 03, 2014 7:54 pm

Quick checked the CPAP display this morning and read therapy hours and AHI. All looked correct compared to previous day sleep session. The problem is when I plugged in the SD card in computer and imported into Sleepyhead I get nothing.
Steps I took:
1. Formatted another SD card and inserted into CPAP and allowed time to export data to card. Seemed like it took longer than usual.
2. Inserted that card into computer and attempted to import into SH. Still nothing.
3. The CPAP machine display shows the previous nights sleep info but may not be writing to SD card.

Tried another card. Storage space on SD cards have plenty of space. I will try again tonight but could my CPAP machine stop writing to SD card. Has anyone had this problem?

Oh another observation, when checking the timestamp on the latest file on the SD card, the date is Dec 4. It is actually Dec 3 and the sleep session occurred beginning Dec 2, which would be the date I would direct to Sleepyhead.

Could it be possible that the CPAP machine has lost track of the date?

_________________
Machine: AirSense 11 Autoset
Mask: ResMed AirFit F30i Full Face CPAP Mask with Headgear
Additional Comments: OSCAR v1.4.0
I LOVE OSCAR

User avatar
Pugsy
Posts: 65413
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Oh No- Machine not writing to SD card?

Post by Pugsy » Wed Dec 03, 2014 8:23 pm

It's unlikely that the machine has lost track of the date. It has an internal clock with an internal battery and at most what happens is it starts losing time and not gaining.

Which SleepyHead version are you using and is it the Windows version or Mac version.

If you see a file dated 12/4 that is odd but if there is a file there the SleepyHead should find it..
How many files do you see in the P0 folder?
Tell me the steps you are doing when you import and what message are you getting...if any.
Sounds like SleepyHead is probably the one that is confused. If you could get me a copy of your SD card contents I can use my SleepyHead as well as Encore and see if anything is there. Easiest way is to zip/compress the entire SD card and upload it to some place like dropbox..or send it to me as an attachment in an email. If you want to use personal email...send me a private message and I will send you my private email address.

\

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
palerider
Posts: 32308
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Oh No- Machine not writing to SD card?

Post by palerider » Wed Dec 03, 2014 8:24 pm

coolbranch wrote:Oh another observation, when checking the timestamp on the latest file on the SD card, the date is Dec 4. It is actually Dec 3 and the sleep session occurred beginning Dec 2, which would be the date I would direct to Sleepyhead.

Could it be possible that the CPAP machine has lost track of the date?
prs1 machines are set to GMT, so it probably already IS dec4th as far as the machine is concerned.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
coolbranch
Posts: 176
Joined: Sat Jul 12, 2014 9:59 am
Location: SE Alabama

Re: Oh No- Machine not writing to SD card?

Post by coolbranch » Wed Dec 03, 2014 8:48 pm

Looking at the p1 folder on the SD card, each date's import has a file extension of .001, .002 and .005 each with the identical dates and timestamps.

Last nights write to the SD card is interesting. The .001 and .005 files are dated 12/3/2014 and have timestamp of 11:07 AM but the .002 file is dated 12/4/2014 and its timestamp is 1:18 AM.

I scrolled back several weeks and last night is the only write to the card that is inconsistent with date and time. A glitch?

I am using the lastest beta of SH.

SleepyHead v0.9.8-1 (testing)

Build Date: Oct 3 2014 17:17:35
Revision: 4595153 (master branch)
Graphics Engine: OpenGL
Data Folder Location: C:\Users\wrsherrill\Dropbox\SleepyHeadData-Testing

I place SH in Dropbox so I can view SH on multiple boxes. Mark Watkins tipped me off that he does this.

Thank you for your help.

_________________
Machine: AirSense 11 Autoset
Mask: ResMed AirFit F30i Full Face CPAP Mask with Headgear
Additional Comments: OSCAR v1.4.0
I LOVE OSCAR

User avatar
coolbranch
Posts: 176
Joined: Sat Jul 12, 2014 9:59 am
Location: SE Alabama

Re: Oh No- Machine not writing to SD card?

Post by coolbranch » Wed Dec 03, 2014 8:49 pm

Using Win 7.

_________________
Machine: AirSense 11 Autoset
Mask: ResMed AirFit F30i Full Face CPAP Mask with Headgear
Additional Comments: OSCAR v1.4.0
I LOVE OSCAR

Guest

Re: Oh No- Machine not writing to SD card?

Post by Guest » Wed Dec 03, 2014 8:55 pm

coolbranch wrote:The CPAP machine display shows the previous nights sleep info but may not be writing to SD card.
Exactly when are you removing the card from the machine?

User avatar
coolbranch
Posts: 176
Joined: Sat Jul 12, 2014 9:59 am
Location: SE Alabama

Re: Oh No- Machine not writing to SD card?

