Black ring around face

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
gilmour
Posts: 11
Joined: Thu Nov 27, 2014 5:24 am

Black ring around face

Post by gilmour » Thu Nov 27, 2014 5:34 am

My cpap machine is sitting on my night stand just above our forced air furnace register. Not sure where the black ring is coming from on my face just around my beard at the bottom of my mask. My filter is also black. Cannot wash the black outof the filter either. Any ideas??

Guest

Re: Black ring around face

Post by Guest » Thu Nov 27, 2014 5:53 am

gilmour wrote:My cpap machine is sitting on my night stand just above our forced air furnace register. Not sure where the black ring is coming from on my face just around my beard at the bottom of my mask. My filter is also black. Cannot wash the black outof the filter either. Any ideas??
Yea sure, if the filter it black it is the air coming into the machine that is contaminated.

Which make/model machine do you have?

gilmour
Posts: 11
Joined: Thu Nov 27, 2014 5:24 am

Re: Black ring around face

Post by gilmour » Thu Nov 27, 2014 6:00 am

I will post what type of machine I have soon.

User avatar
VVV
Posts: 542
Joined: Mon Mar 14, 2011 11:28 am

Re: Black ring around face

Post by VVV » Thu Nov 27, 2014 8:16 am

gilmour wrote:Any ideas??
Yes. Call a trusted HVAC man.
.....................................V

User avatar
Jack_Russel_Fan
Posts: 104
Joined: Wed Apr 09, 2014 2:45 am
Location: Ellijay, GA
Contact:

Re: Black ring around face

Post by Jack_Russel_Fan » Thu Nov 27, 2014 9:10 am

I do not want to overly alarm you, but if you have a gas or oil furnace you need to have your furnace checked immediately, or shut it down until you can! If what you are seeing is soot, your furnace could be producing deadly carbon monoxide.


Tom - Retired heating and AC contractor.

_________________
Mask: TAP PAP Nasal Pillow CPAP Mask with Improved Stability Mouthpiece
Additional Comments: CMS-50F Pulse Oximeter, O2 when sleeping, SleepyHead, Sleep Master
Author of Sleep Master wireless xpap data transfer program.
Machine: PRS1 760P

gilmour
Posts: 11
Joined: Thu Nov 27, 2014 5:24 am

Re: Black ring around face

Post by gilmour » Thu Nov 27, 2014 9:45 am

That is what I thought. I already had an experience with carbon monoxide poisoning at my mothers home. The registers and walls turn black. We do not have that here. Other posts said it was possibly mold spores, dust or even from burning candles. My follow up visit is afound the corner and I need to discuss these matters with them. Any help or same experiences would be most helpful.
We also have a dog that sheds a lot and creates dust.

Guest

Re: Black ring around face

Post by Guest » Thu Nov 27, 2014 9:49 am

gilmour wrote:My follow up visit is afound the corner and I need to discuss these matters with them.
"They" will only tell you it is not the machine.

User avatar
englandsf
Posts: 1016
Joined: Thu Jul 10, 2014 6:17 pm
Location: Austin TX

Re: Black ring around face

Post by englandsf » Thu Nov 27, 2014 10:13 am

What happens after one night if you replace the filter?

_________________
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: Sleepyhead SW. NeilMed and Alkalol Nasal rinses. Veramyst. AutoPAP 11-20 cms. Started June '14, untreated AHI 31-38, with PAP around 1.

gilmour
Posts: 11
Joined: Thu Nov 27, 2014 5:24 am

Re: Black ring around face

Post by gilmour » Thu Nov 27, 2014 11:55 am

I know this is bad. That is the same filter since I started the cpap. I will be getting more on Monday.

User avatar
BlackSpinner
Posts: 9742
Joined: Sat Apr 25, 2009 5:44 pm
Location: Edmonton Alberta
Contact:

Re: Black ring around face

Post by BlackSpinner » Thu Nov 27, 2014 2:06 pm

gilmour wrote:I know this is bad. That is the same filter since I started the cpap. I will be getting more on Monday.
What you are needing to do on Monday is get the furnace problem fixed! Everything you see in the filter you are breathing into your lungs during the day!!!!!

_________________
Machine: PR System One REMStar 60 Series Auto CPAP Machine
Additional Comments: Quatro mask for colds & flus S8 elite for back up
71. The lame can ride on horseback, the one-handed drive cattle. The deaf, fight and be useful. To be blind is better than to be burnt on the pyre. No one gets good from a corpse. The Havamal

gilmour
Posts: 11
Joined: Thu Nov 27, 2014 5:24 am

Re: Black ring around face

Post by gilmour » Thu Nov 27, 2014 5:53 pm

We have 3 carbon monoxide detecters in our home. So I am ruling out the furnace is at fault. I will get new filters tomorrow and see what happens then.

Guest

Re: Black ring around face

Post by Guest » Thu Nov 27, 2014 7:04 pm

gilmour wrote:We have 3 carbon monoxide detecters in our home. So I am ruling out the furnace is at fault. I will get new filters tomorrow and see what happens then.
Burning candles are ya?

gilmour
Posts: 11
Joined: Thu Nov 27, 2014 5:24 am

Re: Black ring around face

Post by gilmour » Thu Nov 27, 2014 7:15 pm

Yes we are burning candles. Hard to believe soot from them would get in our bedroom.

Guest

Re: Black ring around face

Post by Guest » Thu Nov 27, 2014 7:32 pm

gilmour wrote:Yes we are burning candles. Hard to believe soot from them would get in our bedroom.
Ok then.... believe it or not. How many do you burn?

gilmour
Posts: 11
Joined: Thu Nov 27, 2014 5:24 am

Re: Black ring around face

Post by gilmour » Thu Nov 27, 2014 7:47 pm

2~~~~~both in the living room.