Post by coolbranch » Wed Dec 03, 2014 9:01 pm

Maybe I snatched the card too soon after turning off the CPAP this morning? Don't know how that would affect the timestamp. It certainly may affect premature import of the machine data. However, I took the card back to the machine this afternoon to try again. I also unplugged the CPAP for a while and tried to no avail.

_________________
Machine: AirSense 11 Autoset
Mask: ResMed AirFit F30i Full Face CPAP Mask with Headgear
Additional Comments: OSCAR v1.4.0
I LOVE OSCAR

User avatar
Pugsy
Posts: 65413
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Oh No- Machine not writing to SD card?

Post by Pugsy » Wed Dec 03, 2014 9:41 pm

There's nothing wrong with the machine.
I have your SleepyHead data in front of me and that includes last night and it is showing 12/2 as it should.
The one file with the weird date...I don't know why but it doesn't seem to be hurting anything.

All I did was unzip it and tell SleepyHead to go to the unzipped P-Series folder and it imported straight away without issues.
Everything is all there.

Sounds like SleepyHead needs to be told again where to find the P-Series folder or you are directing it too far into the folder...needs to be the P-Series folder...and don't do what I did a while back and try it with P series folder (without the hyphen) cause that won't work.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
coolbranch
Posts: 176
Joined: Sat Jul 12, 2014 9:59 am
Location: SE Alabama

Re: Oh No- Machine not writing to SD card?

Post by coolbranch » Wed Dec 03, 2014 9:55 pm

What was AHI on the last night?

_________________
Machine: AirSense 11 Autoset
Mask: ResMed AirFit F30i Full Face CPAP Mask with Headgear
Additional Comments: OSCAR v1.4.0
I LOVE OSCAR

User avatar
Pugsy
Posts: 65413
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Oh No- Machine not writing to SD card?

Post by Pugsy » Wed Dec 03, 2014 9:59 pm

Your AHI last night...0.51 ....pretty darn good.

Image

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
coolbranch
Posts: 176
Joined: Sat Jul 12, 2014 9:59 am
Location: SE Alabama

Re: Oh No- Machine not writing to SD card?

Post by coolbranch » Wed Dec 03, 2014 10:05 pm

That data is dead on according to the CPAP machine display. Thank you so much for your assistance. Since it is bedtime and tomorrow comes early, I will have to tinker with it tomorrow. I will probably reboot myself and the computer and get everything corrected then. Talk to you soon. Thanks again!

_________________
Machine: AirSense 11 Autoset
Mask: ResMed AirFit F30i Full Face CPAP Mask with Headgear
Additional Comments: OSCAR v1.4.0
I LOVE OSCAR

User avatar
coolbranch
Posts: 176
Joined: Sat Jul 12, 2014 9:59 am
Location: SE Alabama

Re: Oh No- Machine not writing to SD card?

Post by coolbranch » Thu Dec 04, 2014 5:13 am

Oh well. I imported Dec 3 (last nights) data into SH without a hitch this morning. Never was able to get SH to pick up Tuesday night's data. I have copy of SD card with all and up to Tuesday night and tried to pick up Tuesday night again. Sleepyhead message states it already has Tues night, but program will not show it.

I'll probably never know and chalk it up to a glitch and move on. It seems to be working now. Thanks for everyone's input.

_________________
Machine: AirSense 11 Autoset
Mask: ResMed AirFit F30i Full Face CPAP Mask with Headgear
Additional Comments: OSCAR v1.4.0
I LOVE OSCAR

User avatar
coolbranch
Posts: 176
Joined: Sat Jul 12, 2014 9:59 am
Location: SE Alabama

Re: Oh No- Machine not writing to SD card?

Post by coolbranch » Thu Dec 04, 2014 5:44 am

Sleepmapper picked up the missing night from SD card but SH will not.

_________________
Machine: AirSense 11 Autoset
Mask: ResMed AirFit F30i Full Face CPAP Mask with Headgear
Additional Comments: OSCAR v1.4.0
I LOVE OSCAR

Guest

Re: Oh No- Machine not writing to SD card?

Post by Guest » Thu Dec 04, 2014 7:29 am

coolbranch wrote:Sleepmapper picked up the missing night from SD card but SH will not.
I suspect it is something you are doing. You need to tell us in detail what steps you take and when in detail.

User avatar
Pugsy
Posts: 65413
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Oh No- Machine not writing to SD card?

Post by Pugsy » Thu Dec 04, 2014 8:04 am

Try creating a new profile in Sleepyhead and importing to the new profile.

I am using the same version of SleepyHead you are using and I have Win 7 on my computer...and it worked for me.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.